Is EPR best avoided?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Hannibal 2
Posts: 485
Joined: Sun Nov 20, 2016 4:46 am
Location: Derbyshire UK

Is EPR best avoided?

Post by Hannibal 2 » Fri Mar 24, 2017 4:35 am

My machine was set to the maximum EPR setting of 3cmH20, I know this is a comfort feature but does it compromise my therapy? I'm trying to edge up my minimum pressure gradually closer to my 95 per cent pressure, within 2cmH20. Am I better off without EPR if I find I can manage without? Maybe drop it gradually and then switch it off?

_________________
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Air 10 ClimateLineAir, SleepyHead Software
"Welcome my son, welcome to the machine!
Where have you been? It's alright we know where you've been!"
(You've been in the pipeline...)

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Is EPR best avoided?

Post by Jay Aitchsee » Fri Mar 24, 2017 5:14 am

EPR is a "try it and see" feature. Using EPR will effectively lower your EPAP and, therefore, your average therapy pressure. If you like it, you may have to raise your Min pressure setting slightly to compensate in preventing obstructive events. Some find Central Apneas increase with EPR, others do not. EPR in and of itself does not "compromise" your therapy and could enhance your therapy if it is more comfortable for you. Try it and see, or in this case, turn it off or lower it and see.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

rick blaine
Posts: 614
Joined: Tue Dec 15, 2015 7:30 am

Re: Is EPR best avoided?

Post by rick blaine » Fri Mar 24, 2017 6:12 am

It is, as Jay sez, a case of suck it and see. Some people find EPR not just more comfortable but getting on for absolutely necessary - to the point where, if 3 cm of relief isn't enough, they might could switch to a bi-level machine (with an up to 6 cm expiration pressure drop). Other people find the effect isn't that much different.

And a few - such as myself, have turned A-flex (the equivalent on my machine to the EPR on yours*) right off. It seems that, with older age, my sleeping resp rate can go above 20, and when it does, the algorithm doesn't quite mesh.

You might - since you're still in the learning/ exploring phase with your new machine - try it with the EPR off. Given normal lung function, breathing out against, say, a pressure of 15 cm is (a) not impossible, and (b) something your chest and diaphragm muscles soon adapt to. So give it try. Find out not the 'text book' answer but the Hannibal 2 answer.

* And, yes, for all you obsessive techno mavens, I do know it is not the exact equivalent.

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: Is EPR best avoided?

Post by DreamStalker » Fri Mar 24, 2017 6:21 am

It was for me. It doesn't take long to get used to breathing under pressure and after that you realize you really don't need exhalation relief. I think I turned mine off after a few months but I could have turned it off after a week or two I think.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
Hannibal 2
Posts: 485
Joined: Sun Nov 20, 2016 4:46 am
Location: Derbyshire UK

Re: Is EPR best avoided?

Post by Hannibal 2 » Fri Mar 24, 2017 7:05 am

Thank you for your helpful responses.

Rick, Ironically I was re-reading some of the advice you offered when I was about to purchase my first APAP machine a few weeks ago. You suggested trying to set my minimum pressure 0.5 to 2cmH20 below my average 95 per cent. This is what I am attempting to try to optimize my therapy, I will also try then to lower my EPR gradually to see how I cope with that.

I'll also keep watch on how centrals are affected either way.

_________________
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Air 10 ClimateLineAir, SleepyHead Software
"Welcome my son, welcome to the machine!
Where have you been? It's alright we know where you've been!"
(You've been in the pipeline...)

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Is EPR best avoided?

Post by palerider » Fri Mar 24, 2017 8:51 am

rick blaine wrote:It is, as Jay sez, a case of suck it and see. Some people find EPR not just more comfortable but getting on for absolutely necessary - to the point where, if 3 cm of relief isn't enough, they might could switch to a bi-level machine (with an up to 6 cm expiration pressure drop). Other people find the effect isn't that much different. * And, yes, for all you obsessive techno mavens, I do know it is not the exact equivalent.
I don't know where you're getting your "information" but, heres a couple of facts.

bilevel machines can have much more than 6cm pressure support. you can set the resmed s machine to an inhalation pressure of 25 and exhalation of 3, giving you an expiration pressure drop of 22cm.

and as for pressure support and epr being different, which of the following is epr, which is pressure support?

Image

the third line doesn't count, that's an experimentation of rise time.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Fri Mar 24, 2017 10:56 am, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Is EPR best avoided?

Post by palerider » Fri Mar 24, 2017 8:52 am

Hannibal 2 wrote:Thank you for your helpful responses.

Rick, Ironically I was re-reading some of the advice you offered when I was about to purchase my first APAP machine a few weeks ago. You suggested trying to set my minimum pressure 0.5 to 2cmH20 below my average 95 per cent. This is what I am attempting to try to optimize my therapy,
that is dumb advice.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Hannibal 2
Posts: 485
Joined: Sun Nov 20, 2016 4:46 am
Location: Derbyshire UK

Re: Is EPR best avoided?

Post by Hannibal 2 » Fri Mar 24, 2017 9:10 am

palerider wrote:
Hannibal 2 wrote:Thank you for your helpful responses.

Rick, Ironically I was re-reading some of the advice you offered when I was about to purchase my first APAP machine a few weeks ago. You suggested trying to set my minimum pressure 0.5 to 2cmH20 below my average 95 per cent. This is what I am attempting to try to optimize my therapy,
that is dumb advice.
Because? ....And what would your advice be?

Edit: should read 0.5 to 2 below the average pressure, sound better?

_________________
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Air 10 ClimateLineAir, SleepyHead Software
"Welcome my son, welcome to the machine!
Where have you been? It's alright we know where you've been!"
(You've been in the pipeline...)

User avatar
Pugsy
Posts: 64163
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Is EPR best avoided?

Post by Pugsy » Fri Mar 24, 2017 9:24 am

Hannibal 2 wrote:
Because? ....And what would your advice be?
Because you will be chasing your tail in terms of pressure and the more you increase the minimum the more the 95% pressure will increase too. There is nothing magical about that 95% number....it's just a number and a number that is easily skewed for any number of reasons. Remember the definition "at OR BELOW that pressure for 95% of the night".
The "or below" part is really important. It is NOT where you spent 95% of the night.

From your Mar 20 report...the 95% number is 14.74. Do you really want to use a 14 cm minimum? You could but it would be way overkill.

If you want to set a target for the minimum...look at the median average..it's a better target for a minimum.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Is EPR best avoided?

Post by palerider » Fri Mar 24, 2017 10:27 am

Hannibal 2 wrote:
palerider wrote:
Hannibal 2 wrote:Thank you for your helpful responses.

Rick, Ironically I was re-reading some of the advice you offered when I was about to purchase my first APAP machine a few weeks ago. You suggested trying to set my minimum pressure 0.5 to 2cmH20 below my average 95 per cent. This is what I am attempting to try to optimize my therapy,
that is dumb advice.
Because? ....And what would your advice be?
because it comes from someone that doesn't understand what the 95% number means, all the 95% number represents is a maximum pressure discarding any short term spikes.

like throwing out the highest and lowest judges score in the olympics.

it's a stupid, ignorant shortcut to try and get a minimum pressure without actually understanding anything anything about how the night progresses.
Hannibal 2 wrote:Edit: should read 0.5 to 2 below the average pressure, sound better?
no.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Oltremare
Posts: 187
Joined: Tue Dec 29, 2015 7:01 pm
Location: On the other side of the sea - Italy

Re: Is EPR best avoided?

Post by Oltremare » Fri Mar 24, 2017 10:49 am

Hello to all.
The EPR management (or equivalent) is a subjective thing.
But I don't understand why you have doubts on the use of the EPR?
There are some people who have complications for their CA when they use the EPR, but everyone else, why not use EPR? It 'a prejudice?
EPR for me is a great thing!
I could never do without it.
I tried to go from 3 to 2 but I did not succeed!
After two nights, I had to go to 3 but I have no CA.

Oltremare

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Quattro FX FF mask > CMS50D+ pulse oximeter > Climate line > Settings: auto-CPAP pressure Min 11.00 Max 18.0 EPR 3
I hope you will forgive my language errors. To write in English I use a translator online. Thank you

User avatar
Pugsy
Posts: 64163
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Is EPR best avoided?

Post by Pugsy » Fri Mar 24, 2017 11:23 am

Hannibal 2 wrote:Edit: should read 0.5 to 2 below the average pressure, sound better?
Are you aware that the "average" shown in SleepyHead defaults to the median average and not an overall normal average that I think you are thinking about.
You can change the median average tor reflect a "normal" average if you wish. Preferences/CPAP tab.

Have you see this?
http://adventures-in-hosehead-land.blog ... de-to.html

FWIW....IF (big IF) I was targeting anything for an APAP minimum pressure setting it would most likely be the median average.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Hannibal 2
Posts: 485
Joined: Sun Nov 20, 2016 4:46 am
Location: Derbyshire UK

Re: Is EPR best avoided?

Post by Hannibal 2 » Fri Mar 24, 2017 11:28 am

Pugsy wrote:
If you want to set a target for the minimum...look at the median average..it's a better target for a minimum.
Thanks for clarifying, and apologies to Rick for misquoting him - he said average pressure, not average 95 per cent pressure.

So I'll look at my median over the past week and take the average of that.

_________________
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Air 10 ClimateLineAir, SleepyHead Software
"Welcome my son, welcome to the machine!
Where have you been? It's alright we know where you've been!"
(You've been in the pipeline...)

User avatar
Hannibal 2
Posts: 485
Joined: Sun Nov 20, 2016 4:46 am
Location: Derbyshire UK

Re: Is EPR best avoided?

Post by Hannibal 2 » Fri Mar 24, 2017 11:30 am

Pugsy wrote:
Hannibal 2 wrote:Edit: should read 0.5 to 2 below the average pressure, sound better?
Are you aware that the "average" shown in SleepyHead defaults to the median average and not an overall normal average that I think you are thinking about.
You can change the median average tor reflect a "normal" average if you wish. Preferences/CPAP tab.

Have you see this?
http://adventures-in-hosehead-land.blog ... de-to.html

FWIW....IF (big IF) I was targeting anything for an APAP minimum pressure setting it would most likely be the median average.
Will take a look later, thanks

_________________
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Air 10 ClimateLineAir, SleepyHead Software
"Welcome my son, welcome to the machine!
Where have you been? It's alright we know where you've been!"
(You've been in the pipeline...)

User avatar
Hannibal 2
Posts: 485
Joined: Sun Nov 20, 2016 4:46 am
Location: Derbyshire UK

Re: Is EPR best avoided?

Post by Hannibal 2 » Fri Mar 24, 2017 11:33 am

Oltremare wrote:Hello to all.
The EPR management (or equivalent) is a subjective thing.
But I don't understand why you have doubts on the use of the EPR?
There are some people who have complications for their CA when they use the EPR, but everyone else, why not use EPR? It 'a prejudice?
EPR for me is a great thing!
I could never do without it.
I tried to go from 3 to 2 but I did not succeed!
After two nights, I had to go to 3 but I have no CA.

Oltremare
I don't have doubts, I know it's good for some people, just wondering if it's better to try and adjust to not using it.