Connecting O2 to CPAP

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Connecting O2 to CPAP

Post by wardmiller » Sat Mar 17, 2018 1:39 pm

I use O2 all during the day. Now I notice my saturation during the night is slipping down and when i tell my doc he probably will want me to connect up at night, too. I have been using a ResMed Quattro FX ffm for several years and it works just fine for me.

How do I connect the O2 tube to the mask, or CPAP? Do I have to get a different mask? I hate to again go through getting one that fits.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

User avatar
Pugsy
Posts: 64145
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Connecting O2 to CPAP

Post by Pugsy » Sat Mar 17, 2018 1:49 pm

What kind of hose are you using on your S9?
Heated or non heated?

If non heated you can use something like this.
https://www.cpap.com/productpage/oxygen ... apter.html
at the long hose to humidifier junction and save on having to snake the O2 hose along side your cpap hose.

Or if using a heated hose you can of course stick it at the mask/long hose junction.

Or get the heated hose that has a built in little port for hooking up the O2 line.
https://www.cpap.com/productpage/resmed ... rt-S9.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: Connecting O2 to CPAP

Post by wardmiller » Sat Mar 17, 2018 2:09 pm

I'm using a non-heated hose. That adapter you first cited looks like the solution. Thanks.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: Connecting O2 to CPAP

Post by wardmiller » Sun Mar 18, 2018 7:36 am

Pugsy, I was just about to order that T-shaped hose connection, then I had a thought. The CPAP hose puts out quite an airflow. You can feel it several feet away. The O2 machine puts out such little pressure, at 2 l/min I can't feel it when I put the tubing to my cheek. I check it by putting the tubing end in a glass of water and look for bubbles. It would appear to me the CPAP would prevent the O2 flow from exiting the tubing and entering the hose. Stated another way, if there were a flapper door in the connector where the tubing and the hose join, the CPAP air would push the door toward the tubing.

I know this is a lot like the small river flowing into the big river, and it does flow, but we are talking about much, much stronger forces there. With my O2 concentrator, it is continuous flow and there is no way I can tell if the O2 is actually leaving the tubing.

What do you think about it?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

User avatar
Pugsy
Posts: 64145
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Connecting O2 to CPAP

Post by Pugsy » Sun Mar 18, 2018 7:47 am

People use those adapters all the time and they seem to work just fine.
We have a lot of people here on the forum using them.

Also when you first turn the machine on it does seem like there is a massive air stream to fight but that's really only until the seal is made and once the seal is made the forceful movement of the air isn't nearly so significant.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: Connecting O2 to CPAP

Post by greatunclebill » Sun Mar 18, 2018 8:21 am

wardmiller wrote:
Sun Mar 18, 2018 7:36 am
Pugsy, I was just about to order that T-shaped hose connection, then I had a thought. The CPAP hose puts out quite an airflow. You can feel it several feet away. The O2 machine puts out such little pressure, at 2 l/min I can't feel it when I put the tubing to my cheek. I check it by putting the tubing end in a glass of water and look for bubbles. It would appear to me the CPAP would prevent the O2 flow from exiting the tubing and entering the hose. Stated another way, if there were a flapper door in the connector where the tubing and the hose join, the CPAP air would push the door toward the tubing.

I know this is a lot like the small river flowing into the big river, and it does flow, but we are talking about much, much stronger forces there. With my O2 concentrator, it is continuous flow and there is no way I can tell if the O2 is actually leaving the tubing.

What do you think about it?
this is often asked by people new to the o2 game. it's apples and oranges. cpap is very low pressure and high volume. think of standing in front of a fan. o2 is higher pressure and low volume. the high cpap pressure that people complain about is only .3psi. the o2 pressure whether out of a cylinder or concentrater is many times higher than the cpap pressure. the 2L per hour flow from the concentrator is volume, not pressure. bottom line: plug it in, turn it on and go to sleep. it blends together just fine.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: Connecting O2 to CPAP

Post by wardmiller » Sun Mar 18, 2018 9:53 am

Pugsy and Bill, thank you for those replies. I'll order one today.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Connecting O2 to CPAP

Post by Goofproof » Sun Mar 18, 2018 12:44 pm

wardmiller wrote:
Sun Mar 18, 2018 7:36 am
The O2 machine puts out such little pressure, at 2 l/min I can't feel it when I put the tubing to my cheek.

I know this is a lot like the small river flowing into the big river, and it does flow, but we are talking about much, much stronger forces there. With my O2 concentrator, it is continuous flow and there is no way I can tell if the O2 is actually leaving the tubing.

What do you think about it?
Totally Flawed thinking!

1. The O2 Concentrator pressure and flow volume are two different animals. Flow is the amount of something, pressure is the push behind it. The O2 Concentrator, doesn't build it pressure until it's flow is restricted, the pressure is measured in .cm of water level, the Flow is controlled by a regulator gauge. they are built into the O2 unit or mounted on the tank. If yours doesn't have one, get one.

About the Small River flowing into the Big River, What makes a Big River, Water flowing into it! How does water flow, down hill, from higher to lower. XPAPS are supposed to be used with Full Flow O2 Only!

The O2 Concentrator pressure IS higher than your XPAP pressure, that's why the O2 goes into the XPAP airstream. The reason the Small river or Stream goes into the Big River is the same the Small river or Stream's level is higher than the Big River, It has more pressure.

Put your finger on the O2 hose and seal it off when it's running you will feel the pressure building up, the pressure exceeds the XPAP pressure by a lot, the flow does not, the XPAP flow is much higher.

I use 3 L of O2, because I realize much of the added O2 is washed out of the mask by the mask vent, it's common sense, but that's not taught in med school, it's learned by life, or not! Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: Connecting O2 to CPAP

Post by wardmiller » Sun Mar 18, 2018 2:35 pm

Goofproof wrote:
Sun Mar 18, 2018 12:44 pm

Totally Flawed thinking!
...
I use 3 L of O2, because I realize much of the added O2 is washed out of the mask by the mask vent, it's common sense, but that's not taught in med school, it's learned by life, or not! Jim
Thanks for the education.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Connecting O2 to CPAP

Post by Goofproof » Sun Mar 18, 2018 3:22 pm

wardmiller wrote:
Sun Mar 18, 2018 2:35 pm
Goofproof wrote:
Sun Mar 18, 2018 12:44 pm

Totally Flawed thinking!
...
I use 3 L of O2, because I realize much of the added O2 is washed out of the mask by the mask vent, it's common sense, but that's not taught in med school, it's learned by life, or not! Jim
Thanks for the education.
I start the O2 unit, start the XPAP at my full pressure, put the mask on, then adjust my O2 flow on the O2 unit. that way i'm sure the O2 is correct, As far as the O2 rate 2 L is fine, I just want more. Jim

Also there are more than one kind of bleed in adapter, mine is safer than most it has a check valve built in. When you turn the XPAP off, the check valve stops th O2 flow from going back into the XPAP by dumping the added O2 into the room, downside the Smartstart won't work with it, no problem, (I can't find one like it, just cheap ones). I always turn my own on. I also keep the O2 unit in the next room (50 feet line max), keeping the heat and noise out of the bedroom.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Connecting O2 to CPAP

Post by palerider » Sun Mar 18, 2018 4:32 pm

wardmiller wrote:
Sun Mar 18, 2018 7:36 am
Pugsy, I was just about to order that T-shaped hose connection, then I had a thought. The CPAP hose puts out quite an airflow. You can feel it several feet away. The O2 machine puts out such little pressure, at 2 l/min I can't feel it when I put the tubing to my cheek. I check it by putting the tubing end in a glass of water and look for bubbles. It would appear to me the CPAP would prevent the O2 flow from exiting the tubing and entering the hose. Stated another way, if there were a flapper door in the connector where the tubing and the hose join, the CPAP air would push the door toward the tubing.

I know this is a lot like the small river flowing into the big river, and it does flow, but we are talking about much, much stronger forces there. With my O2 concentrator, it is continuous flow and there is no way I can tell if the O2 is actually leaving the tubing.

What do you think about it?
The oxygen concentrator has *much* higher pressure than the cpap.

How much air you can feel blowing out the cpap hose when it's not connected is totally irrelevant, because it's running at full speed, trying to pressurize your entire room.

If those O2 adapters didn't work (they do) they wouldn't sell them, and Pugsy wouldn't have recommended them.

(my housemate uses one on his cpap.)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Connecting O2 to CPAP

Post by palerider » Sun Mar 18, 2018 4:36 pm

palerider wrote:
Sun Mar 18, 2018 4:32 pm
wardmiller wrote:
Sun Mar 18, 2018 7:36 am
Pugsy, I was just about to order that T-shaped hose connection, then I had a thought. The CPAP hose puts out quite an airflow. You can feel it several feet away. The O2 machine puts out such little pressure, at 2 l/min I can't feel it when I put the tubing to my cheek. I check it by putting the tubing end in a glass of water and look for bubbles. It would appear to me the CPAP would prevent the O2 flow from exiting the tubing and entering the hose. Stated another way, if there were a flapper door in the connector where the tubing and the hose join, the CPAP air would push the door toward the tubing.

I know this is a lot like the small river flowing into the big river, and it does flow, but we are talking about much, much stronger forces there. With my O2 concentrator, it is continuous flow and there is no way I can tell if the O2 is actually leaving the tubing.

What do you think about it?
The oxygen concentrator has *much* higher pressure than the cpap.

How much air you can feel blowing out the cpap hose when it's not connected is totally irrelevant, because it's running at full speed, trying to pressurize your entire room.

If those O2 adapters didn't work (they do) they wouldn't sell them, and Pugsy wouldn't have recommended them.

(my housemate uses one on his cpap.)

As to how you can tell whether it's flowing or not, you look at the flowmeter on the concentrator.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: Connecting O2 to CPAP

Post by wardmiller » Mon Mar 19, 2018 7:26 am

greatunclebill wrote:
Sun Mar 18, 2018 8:21 am

. . . the 2L per hour flow from the concentrator. . .
I believe you meant "per minute", not "per hour". Right?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.

User avatar
ChicagoGranny
Posts: 14516
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Connecting O2 to CPAP

Post by ChicagoGranny » Mon Mar 19, 2018 7:43 am

wardmiller wrote:
Sun Mar 18, 2018 7:36 am
The CPAP hose puts out quite an airflow. You can feel it several feet away. The O2 machine puts out such little pressure, at 2 l/min I can't feel it when I put the tubing to my cheek. I check it by putting the tubing end in a glass of water and look for bubbles. It would appear to me the CPAP would prevent the O2 flow from exiting the tubing and entering the hose.

I know this is a lot like the small river flowing into the big river, and it does flow, but we are talking about much, much stronger forces there. With my O2 concentrator, it is continuous flow and there is no way I can tell if the O2 is actually leaving the tubing.
Actually, the oxygen concentrator creates a pressure many times higher than any CPAP pressure. The concentrator has a pressure reducing valve to allow the correct amount of oxygen to flow through the hose that is connected to the patient.

Concentrator/CPAP systems are a proven science with a history of thousands of patients.

wardmiller
Posts: 294
Joined: Mon Sep 17, 2012 6:29 pm
Location: Upstate New York

Re: Connecting O2 to CPAP

Post by wardmiller » Mon Mar 19, 2018 7:52 am

palerider wrote:
Sun Mar 18, 2018 4:32 pm

The oxygen concentrator has *much* higher pressure than the cpap.

How much air you can feel blowing out the cpap hose when it's not connected is totally irrelevant, because it's running at full speed, trying to pressurize your entire room.
Just for the record, after I have worn the CPAP mask for many minutes, I can feel the airflow out of the mask's vent 16" away or more. When set to 2 L/min, I can not feel oxygen coming out of the concentrator's tubing, but I know it is working if I put the tubing in a glass of water and see bubbles.
palerider wrote:
Sun Mar 18, 2018 4:32 pm
If those O2 adapters didn't work (they do) they wouldn't sell them. . .
I don't know where you lived, but around here we can buy plenty of stuff that does not work as advertised. Some is even mentioned on this forum from time to time.

Also, please note the advertisement say, "Return On Warranty Issues Only". Does not provide for return if it fails to work as advertised.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed VPAP Auto, Series 9 device, with H5i Heated Humidifier; RedMed Mirage Quattro full-face mask.