Sudden burst of CA's because left arm was out of bed!
Sudden burst of CA's because left arm was out of bed!
I wonder if anyone can explain this brief weird set of occurrences.
2 nights ago I was woken at 6:04am by an arousal, and felt the need to switch from my right side to my left side. I confirmed this for my records with 5 quick sniffs followed by 4 more (easily seen on a Flow Rate chart).
Just 70 seconds later I was awoken (not from real sleep, I suspect) by an oximeter vibration because my SpO2 had dropped to 89%. I was on my left side with my left wrist (the one with the oximeter on) hanging slightly off the bed and my right arm was across it. I lifted my right arm off it and then lifted my left wrist in the air, and the vibration stopped and SpO2 immediately jumped to 98%!
I thought this odd so, after about 3 minutes, I deliberately put my left wrist back hanging off the bed. That didn't seem to cause a vibration (though it looks like O2 did drop a lot). After mulling that over for another few minutes, I put my left wrist back hanging off the bed and laid my right arm across it. In less than a minute, the oximeter vibrated again!
I repeated this again and got the same result - and decided it was time to get up.
When I got all the data from the CPAP and the oximeter and SomnoPose into Oscar (see attached chart showing the most relevant clips), I realised the situation was even more complex than I thought. During this 20 minute episode, the ResMed 10 said I had NINE CA's (something I don't often get actually). I suspect none of these were real, especially as I was mostly awake during this time.
So is the oximeter really telling the truth about my body being short of oxygen, or is it just responding erroneously to a slightly cut-off blood supply to that hand?
Here is an expanded view of the first CA.
2 nights ago I was woken at 6:04am by an arousal, and felt the need to switch from my right side to my left side. I confirmed this for my records with 5 quick sniffs followed by 4 more (easily seen on a Flow Rate chart).
Just 70 seconds later I was awoken (not from real sleep, I suspect) by an oximeter vibration because my SpO2 had dropped to 89%. I was on my left side with my left wrist (the one with the oximeter on) hanging slightly off the bed and my right arm was across it. I lifted my right arm off it and then lifted my left wrist in the air, and the vibration stopped and SpO2 immediately jumped to 98%!
I thought this odd so, after about 3 minutes, I deliberately put my left wrist back hanging off the bed. That didn't seem to cause a vibration (though it looks like O2 did drop a lot). After mulling that over for another few minutes, I put my left wrist back hanging off the bed and laid my right arm across it. In less than a minute, the oximeter vibrated again!
I repeated this again and got the same result - and decided it was time to get up.
When I got all the data from the CPAP and the oximeter and SomnoPose into Oscar (see attached chart showing the most relevant clips), I realised the situation was even more complex than I thought. During this 20 minute episode, the ResMed 10 said I had NINE CA's (something I don't often get actually). I suspect none of these were real, especially as I was mostly awake during this time.
So is the oximeter really telling the truth about my body being short of oxygen, or is it just responding erroneously to a slightly cut-off blood supply to that hand?
Here is an expanded view of the first CA.
_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + P10 Nasal Pillow Mask Bundle |
Additional Comments: ResMed F20 also; Oscar; SleepHQ; Wellue SleepU oximeter; iphone sleep stages |
- ChicagoGranny
- Posts: 14556
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Sudden burst of CA's because left arm was out of bed!
Why do people worry about such one-offs as this?
Re: Sudden burst of CA's because left arm was out of bed!
CAs while you are awake are meaningless.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
Re: Sudden burst of CA's because left arm was out of bed!
Yup, why didn't Dr Alexander Fleming simply ignore the mould growing on the Petri dish? It was only a one-off observation.
_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + P10 Nasal Pillow Mask Bundle |
Additional Comments: ResMed F20 also; Oscar; SleepHQ; Wellue SleepU oximeter; iphone sleep stages |
Re: Sudden burst of CA's because left arm was out of bed!
CAs while you are awake are meaningless.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
- ChicagoGranny
- Posts: 14556
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Sudden burst of CA's because left arm was out of bed!
It certainly wasn't a one-off. The mold and its curative properties had been observed for centuries. Fleming was the first to identify which substance in the mold had the bacteria-fighting characteristics. It was a decade later that a team of scientists that did not involve Fleming developed methods for cultivating this substance and making it available for use.
Please update this thread when your arm-hanging-off-mattress research results in a major step forward in medical history.
(This comment is meant in a lighthearted way. )
Re: Sudden burst of CA's because left arm was out of bed!
As was mine....to help me resist the urge to ask if it was normal on this forum for long-term residents to shoot newbies down in flames by criticising one of their first pleas for help!(This comment is meant in a lighthearted way. )
_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + P10 Nasal Pillow Mask Bundle |
Additional Comments: ResMed F20 also; Oscar; SleepHQ; Wellue SleepU oximeter; iphone sleep stages |
- vandownbytheriver
- Posts: 395
- Joined: Sat Feb 03, 2024 11:42 pm
Re: Sudden burst of CA's because left arm was out of bed!
Yes, I've had my O2Ring go off from having the limb go to sleep.
Yes, I've had the machine register CA's from holding my breath while moving around. I've had to train myself to breathe during my toss-n-turns if it's taking a while. The Movement graph is my hillbilly EEG... see Movement? It was an arousal period.
If you're on the edge of CA's anyway then an exertion can cause a cycle of them. Say you're running high EPR (we don't know your pressures) and medium pressures... if you have an exertion (rolling over or re-positioning yourself) you'll breathe more... then less... then more... etc. EPR 3 tends to make this oscillation go longer. This can happen when you're awake... your body follows its blood pH whether you're asleep or not. Curious to know your pressures.
Yes, I've had the machine register CA's from holding my breath while moving around. I've had to train myself to breathe during my toss-n-turns if it's taking a while. The Movement graph is my hillbilly EEG... see Movement? It was an arousal period.
If you're on the edge of CA's anyway then an exertion can cause a cycle of them. Say you're running high EPR (we don't know your pressures) and medium pressures... if you have an exertion (rolling over or re-positioning yourself) you'll breathe more... then less... then more... etc. EPR 3 tends to make this oscillation go longer. This can happen when you're awake... your body follows its blood pH whether you're asleep or not. Curious to know your pressures.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
- ChicagoGranny
- Posts: 14556
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
- chunkyfrog
- Posts: 34451
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Sudden burst of CA's because left arm was out of bed!
Switch the oximeter to a finger on the other hand.
Or put the thing in a drawer, where mine is.
Or put the thing in a drawer, where mine is.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Sudden burst of CA's because left arm was out of bed!
I wondered at first if it could have been that simple, but the 'CA' issue (and little change in Pulse Rate) surely suggests it wasn't a localised problem?vandownbytheriver wrote: ↑Tue Apr 30, 2024 3:26 amYes, I've had my O2Ring go off from having the limb go to sleep.
I too have tried to learn to keep breathing when moving around, though I know I sometimes fail.Yes, I've had the machine register CA's from holding my breath while moving around. I've had to train myself to breathe during my toss-n-turns if it's taking a while.
I'm not sure which of the many movements you are referring to, but I did say it all started with an arousal.The Movement graph is my hillbilly EEG... see Movement? It was an arousal period.
I wasn't aware of this oscillation issue. Here's a standard Oscar screenshot for that whole night. Sadly, EPR gets cutoff, but it was 3.If you're on the edge of CA's anyway then an exertion can cause a cycle of them. Say you're running high EPR (we don't know your pressures) and medium pressures... if you have an exertion (rolling over or re-positioning yourself) you'll breathe more... then less... then more... etc. EPR 3 tends to make this oscillation go longer. This can happen when you're awake... your body follows its blood pH whether you're asleep or not. Curious to know your pressures.
_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + P10 Nasal Pillow Mask Bundle |
Additional Comments: ResMed F20 also; Oscar; SleepHQ; Wellue SleepU oximeter; iphone sleep stages |
- vandownbytheriver
- Posts: 395
- Joined: Sat Feb 03, 2024 11:42 pm
Re: Sudden burst of CA's because left arm was out of bed!
Strictly referring to your zoom in.
That was my guess... low pressure, max EPR. Exertion, resulting in some extra breathing, resulting in a cycle of CA's.I wasn't aware of this oscillation issue. Here's a standard Oscar screenshot for that whole night. Sadly, EPR gets cutoff, but it was 3.
SleepHQ is free, and allows a hi-res sharing of data. I use Oscar for at home and SHQ for link sharing. Here's a night from December, before I went bi-level:
https://sleephq.com/public/1f8e789b-83f ... 2c35847870
Note you can click and drag and zoom in like Oscar. 'r' resets the graphs.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
Re: Sudden burst of CA's because left arm was out of bed!
Sadly, the free version doesn't let me include my oximeter data or my SomnoPose data....and, for what you get, the Pro version's $15/month seems a lot to a pensioner.SleepHQ is free, and allows a hi-res sharing of data.
_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + P10 Nasal Pillow Mask Bundle |
Additional Comments: ResMed F20 also; Oscar; SleepHQ; Wellue SleepU oximeter; iphone sleep stages |