BiPAP not picking up inspiration sometimes

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
unputdownnable
Posts: 15
Joined: Wed Jan 03, 2024 3:04 pm

BiPAP not picking up inspiration sometimes

Post by unputdownnable » Wed Jan 03, 2024 3:11 pm

Hi! I'm migrating from CPAP to BiPAP and have been struggling with the device not picking up my inspiration and not triggering the IPAP often. Could this be a breathing problem or a device issue? I initially noticed a clogged nostril but even after decongestion it kept happening.
Last edited by unputdownnable on Wed Jan 03, 2024 10:00 pm, edited 1 time in total.

User avatar
lazarus
Posts: 1305
Joined: Sat Mar 12, 2022 3:21 pm
Location: Just above NJ.

Re: BiPAP not picking up inspiration sometimes

Post by lazarus » Wed Jan 03, 2024 5:07 pm

Knowing brand and model of machine might be helpful to any responders.
The people who confuse "entomology" and "etymology" really bug me beyond words.
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p

User avatar
ChicagoGranny
Posts: 14499
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: BiPAP not picking up inspiration sometimes

Post by ChicagoGranny » Wed Jan 03, 2024 5:25 pm

unputdownnable wrote:
Wed Jan 03, 2024 3:11 pm
Hi! I'm migrating from CPAP to BiPAP and has struggling with the device not picking up my inspiration and not triggering the IPAP often. Could this be a breathing problem or a device issue? I initially noticed a clogged nostril but even after decongestion it kept happening.
Many possibilities. Please state machine and mask models, mode setting, pressure settings including ramp, ...

If you don't know how to get into the clinical menu to check the settings, find a youtube video for your model machine explaining it.

unputdownnable
Posts: 15
Joined: Wed Jan 03, 2024 3:04 pm

Re: BiPAP not picking up inspiration sometimes

Post by unputdownnable » Wed Jan 03, 2024 6:21 pm

Here is my setup:

Device: BPAP AUTO GII BMC T-20A Ser. (https://en.bmc-medical.com/products/npa ... t-25a.html)
Mode: S
Mask: Phillips Nasal DreamWear
EPAP: 6 cmH2O
IPAP: 10.5 cmH2O
Ramp: 10min
I Sense: 8
E Sense: 8
Hose length: ~1.75m (no bacterial filter)

P.S.: BTW, the hose is passing through a hanger on the wall so that it does not get too lengthy on the bed.
Last edited by unputdownnable on Wed Jan 03, 2024 8:22 pm, edited 6 times in total.

User avatar
ChicagoGranny
Posts: 14499
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: BiPAP not picking up inspiration sometimes

Post by ChicagoGranny » Wed Jan 03, 2024 6:36 pm

Is it the AirCurve 10V? The model name should be on the front of the machine. https://www.cpap.com/productpage/resmed ... el-machine

Can you post a typical day in imgur.com and share the link in this thread? Instructions in my signature.
First read: viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

Then get a free account: https://home.sleephq.com/

Post links in this thread for members to see and make recommendations.

User avatar
lazarus
Posts: 1305
Joined: Sat Mar 12, 2022 3:21 pm
Location: Just above NJ.

Re: BiPAP not picking up inspiration sometimes

Post by lazarus » Wed Jan 03, 2024 6:55 pm

As a general rule, ResMed bilevels have a setting in the clinician menu for adjusting trigger sensitivity to help it recognize when you breathe in. There should also be a sensitivity adjustment for cycle, when you breathe out.
"Under normal conditions, the device triggers (initiates IPAP) and cycles (terminates IPAP and changes to EPAP) as it senses the change in patient flow. . . . The device has adjustable trigger/cycle sensitivity to optimize the sensing level according to patient conditions."--https://www.resmed.com/en-us/healthcare ... echnology/
I believe most users do well with "medium" for both, although a few users are more comfortable with changing those for increased or decreased sensitivity.
The people who confuse "entomology" and "etymology" really bug me beyond words.
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p

unputdownnable
Posts: 15
Joined: Wed Jan 03, 2024 3:04 pm

Re: BiPAP not picking up inspiration sometimes

Post by unputdownnable » Wed Jan 03, 2024 7:42 pm

My device has sensitivities ranging from 1 to 8. I have played with both "I Sensitivity" and "E Sensitivity" for many days. Both are at maximum lately (=8). But still, every once in a while when I finish breathing out and start to breathe in, it does not switch the pressure and it gets hard to breathe in.

User avatar
SleepGeek
Posts: 1049
Joined: Mon Feb 09, 2015 2:56 pm

Re: BiPAP not picking up inspiration sometimes

Post by SleepGeek » Wed Jan 03, 2024 7:44 pm

unputdownnable wrote:
Wed Jan 03, 2024 3:11 pm
Hi! I'm migrating from CPAP to BiPAP and has struggling with the device not picking up my inspiration and not triggering the IPAP often. Could this be a breathing problem or a device issue? I initially noticed a clogged nostril but even after decongestion it kept happening.
How long is the hose?
Do you have anything in the hose line like a bacterial filter?
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: What exactly do you think you're adding to this thread?

unputdownnable
Posts: 15
Joined: Wed Jan 03, 2024 3:04 pm

Re: BiPAP not picking up inspiration sometimes

Post by unputdownnable » Wed Jan 03, 2024 7:50 pm

ChicagoGranny wrote:
Wed Jan 03, 2024 6:36 pm
Is it the AirCurve 10V? The model name should be on the front of the machine. https://www.cpap.com/productpage/resmed ... el-machine

Can you post a typical day in imgur.com and share the link in this thread? Instructions in my signature.
It's actually "BPAP AUTO GII BMC T-20A Ser.". I will have a look at the instructions.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: BiPAP not picking up inspiration sometimes

Post by palerider » Wed Jan 03, 2024 7:57 pm

unputdownnable wrote:
Wed Jan 03, 2024 6:21 pm
Here is my setup:

Device: Resmed T20A
I'm not aware of any such model. what's it say on the front of the machine?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

unputdownnable
Posts: 15
Joined: Wed Jan 03, 2024 3:04 pm

Re: BiPAP not picking up inspiration sometimes

Post by unputdownnable » Wed Jan 03, 2024 8:01 pm

SleepGeek wrote:
Wed Jan 03, 2024 7:44 pm
unputdownnable wrote:
Wed Jan 03, 2024 3:11 pm
Hi! I'm migrating from CPAP to BiPAP and has struggling with the device not picking up my inspiration and not triggering the IPAP often. Could this be a breathing problem or a device issue? I initially noticed a clogged nostril but even after decongestion it kept happening.
How long is the hose?
Do you have anything in the hose line like a bacterial filter?
Hose is about 1.7-1.8m. No bacterial filter.

unputdownnable
Posts: 15
Joined: Wed Jan 03, 2024 3:04 pm

Re: BiPAP not picking up inspiration sometimes

Post by unputdownnable » Wed Jan 03, 2024 8:07 pm

palerider wrote:
Wed Jan 03, 2024 7:57 pm
unputdownnable wrote:
Wed Jan 03, 2024 6:21 pm
Here is my setup:

Device: Resmed T20A
I'm not aware of any such model. what's it say on the front of the machine?
I updated my post with more details and the link to the device model.

User avatar
ozij
Posts: 10169
Joined: Fri Mar 18, 2005 11:52 pm

Re: BiPAP not picking up inspiration sometimes

Post by ozij » Wed Jan 03, 2024 10:27 pm

RESmart GII BPAP System (T-20A / T-25A)

https://en.bmc-medical.com/contact/

International Headquarters
BMC Medical Co., Ltd.
Add: Room 110 Tower A Fengyu Building, No.115 Fucheng Road, Haidian, 100036 Beijing, People's Republic of China

Phone: +86 18622933511
Fax: 86-10-5166 3880 ext. 810
Email: intl@bmc-medical.com
For Technical service: service1@bmc-medical.com

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
ChicagoGranny
Posts: 14499
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: BiPAP not picking up inspiration sometimes

Post by ChicagoGranny » Thu Jan 04, 2024 12:28 pm

unputdownnable wrote:
Wed Jan 03, 2024 7:50 pm
It's actually "BPAP AUTO GII BMC T-20A Ser.".
BMC's ( :( ) machines are not supported by OSCAR.

unputdownnable
Posts: 15
Joined: Wed Jan 03, 2024 3:04 pm

Re: BiPAP not picking up inspiration sometimes

Post by unputdownnable » Fri Jan 05, 2024 2:06 pm

unputdownnable wrote:
Wed Jan 03, 2024 3:11 pm
Hi! I'm migrating from CPAP to BiPAP and have been struggling with the device not picking up my inspiration and not triggering the IPAP often. Could this be a breathing problem or a device issue? I initially noticed a clogged nostril but even after decongestion it kept happening.
Just wanted to share that I've noticed that this was happening during ramp only. I've made 2 changes that seem to not trigger the issue anymore for the last 2 nights:
1) Increased the ramp time from 10min to 15min.
2) Reduced the "E Sensibility" from 8 (max) to 7.

If the issue comes back I will update in the thread.