Does EPR truly lower Central Apnea?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ozij
Posts: 10190
Joined: Fri Mar 18, 2005 11:52 pm

Re: Does EPR truly lower Central Apnea?

Post by ozij » Mon Aug 14, 2023 10:45 am

Pugsy wrote:
Mon Aug 14, 2023 10:29 am
ozij wrote:
Mon Aug 14, 2023 10:10 am
Why no EPAP line when she's at less than 7, and EPR is presumably functioning at less than 3? Is that an OSCAR issue?
I don't know.
EPAP could be at 4 cm so we can't see the green EPAP line when it is on the baseline number. I think that is probably why we can't "see" it...but it's probably there at the baseline.
But obviously there are times where EPR can kick in but we don't see it in action at these settings.
Got it. Thanks.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Does EPR truly lower Central Apnea?

Post by palerider » Mon Aug 14, 2023 2:21 pm

ozij wrote:
Mon Aug 14, 2023 10:10 am
Pugsy wrote:
Mon Aug 14, 2023 8:00 am
Essentially you get the amount of reduction available within the constraints of the machine itself.
You don't have to be using the 7 cm to get any exhale reduction...it's just you have to be using at least 7 to get the full 3 cm drop.
If using 7 cm minimum and setting of 3 EPR then you can get the full 3 cm drop
If using 6 cm minimum and setting of 3 EPR then you will only get a 2 cm drop while at 6 cm.
If using 5 cm minimum and setting of 3 EPR then you will only get a 1 cm drop while at 5 cm.
Something I don't understand:
When using EPR at higher pressures, I get a red line for IPAP in the Pressure chart, and a green one for the EPAP, the green line being an EPR's distance below the red.

RES9372 has EPR set to 3, full time. The Pressure chart only shows a minimal blip of a green EPAP line at about 06:15, when she reaches IPAP 7.

Why no EPAP line when she's at less than 7, and EPR is presumably functioning at less than 3? Is that an OSCAR issue?
It's hidden behind the baseline of the chart at 4. If the chart minimum was set to 3, you'd see the EPAP at 4.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
lazarus
Posts: 1327
Joined: Sat Mar 12, 2022 3:21 pm
Location: Just above NJ.

Re: Does EPR truly lower Central Apnea?

Post by lazarus » Mon Aug 14, 2023 6:51 pm

Re: Does EPR truly lower Central Apnea?
"In our experience, EPR may be beneficial for some patients and prevents the need to convert to BPAP for tolerance. Some patients only need EPR during the ramp period. In other patients, the drop in pressure may induce more treatment-emergent central events and intolerance, so we recommend that the patient be treated with and without EPR on setup to determine what is most comfortable rather than be set for all patients. Similarly, some patients find starting at low pressures suffocating (often elicited by asking if they feel they have to suck in air) and prefer to start with no ramp or a higher start ramp setting, which can be tested on setup."--Kryger's Principles and Practice of Sleep Medicine, p. 1262. Meir H. Kryger, Thomas Roth, Cathy A Goldstein
Elsevier Health Sciences, Dec 16, 2021. https://books.google.com/books?id=yDVVE ... ea&f=false

dataq1
Posts: 829
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Does EPR truly lower Central Apnea?

Post by dataq1 » Tue Aug 15, 2023 1:54 pm

going back to the very beginning of this topic.....
Did your doctor prescribe the constant 4 cm that you showed in your first post?
If not, what did he prescribe?
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

RES9372
Posts: 21
Joined: Fri Jun 16, 2023 9:49 am

Re: Does EPR truly lower Central Apnea?

Post by RES9372 » Wed Aug 16, 2023 8:30 pm

dataq1 wrote:
Tue Aug 15, 2023 1:54 pm
going back to the very beginning of this topic.....
Did your doctor prescribe the constant 4 cm that you showed in your first post?
If not, what did he prescribe?
They prescribed a fixed pressure of 7. But my AHI was even higher on that.

User avatar
Captain_Midnight
Posts: 761
Joined: Sat Jan 28, 2006 8:10 pm
Location: The Great State of Idaho

Re: Does EPR truly lower Central Apnea?

Post by Captain_Midnight » Wed Aug 16, 2023 8:50 pm

I'm a rare visitor hereabouts, however this is a topic of interest to me.

I recall reading, maybe from here, that when one uses APAP, the lower pressure needs to be increased (I think) 1 cm for each level of EPR.

I use an EPR of 1, and I did increase my lower APAP setting by one from my original titrated pressure (don't set your lower pressure at zero). My AHI stays at a consistent 0.4, if that's an indicator.

Just my observation, not a recommendation. As always, do your own DD.

Regards all - Capt Midnight

_________________
Machine: DreamStation Auto CPAP Machine
Mask: ComfortGel Blue Nasal CPAP Mask with Headgear
Additional Comments: APAP range = 10 - 12.5 In H20

dataq1
Posts: 829
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Does EPR truly lower Central Apnea?

Post by dataq1 » Wed Aug 16, 2023 9:47 pm

RES9372 wrote:
Wed Aug 16, 2023 8:30 pm
dataq1 wrote:
Tue Aug 15, 2023 1:54 pm
going back to the very beginning of this topic.....
Did your doctor prescribe the constant 4 cm that you showed in your first post?
If not, what did he prescribe?
They prescribed a fixed pressure of 7. But my AHI was even higher on that.
Ok, so you started at 7 cm constant, but to achieve a lower AHI, you changed to 4 cm constant.
Then you began to experiment with EPR settings.
Now (8-13-23) your settings are on Auto (5.0 to 9.2) , Full time EPR of 3 and your seeing Clear Airway Index of 6.4 and Obstructive Index of 0.0
Is that a fair summary?
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

RES9372
Posts: 21
Joined: Fri Jun 16, 2023 9:49 am

Re: Does EPR truly lower Central Apnea?

Post by RES9372 » Thu Aug 17, 2023 4:37 pm

dataq1 wrote:
Wed Aug 16, 2023 9:47 pm
RES9372 wrote:
Wed Aug 16, 2023 8:30 pm
dataq1 wrote:
Tue Aug 15, 2023 1:54 pm
going back to the very beginning of this topic.....
Did your doctor prescribe the constant 4 cm that you showed in your first post?
If not, what did he prescribe?
They prescribed a fixed pressure of 7. But my AHI was even higher on that.
Ok, so you started at 7 cm constant, but to achieve a lower AHI, you changed to 4 cm constant.
Then you began to experiment with EPR settings.
Now (8-13-23) your settings are on Auto (5.0 to 9.2) , Full time EPR of 3 and your seeing Clear Airway Index of 6.4 and Obstructive Index of 0.0
Is that a fair summary?
Yep exactly.

RES9372
Posts: 21
Joined: Fri Jun 16, 2023 9:49 am

Re: Does EPR truly lower Central Apnea?

Post by RES9372 » Thu Aug 17, 2023 6:48 pm

Here is the latest screen shot from last night. I am wondering if I should raise the minimum pressure?

Image

dataq1
Posts: 829
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Does EPR truly lower Central Apnea?

Post by dataq1 » Thu Aug 17, 2023 8:48 pm

RES9372 wrote:
Thu Aug 17, 2023 6:48 pm
Here is the latest screen shot from last night. I am wondering if I should raise the minimum pressure?
Not responsive to you question, but could you change the Y-axis on your pressure curve to 0-15, makes it easier to see the EPR.

It is my personal preference to avoid changing parameters (pressure/epr settings) every day. Again, it's just what I do, but I'd very leery of trying to draw conclusions from one or two days. Again, it's just my preference, but unless some setting is very very bad, I (personally) would leave conditions alone for a week before considering further jiggering.

That said, are you certain that the clear airway events you are experiencing are NOT a result of sleep-wake-junk?
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
ozij
Posts: 10190
Joined: Fri Mar 18, 2005 11:52 pm

Re: Does EPR truly lower Central Apnea?

Post by ozij » Thu Aug 17, 2023 9:37 pm

dataq1 wrote:
Thu Aug 17, 2023 8:48 pm

Not responsive to you question, but could you change the Y-axis on your pressure curve to 0-15, makes it easier to see the EPR.
If the machine does not go beneath a pressure of 4, changing the axis show anything underneath 4 is unnecessary and will not teach you anything about your therapy.
RES9372 wrote:
Thu Aug 17, 2023 6:48 pm
Here is the latest screen shot from last night. I am wondering if I should raise the minimum pressure?
How long have you been on this setting?
Stick to it for a few day - your body has to get used to it

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

dataq1
Posts: 829
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Does EPR truly lower Central Apnea?

Post by dataq1 » Thu Aug 17, 2023 9:57 pm

ozij wrote:
Thu Aug 17, 2023 9:37 pm
If the machine does not go beneath a pressure of 4, changing the axis show anything underneath 4 is unnecessary and will not teach you anything about your therapy.
True, but it would avoid people having to ask ‘where is the EPR curve ?’
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Does EPR truly lower Central Apnea?

Post by palerider » Fri Aug 18, 2023 12:15 pm

A minimum pressure of 7, maybe 8 would give you less breathing issues that disturb sleep.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.