This info is the first night using my new full face mask, Amera View. My old mask, DreamWear leaked when I opened my mouth while sleeping, so doc ordered it. Can someone tell me if the new mask is working properly? I think the settings are okay, but not sure. The tech increased the pressure from 5-8 to 5-10 recently.

AHIApnea Hypopnea Index 2.17
ResMed S9
AutoSetModel 36005 - 23131184160
PAP Mode: APAP
Min 5 Max 10 (cmH2O)
Date Sleep Wake Hours
3/27/2017 16:44:59 09:25:56 08:17:07
Large Leak 0.25%
Clear Airway 0.00
Obstructive 1.45
Unclassified Apnea 0.00
Hypopnea 0.72

Event Breakdown

Statistics
Channel Min Med 95% Max
PressureTherapy Pressure (cmH2O)
W-Avg: 8.21 5.00 8.82 10.00 10.00
EPAPExpiratory Pressure (cmH2O)
W-Avg: 5.53 4.00 5.82 7.00 7.00
Minute Vent.Amount of air displaced per minute (L/min)
W-Avg: 6.49 1.25 6.25 8.62 21.62
Resp. RateRate of breaths per minute (Breaths/min)
W-Avg: 18.03 2.00 18.00 22.00 47.00
Flow Limit.Graph showing severity of flow limitations (Severity (0-1))
W-Avg: 0.05 0.00 0.00 0.36 0.97
Leak RateRate of detected mask leakage (L/min)
W-Avg: 2.24 0.00 1.20 6.00 68.40
SnoreGraph displaying snore volume (?)
W-Avg: 0.03 0.00 0.04 0.06 0.40
Insp. TimeTime taken to breathe in (Seconds)
W-Avg: 1.59 0.04 1.50 2.22 9.96
Exp. TimeTime taken to breathe out (Seconds)
W-Avg: 1.81 0.04 1.80 2.38 4.36
Tidal VolumeAmount of air displaced per breath (ml)
W-Avg: 356.77 60.00 340.00 520.00 1180.00

Total time in apnea 00:06:20
Time over leak redline 0.255%
Machine Settings

ModeCPAP Mode APAP
Pressure MinMinimum Therapy Pressure 5 cmH2O
Pressure MaxMaximum Therapy Pressure 10 cmH2O
EPRResMed Exhale Pressure Relief Full Time
EPR LevelExhale Pressure Relief Level 3cmH2O
Session Information



CPAP Sessions
On Date Start End Duration
ResMed Session #1490647498
3/27/2017 16:44:59 17:05:01 00h 20m 02s
ResMed Session #1490648771
3/27/2017 17:06:11 17:11:13 00h 05m 02s
ResMed Session #1490650683
3/27/2017 17:38:04 18:44:06 01h 06m 02s
ResMed Session #1490683195
3/28/2017 02:39:55 09:25:56 06h 46m 01s
20170328_023955_PLD.edf
20170328_023955_BRP.edf
20170328_023955_SAD.edf
20170328_023955_EVE.edf
STR.edf

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Humidifier: XT Heated Humidifier
Additional Comments: My headgear varies (STILL!)
Last edited by DeeCPAP on Tue Mar 28, 2017 10:38 am, edited 1 time in total.
Resmed S9 with humidifier and in need of the right mask.

User avatar
Pugsy
Posts: 64153
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Data From New Mask

Post by Pugsy » Tue Mar 28, 2017 10:35 am

DeeCPAP wrote:Large Leak 0.25%
In terms of large leak or time spent in large leak....not even 1% of the night was spent in large leak.
So sealing very well.

In terms of AHI....very decent therapy. You may or may not end up wanting a slight increase in the therapy pressure but at this point I wouldn't do anything.
DeeCPAP wrote:Clear Airway 0.00
Obstructive 1.45
Unclassified Apnea 0.00
Hypopnea 0.72
So yeah..looks like it is doing what it is supposed to do.

How was your sleep quality last night? Wake often or sleep soundly? It took me a few nights to get the point where I was sleeping soundly with the Amara View because I wasn't used to having so much stuff on my face but once I did I slept fairly decently with it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Data From New Mask

Post by DeeCPAP » Tue Mar 28, 2017 10:45 am

DeeCPAP wrote:This info is the first night using my new full face mask, Amera View. My old mask, DreamWear leaked when I opened my mouth while sleeping, so doc ordered it. Can someone tell me if the new mask is working properly? I think the settings are okay, but not sure. The tech increased the pressure from 5-8 to 5-10 recently.

AHIApnea Hypopnea Index 2.17
ResMed S9
AutoSetModel 36005 - 23131184160
PAP Mode: APAP
Min 5 Max 10 (cmH2O)
Date Sleep Wake Hours
3/27/2017 16:44:59 09:25:56 08:17:07
Large Leak 0.25%
Clear Airway 0.00
Obstructive 1.45
Unclassified Apnea 0.00
Hypopnea 0.72

Event Breakdown

Statistics
Channel Min Med 95% Max
PressureTherapy Pressure (cmH2O)
W-Avg: 8.21 5.00 8.82 10.00 10.00
EPAPExpiratory Pressure (cmH2O)
W-Avg: 5.53 4.00 5.82 7.00 7.00
Minute Vent.Amount of air displaced per minute (L/min)
W-Avg: 6.49 1.25 6.25 8.62 21.62
Resp. RateRate of breaths per minute (Breaths/min)
W-Avg: 18.03 2.00 18.00 22.00 47.00
Flow Limit.Graph showing severity of flow limitations (Severity (0-1))
W-Avg: 0.05 0.00 0.00 0.36 0.97
Leak RateRate of detected mask leakage (L/min)
W-Avg: 2.24 0.00 1.20 6.00 68.40
SnoreGraph displaying snore volume (?)
W-Avg: 0.03 0.00 0.04 0.06 0.40
Insp. TimeTime taken to breathe in (Seconds)
W-Avg: 1.59 0.04 1.50 2.22 9.96
Exp. TimeTime taken to breathe out (Seconds)
W-Avg: 1.81 0.04 1.80 2.38 4.36
Tidal VolumeAmount of air displaced per breath (ml)
W-Avg: 356.77 60.00 340.00 520.00 1180.00

Total time in apnea 00:06:20
Time over leak redline 0.255%
Machine Settings

ModeCPAP Mode APAP
Pressure MinMinimum Therapy Pressure 5 cmH2O
Pressure MaxMaximum Therapy Pressure 10 cmH2O
EPRResMed Exhale Pressure Relief Full Time
EPR LevelExhale Pressure Relief Level 3cmH2O
Session Information



CPAP Sessions
On Date Start End Duration
ResMed Session #1490647498
3/27/2017 16:44:59 17:05:01 00h 20m 02s
ResMed Session #1490648771
3/27/2017 17:06:11 17:11:13 00h 05m 02s
ResMed Session #1490650683
3/27/2017 17:38:04 18:44:06 01h 06m 02s
ResMed Session #1490683195
3/28/2017 02:39:55 09:25:56 06h 46m 01s
20170328_023955_PLD.edf
20170328_023955_BRP.edf
20170328_023955_SAD.edf
20170328_023955_EVE.edf
STR.edf
THANK YOU, PUGSY! As for sleeping, yes, it was okay (so was sleeping with the DreamWeaver but the data showed large leaks. However, the mask is tight, so it made the tip of my nose red and it'll probably will stay red for hours due to sensitive skin. I'm afraid to try a larger cushion because the rest of the mask may then be too loose (and leak). Is there a solution? Thanks

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Humidifier: XT Heated Humidifier
Additional Comments: My headgear varies (STILL!)
Resmed S9 with humidifier and in need of the right mask.

User avatar
Pugsy
Posts: 64153
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Data From New Mask

Post by Pugsy » Tue Mar 28, 2017 10:55 am

A little bit of Lansinoh lanolin ointment/cream on the tip of the nose area ...maybe?
Fiddle with the strap tension just a little. It shouldn't be horribly tight to seal. I found that it sealed quite well with minimal tension.
Too tight can cause the skin to become abraded and you don't want that.

Don't discount trying a different size cushion too quickly...it might be just the ticket to fitting better.
From what I have read the frame part of the mask is universal size and the next size large is just a little "deeper" and it might not touch the tip of your nose in the same way.

Did you get the fit pack with all sizes of cushions or just the one cushion size?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

prodigyplace
Posts: 1739
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: Data From New Mask

Post by prodigyplace » Tue Mar 28, 2017 11:01 am

What size cushion do you have? I have an Amara View that does not work for me. I have the small & medium masks, both very lightly used.
If it would not freak you out, I might me able to send you the Medium mask for the cost of shipping.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Data From New Mask

Post by DeeCPAP » Tue Mar 28, 2017 12:03 pm

Pugsy wrote:A little bit of Lansinoh lanolin ointment/cream on the tip of the nose area ...maybe?
Fiddle with the strap tension just a little. It shouldn't be horribly tight to seal. I found that it sealed quite well with minimal tension.
Too tight can cause the skin to become abraded and you don't want that.

Don't discount trying a different size cushion too quickly...it might be just the ticket to fitting better.
From what I have read the frame part of the mask is universal size and the next size large is just a little "deeper" and it might not touch the tip of your nose in the same way.

Did you get the fit pack with all sizes of cushions or just the one cushion size?
I didn't get the fit pack. Doc simply ordered "medium". I don't know why docs don't know there tends to be an
issue with sizes and should order the pack, but they're not trained well enough to know -- a real shame.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Humidifier: XT Heated Humidifier
Additional Comments: My headgear varies (STILL!)
Resmed S9 with humidifier and in need of the right mask.

User avatar
Pugsy
Posts: 64153
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Data From New Mask

Post by Pugsy » Tue Mar 28, 2017 12:08 pm

Did your mask package come with the little guide for measuring optimal cushion size?
If not I think that there is one you can print off and check sizing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Data From New Mask

Post by DeeCPAP » Wed Mar 29, 2017 10:11 am

Pugsy wrote:A little bit of Lansinoh lanolin ointment/cream on the tip of the nose area ...maybe?
Fiddle with the strap tension just a little. It shouldn't be horribly tight to seal. I found that it sealed quite well with minimal tension.
Too tight can cause the skin to become abraded and you don't want that.

Don't discount trying a different size cushion too quickly...it might be just the ticket to fitting better.
From what I have read the frame part of the mask is universal size and the next size large is just a little "deeper" and it might not touch the tip of your nose in the same way.

Did you get the fit pack with all sizes of cushions or just the one cushion size?

This is NEWER data from last night. I think this mask (medium) didn't work as well as the SMALL, but the small felt too tight.


ResMed S9
AutoSetModel 36005 - 23131184160
PAP Mode: APAP
Min 5 Max 10 (cmH2O)
Date Sleep Wake Hours
3/29/2017 03:33:36 09:40:15 06:04:08
Large Leak 3.73%
Clear Airway 0.16
Obstructive 3.30
Unclassified Apnea 0.00
Hypopnea 0.82

Event Breakdown

Statistics
Channel Min Med 95% Max
PressureTherapy Pressure (cmH2O)
W-Avg: 8.06 5.00 8.44 10.00 10.00
EPAPExpiratory Pressure (cmH2O)
W-Avg: 5.40 4.00 5.44 7.00 7.00
Minute Vent.Amount of air displaced per minute (L/min)
W-Avg: 5.94 0.00 5.62 8.75 22.25
Resp. RateRate of breaths per minute (Breaths/min)
W-Avg: 18.61 0.00 18.20 22.80 50.00
Flow Limit.Graph showing severity of flow limitations (Severity (0-1))
W-Avg: 0.04 0.00 0.00 0.31 0.93
Leak RateRate of detected mask leakage (L/min)
W-Avg: 8.91 0.00 8.40 22.80 78.00
SnoreGraph displaying snore volume (?)
W-Avg: 0.05 0.00 0.04 0.08 0.26
Insp. TimeTime taken to breathe in (Seconds)
W-Avg: 1.59 0.10 1.54 2.08 15.36
Exp. TimeTime taken to breathe out (Seconds)
W-Avg: 1.72 0.04 1.72 2.18 3.44
Tidal VolumeAmount of air displaced per breath (ml)
W-Avg: 310.59 0.00 300.00 440.00 1160.00

Total time in apnea 00:11:25
Time over leak redline 3.726%
Machine Settings

ModeCPAP Mode APAP
Pressure MinMinimum Therapy Pressure 5 cmH2O
Pressure MaxMaximum Therapy Pressure 10 cmH2O
EPRResMed Exhale Pressure Relief Full Time
EPR LevelExhale Pressure Relief Level 3cmH2O
Session Information



CPAP Sessions
On Date Start End Duration
ResMed Session #1490772815
3/29/2017 03:33:36 04:07:38 00h 34m 02s
20170329_033336_PLD.edf
20170329_033335_BRP.edf
20170329_033336_SAD.edf
STR.edf
ResMed Session #1490774927
3/29/2017 04:08:47 06:17:49 02h 09m 02s
20170329_040847_PLD.edf
20170329_040847_BRP.edf
20170329_040847_SAD.edf
20170329_040847_EVE.edf
STR.edf
ResMed Session #1490782705
3/29/2017 06:18:25 09:24:27 03h 06m 02s
20170329_061825_PLD.edf
20170329_061825_BRP.edf
20170329_061825_SAD.edf
20170329_061825_EVE.edf
STR.edf
ResMed Session #1490793912
3/29/2017 09:25:13 09:40:15 00h 15m 02s
20170329_092513_PLD.edf
20170329_092512_BRP.edf
20170329_092513_SAD.edf
STR.edf

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Humidifier: XT Heated Humidifier
Additional Comments: My headgear varies (STILL!)
Resmed S9 with humidifier and in need of the right mask.

User avatar
Pugsy
Posts: 64153
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Data From New Mask

Post by Pugsy » Wed Mar 29, 2017 10:21 am

You had more large leak time last night than the night before which may or may not be related to mask size used.
We can have some fairly significant variations in any of the results without ever changing a thing.
I might have 5% of the night in large leak one night and 15% the next night and the next night never go into large leak territory and never change a thing.
This is why we try to get more than one night of data before we go changing anything.....so we can get an overall pattern or trend ideas as to what is going on.

Use the size that is most comfortable for you...and use it for several nights and then see how the leaks are trending.
The change in size may or may not have been the cause of the change in leak numbers.
It's still not a bad number though...3 % of the night in large leak...not the end of the world at all.
6 hours...3 % would be approx 10 minutes...not the end of the world at all as long as you slept well and leaks didn't wake you up.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Data From New Mask

Post by DeeCPAP » Wed Mar 29, 2017 4:18 pm

Pugsy wrote:You had more large leak time last night than the night before which may or may not be related to mask size used.
We can have some fairly significant variations in any of the results without ever changing a thing.
I might have 5% of the night in large leak one night and 15% the next night and the next night never go into large leak territory and never change a thing.
This is why we try to get more than one night of data before we go changing anything.....so we can get an overall pattern or trend ideas as to what is going on.

Use the size that is most comfortable for you...and use it for several nights and then see how the leaks are trending.
The change in size may or may not have been the cause of the change in leak numbers.
It's still not a bad number though...3 % of the night in large leak...not the end of the world at all.
6 hours...3 % would be approx 10 minutes...not the end of the world at all as long as you slept well and leaks didn't wake you up.
Thanks, Pugsy. As for how I felt today, it felt like I hadn't slept at all, but it just may be the stress I'm under. I'll keep giving it time and upload more numbers. Thanks!

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Humidifier: XT Heated Humidifier
Additional Comments: My headgear varies (STILL!)
Resmed S9 with humidifier and in need of the right mask.

mrbreezeet1
Posts: 223
Joined: Mon Mar 06, 2017 7:35 pm

Re: Data From New Mask

Post by mrbreezeet1 » Mon Apr 17, 2017 10:09 pm

Pugsy wrote:Did your mask package come with the little guide for measuring optimal cushion size?
If not I think that there is one you can print off and check sizing.
There is one on line, there is a "quarter within a credit card" size test also.
But the measuring guide basically just measures the length from your lip, to the tip of your nose.
I have the medium, but seems to need to be a little wider at the ends of my mouth.
She is bringing a large, but looks to me it will not be any wider.
I'll say it again, neither the doctor, or the DME (hope that correct?) seem to be helping me at all.
I have learned more here.
Didn't think there were going to be all these issues.

Sent from my MotoG3 using Tapatalk