Aaaaand again: newbie struggles topic
Aaaaand again: newbie struggles topic
Hi I have been diagnosed July 2017 but have been on therapy on and off since Feb 2018. And I can't understand what's wrong. I have tried 2 masks - first
ComfortGel Blue Full Face and then DreamWear Nasal. I didn't have software until some days ago. I only new therapy is not working. I tried again last night the full face and statistics don't look good to me also.
Please help me figure out what is the problem. I am a side sleeper and masks are really bothering me, I fall asleep hard, wake up too often, go to WC often but still not sure whether while I am asleep, CPAP therapy is working. It is really overwhelming.
Let me know if I should snip other info from statistics.
Thank you all in advance!
ComfortGel Blue Full Face and then DreamWear Nasal. I didn't have software until some days ago. I only new therapy is not working. I tried again last night the full face and statistics don't look good to me also.
Please help me figure out what is the problem. I am a side sleeper and masks are really bothering me, I fall asleep hard, wake up too often, go to WC often but still not sure whether while I am asleep, CPAP therapy is working. It is really overwhelming.
Let me know if I should snip other info from statistics.
Thank you all in advance!
- Attachments
-
- Nasal 18.04.JPG (17.77 KiB) Viewed 4840 times
-
- Full face 21.04.JPG (19.65 KiB) Viewed 4840 times
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Aaaaand again: newbie struggles topic
We need to know the settings you have and see the graphs that show what is going on.
Also, can you tell us what problems you feel you are having? Then we can help with better advice.
Also, can you tell us what problems you feel you are having? Then we can help with better advice.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: Aaaaand again: newbie struggles topic
You need to post the correct data from Sleepyhead for help
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead
how to post your data https://sleep.tnet.com/reference/tips/imgur
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead
how to post your data https://sleep.tnet.com/reference/tips/imgur
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Amara View Full Face CPAP Mask with Headgear |
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose |
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm
Amara View mask
O2 - 2-4 lpm
Re: Aaaaand again: newbie struggles topic
My settings are 4-14 cmH2O.
I feel very tired. Exhausted, to be more specific. Very nervous and irritated and basically depressed.
I read the necessary graphs to show but I didn't find anywhere how to show one if it's missing. I am attaching the view I have now.
https://imgur.com/a/rhB3la1
I feel very tired. Exhausted, to be more specific. Very nervous and irritated and basically depressed.
I read the necessary graphs to show but I didn't find anywhere how to show one if it's missing. I am attaching the view I have now.
https://imgur.com/a/rhB3la1
Re: Aaaaand again: newbie struggles topic
nova4e
Your minimum pressure is very low, It is possible that you are feeling like you aren't getting enough air. My suggestion would be to raise your minimum pressure to 6. I noticed that when your pressure drops down to 4-5 you start having hypopneas. I am sure more knowledgeable people will be along shortly.
Your minimum pressure is very low, It is possible that you are feeling like you aren't getting enough air. My suggestion would be to raise your minimum pressure to 6. I noticed that when your pressure drops down to 4-5 you start having hypopneas. I am sure more knowledgeable people will be along shortly.
Re: Aaaaand again: newbie struggles topic
It is easy to see why you feel poorly.
As gwc says, you need to increase your minimum pressure. Start with 6 and then reevaluate after a few days.
As gwc says, you need to increase your minimum pressure. Start with 6 and then reevaluate after a few days.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/ |
Use data to optimize your xPAP treatment:
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead/shorganize
how to post your data https://sleep.tnet.com/reference/tips/imgur
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead/shorganize
how to post your data https://sleep.tnet.com/reference/tips/imgur
Re: Aaaaand again: newbie struggles topic
Thank you both. Will do this night for sure. And will use my nasal mask, because the full face one is making my nose really sore.
Hope to get the good results finally.
Hope to get the good results finally.
- Okie bipap
- Posts: 3552
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: Aaaaand again: newbie struggles topic
You can use moleskin to protect your nose when using your full face mask.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
- Islandwoman
- Posts: 319
- Joined: Sun Dec 22, 2013 8:15 pm
- Location: Pacific Northwest
Re: Aaaaand again: newbie struggles topic
I had similar problems with the Wisp mask. Discovered that plastic type masks caused irritation. I switched to a Circadiance soft cloth mask and have been using them for nearly five years. I too am a side sleeper and move around a lot.
Re: Aaaaand again: newbie struggles topic
How is the leak? I hear for the first time for cloth masks. Sounds soft and much more comfortable.
Last days I have a really sore neck (and because of that - awful headache) from trying to keep my head so that the pillow doesn't press the mask and it doesn't leak.
Can't remember when I felt good for last time actually
Last days I have a really sore neck (and because of that - awful headache) from trying to keep my head so that the pillow doesn't press the mask and it doesn't leak.
Can't remember when I felt good for last time actually
- chunkyfrog
- Posts: 34390
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Aaaaand again: newbie struggles topic
Sleepweaver cloth cpap masks, by Circadiance.
Tricky to adjust and wear, but if you can master "the dance"--pure heaven!
Tricky to adjust and wear, but if you can master "the dance"--pure heaven!
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Aaaaand again: newbie struggles topic
Here are the last 2 days:
- Sunday https://imgur.com/a/EFloIP2
- Monday https://imgur.com/a/ga8KE9l
The statistics look good, but still I don't feel rested.
Last night I woke up after 4 and couldn't sleep much after. Turned off the CPAP around 5 but still no sleep. What am I doing wrong?!
P.S. Could you tell me how to show the other charts - Flow Rate, Snore & Flow Limit? And also get all other statistics on the left? Like here https://imgur.com/a/sAGQbss
- Sunday https://imgur.com/a/EFloIP2
- Monday https://imgur.com/a/ga8KE9l
The statistics look good, but still I don't feel rested.
Last night I woke up after 4 and couldn't sleep much after. Turned off the CPAP around 5 but still no sleep. What am I doing wrong?!
P.S. Could you tell me how to show the other charts - Flow Rate, Snore & Flow Limit? And also get all other statistics on the left? Like here https://imgur.com/a/sAGQbss
Re: Aaaaand again: newbie struggles topic
With your min pressure set to 6, your data looks good. Give it some time (a few weeks) to see if you feel better.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/ |
Use data to optimize your xPAP treatment:
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead/shorganize
how to post your data https://sleep.tnet.com/reference/tips/imgur
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead/shorganize
how to post your data https://sleep.tnet.com/reference/tips/imgur
Re: Aaaaand again: newbie struggles topic
I think the stuff is missing because your machine doesn't do all those other graphs that the other brands of machines do.nova4e wrote: ↑Mon Apr 23, 2018 9:01 pmCould you tell me how to show the other charts - Flow Rate, Snore & Flow Limit? And also get all other statistics on the left? Like here https://imgur.com/a/sAGQbss
SleepyHead can only report whatever data your machine gathers and the Icon data isn't as comprehensive as the other brands.
The SA on the Icon is SenseAwake...the machine thinks you woke up several times...and you mention waking up and not being able to go back to sleep.
So how many hours of real sleep and not just having the machine on do you think you might be doing?
There's so much more to feeling rested than just getting nice low numbers.
Getting the nice low numbers is the easy part.
Do you take any medications of any kind, if so, what?
How many hours of actual sleep?
Are those hours of real sleep fragmented by very many wake ups? If so, any idea why.
How often do you wake and not be able to go to sleep?
Any other physical or mental health issues going on?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Aaaaand again: newbie struggles topic
Yes, totally agree. It is not just the numbers because there's so much more behind them.
I really don't know how much sleep I get. I do have light sleep though. Many times I turn, I am not actually sleeping, because I try to get the mask set so that it is not pressed to my nose, neither away and leaking. I get the hose positioned so that it is not pressed. And this gets me out of the deep sleep. And next is the mental part - because I am "rushing" to fall asleep again and not waste any more time, but I actually don't succeed, I get frustrated. Then often I need to go to WC. (By the way when you go to WC, do you usually turn of the CPAP machine, or just unplug the hose and leave it to leak?). Then my daughter would sleep-talk or something else ... You get the picture And this is several times each night. I intentionally don't watch the clock on each wake, only sometimes when I go to WC.
Other than that I have Hashimoto's and I take Eutirox daily. No other physical or mental health issues.
I really don't know how much sleep I get. I do have light sleep though. Many times I turn, I am not actually sleeping, because I try to get the mask set so that it is not pressed to my nose, neither away and leaking. I get the hose positioned so that it is not pressed. And this gets me out of the deep sleep. And next is the mental part - because I am "rushing" to fall asleep again and not waste any more time, but I actually don't succeed, I get frustrated. Then often I need to go to WC. (By the way when you go to WC, do you usually turn of the CPAP machine, or just unplug the hose and leave it to leak?). Then my daughter would sleep-talk or something else ... You get the picture And this is several times each night. I intentionally don't watch the clock on each wake, only sometimes when I go to WC.
Other than that I have Hashimoto's and I take Eutirox daily. No other physical or mental health issues.