OSCAR: "CPAP Data Located" NOT
-
- Posts: 6
- Joined: Sat Jun 20, 2020 2:25 pm
OSCAR: "CPAP Data Located" NOT
New CPAP user since June 16, 2020.
I've downloaded OSCAR (onto a Chromebook), gave LINUX permission to the locked SD card, filled out the profile, etc. The data, however, is not being recognized by OSCAR - the process at http://www.apneaboard.com/wiki/index.ph ... FIRST_time stops, and I never see the following "CPAP Data Located" message.
The http://www.apneaboard.com/wiki/index.ph ... leshooting troubleshooting page has a blank "Data won't load correctly or at all" section http://www.apneaboard.com/wiki/index.ph ... _at_all.29.
Thoughts, suggestions, advice?
Thank you! Karen
I've downloaded OSCAR (onto a Chromebook), gave LINUX permission to the locked SD card, filled out the profile, etc. The data, however, is not being recognized by OSCAR - the process at http://www.apneaboard.com/wiki/index.ph ... FIRST_time stops, and I never see the following "CPAP Data Located" message.
The http://www.apneaboard.com/wiki/index.ph ... leshooting troubleshooting page has a blank "Data won't load correctly or at all" section http://www.apneaboard.com/wiki/index.ph ... _at_all.29.
Thoughts, suggestions, advice?
Thank you! Karen
"Karen"
Re: OSCAR: "CPAP Data Located" NOT
Did you click the Import button?
In Linux I don't believe it would show as a drive letter. It'd probably be something like /mnt/SD. though I believe the name is up to you. I am also not sure Linux auto-mounts SD cards as Windows does ... you may have to manually mount it with the mount command.
Once you're sure it is mounted, you should be able to click import and navigate to where the root of the SD card is mounted.
-Kevin
In Linux I don't believe it would show as a drive letter. It'd probably be something like /mnt/SD. though I believe the name is up to you. I am also not sure Linux auto-mounts SD cards as Windows does ... you may have to manually mount it with the mount command.
Once you're sure it is mounted, you should be able to click import and navigate to where the root of the SD card is mounted.
-Kevin
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.
Re: OSCAR: "CPAP Data Located" NOT
welcome to the zoo!
as far as i can tell, you may be the first person on the forum to try to run oscar on a chromebook, though maybe not the first person to run it on linux.
you may get better results posting in this thread-
viewtopic/t179460/OSCAR-v111-is-Released.html
the authors are on here regularly and may "see" your post there first.
good luck!!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
- Posts: 6
- Joined: Sat Jun 20, 2020 2:25 pm
Re: OSCAR: "CPAP Data Located" NOT
I was hoping you'd chime in, Zonker. Thank you, Arlenzonker wrote:welcome to the zoo!
as far as i can tell, you may be the first person on the forum to try to run oscar on a chromebook, though maybe not the first person to run it on linux.
you may get better results posting in this thread-
viewtopic/t179460/OSCAR-v111-is-Released.html
the authors are on here regularly and may "see" your post there first.
good luck!!
Sent from my MHA-L29 using Tapatalk
"Karen"
- BlueDragon
- Posts: 541
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: OSCAR: "CPAP Data Located" NOT
Which version of OSCAR did you download? Chromebooks are not yet on our list of supported computers although support for them is high on the wish list. I'll direct one of the developers familiar with linux and chromebook to your question.Karen Avery wrote: ↑Thu Jun 25, 2020 9:26 amI've downloaded OSCAR (onto a Chromebook), gave LINUX permission to the locked SD card, filled out the profile, etc. The data, however, is not being recognized by OSCAR ....
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
OSCAR 1.5.1 is the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
Re: OSCAR:
or just wait right here and blue dragon will answer!Karen Avery wrote: ↑Thu Jun 25, 2020 12:14 pmI was hoping you'd chime in, Zonker. Thank you, Arlen
Sent from my MHA-L29 using Tapatalk
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- BlueDragon
- Posts: 541
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: OSCAR: "CPAP Data Located" NOT
I've sent you a PM (private message) about helping test a Chromebook version of OSCAR.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
OSCAR 1.5.1 is the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
Re: OSCAR: "CPAP Data Located" NOT
kewel! if you can get a version going for chromebook(in the far future, natch) i'd be interested.BlueDragon wrote: ↑Thu Jun 25, 2020 6:30 pmI've sent you a PM (private message) about helping test a Chromebook version of OSCAR.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: OSCAR: "CPAP Data Located" NOT
Karen,
one of the projects I'm working on is cleaning up the Wiki for installing OSCAR on Linux, on a Chromebook. We know it runs on Chromebook, no question, but we need solid instructions on how-to that can be followed by novices. Linux Pros don't seem too major problems. (I'm sure there are small things they can easily solve). This needs to be solid before we announce it.
one of the projects I'm working on is cleaning up the Wiki for installing OSCAR on Linux, on a Chromebook. We know it runs on Chromebook, no question, but we need solid instructions on how-to that can be followed by novices. Linux Pros don't seem too major problems. (I'm sure there are small things they can easily solve). This needs to be solid before we announce it.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: VAuto 11-25cmH2O PS 3 --- OSCAR software, Many masks - Amara View FFM to P10 Pillows Several Nasal and FFM |
-
- Posts: 6
- Joined: Sat Jun 20, 2020 2:25 pm
Re: OSCAR: "CPAP Data Located" NOT
I would love to help you betatest anything you're working on. BlueDragon PM'ed me about the Apnea Board Forum discussion about OSCAR and Chromebooks, too, so I've signedup there as well.
"Karen"
"Karen"
bonjour wrote: ↑Thu Jun 25, 2020 7:23 pmKaren,
one of the projects I'm working on is cleaning up the Wiki for installing OSCAR on Linux, on a Chromebook. We know it runs on Chromebook, no question, but we need solid instructions on how-to that can be followed by novices. Linux Pros don't seem too major problems. (I'm sure there are small things they can easily solve). This needs to be solid before we announce it.
"Karen"
Re: OSCAR: "CPAP Data Located" NOT
plus, Linux doesn't use backslasheskhauser wrote: ↑Thu Jun 25, 2020 9:41 amDid you click the Import button?
In Linux I don't believe it would show as a drive letter. It'd probably be something like /mnt/SD. though I believe the name is up to you. I am also not sure Linux auto-mounts SD cards as Windows does ... you may have to manually mount it with the mount command.
Once you're sure it is mounted, you should be able to click import and navigate to where the root of the SD card is mounted.
-Kevin
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: OSCAR: "CPAP Data Located" NOT
It's good that you gave Linux permission to access the SD Card and indeed managed to install on a Chromebook. Have you seen this wiki page before?Karen Avery wrote: ↑Thu Jun 25, 2020 9:26 amNew CPAP user since June 16, 2020.
I've downloaded OSCAR (onto a Chromebook), gave LINUX permission to the locked SD card, filled out the profile, etc. The data, however, is not being recognized by OSCAR - the process at http://www.apneaboard.com/wiki/index.ph ... FIRST_time stops, and I never see the following "CPAP Data Located" message. CPAP_data_located.png
The http://www.apneaboard.com/wiki/index.ph ... leshooting troubleshooting page has a blank "Data won't load correctly or at all" section http://www.apneaboard.com/wiki/index.ph ... _at_all.29.
Thoughts, suggestions, advice?
Thank you! Karen
http://www.apneaboard.com/wiki/index.ph ... _Chrome_OS
It contains a header that reads There are 5 topics we need to deal with here:
and you seem to have managed the first 4 steps, but stumbled at the 5th.
That section has the header
Importing SD Card Data
OSCAR is not currently capable of finding your SD Card for you, you have to tell it where it is.
Try working your way through that section, and it should all work. Post back here if you get stuck.
Michael