Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
I changed the filter (using another hypoallergenic thin filter) in my VAuto a couple days ago, and the machine started acting weird.
It wouldn't shut off like it normally would, the fan would run at a lower speed, and I could feel some vibrating in the machine. Additionally, it wouldn't deliver air like normal after it got into that strange new behavior pattern, when I'd inhale, I could feel and hear a vibration in the airflow. A couple times it was making the wheezing noises that so many have complained about.
What I finally figured out is that the new filter was getting sucked down onto the opening, so it was causing a lot more airflow resistance. I popped open the filter cover, while it was acting weird, noticed the filter was sucked into the opening, pulled the filter out, and the airflow picked up to where it always was before when I'd take the mask off, and then smart start shut off the machine, just like it is supposed to do.
playing with it more, I found that if I put the filter in the square opening of the filter cover door, it would get sucked into the intake, If however I sat the filter in the opening of the housing, and then closed the door, it would pinch the bottom edge of the filter in place, and the filter would stay, and the machine would work fine.
I guess pictures are in order.
How I thought I should put in a new filter:
What happened:
What worked:
There needs to be a little more support around the actual air inlet when using just the thin hypoallergenic filters, but pinching the filter in the bottom of the door seems to be working for the time being
There needs to be a little more support around the inner opening,
It wouldn't shut off like it normally would, the fan would run at a lower speed, and I could feel some vibrating in the machine. Additionally, it wouldn't deliver air like normal after it got into that strange new behavior pattern, when I'd inhale, I could feel and hear a vibration in the airflow. A couple times it was making the wheezing noises that so many have complained about.
What I finally figured out is that the new filter was getting sucked down onto the opening, so it was causing a lot more airflow resistance. I popped open the filter cover, while it was acting weird, noticed the filter was sucked into the opening, pulled the filter out, and the airflow picked up to where it always was before when I'd take the mask off, and then smart start shut off the machine, just like it is supposed to do.
playing with it more, I found that if I put the filter in the square opening of the filter cover door, it would get sucked into the intake, If however I sat the filter in the opening of the housing, and then closed the door, it would pinch the bottom edge of the filter in place, and the filter would stay, and the machine would work fine.
I guess pictures are in order.
How I thought I should put in a new filter:
What happened:
What worked:
There needs to be a little more support around the actual air inlet when using just the thin hypoallergenic filters, but pinching the filter in the bottom of the door seems to be working for the time being
There needs to be a little more support around the inner opening,
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
I have the same machine that you do. I've never had an issue with the filter like you describe. I always place the filter on the door of the compartment when replacing it and have never had an issue.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
Do not regret growing older. It is a privilege denied to many...
- raisedfist
- Posts: 1176
- Joined: Wed Jun 15, 2016 7:21 am
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
I read a post about this recently...the hypoallergenic filter getting sucked in. I find my humidifier chamber gets stuck sometimes where I have to yank harder than I'd like to get it out to clean it.
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12
- chunkyfrog
- Posts: 34432
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
Is the filter OEM, or third party?
I don't know what mine are, but never have that problem.
Knock on wood . . .
A brilliant solution, nonetheless!
I don't know what mine are, but never have that problem.
Knock on wood . . .
A brilliant solution, nonetheless!
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- Dog Slobber
- Posts: 3967
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
Thanks for this, we'll add it to the list of things to check for when noisey or odd behaving machines.
Given the HA filter is thinner, more pliable and less porous it would be more prone to getting sucked in. Well done.
Given the HA filter is thinner, more pliable and less porous it would be more prone to getting sucked in. Well done.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
Might be third party. The weird thing is that the previous one from the same package had no problem.chunkyfrog wrote: ↑Wed Oct 05, 2022 9:33 amIs the filter OEM, or third party?
I don't know what mine are, but never have that problem.
Knock on wood . . .
A brilliant solution, nonetheless!
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
The point, as it were, wasn't that this WOULD happen, but that it *CAN* happen, and if the machine starts making unexpected noises (a complaint that gets heard, no pun intended, check the filter.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
thanks for the tip. i don't use hypoallergenic filters and usually put the filter against the port.
so far so good.
now, if i can just remember the tip to pass on when someone complains of noise.
so far so good.
now, if i can just remember the tip to pass on when someone complains of noise.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
PR:
I thought that you were once in the double-filtering camp.
No longer so?
viewtopic/t157518/Doubling-up-on-ResMed ... -idea.html
I thought that you were once in the double-filtering camp.
No longer so?
viewtopic/t157518/Doubling-up-on-ResMed ... -idea.html
_________________
Machine: Airsense 10 Card to Cloud |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Last edited by lazarus on Wed Oct 05, 2022 12:37 pm, edited 1 time in total.
The people who confuse "entomology" and "etymology" really bug me beyond words.
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
I usually do, but didn't have a coarse filter handy the last two times I changed.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
I would make a joke about filtering out coarseness, but I'm no longer paid by the joke here.
_________________
Machine: Airsense 10 Card to Cloud |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Last edited by lazarus on Wed Oct 05, 2022 12:40 pm, edited 1 time in total.
The people who confuse "entomology" and "etymology" really bug me beyond words.
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
Someone get this guy back on the payroll!
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34432
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
Yeah, we need the yuks; and we have no taste . . .
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
I was told this year's entire budget went toward finding Zonker some new gifs.
StarKist doesn't want frogs with good taste . . .
All in all, a very helpful thread on possible problems with filters.
In my case, I never change them often enough.
Something else I'm waay too lazy to put a reminder for inside my phone's calendar.
_________________
Machine: Airsense 10 Card to Cloud |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
The people who confuse "entomology" and "etymology" really bug me beyond words.
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
- chunkyfrog
- Posts: 34432
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically
My strategy is my spiral notebook.
I draw up a manual spreadsheet, with 21 lines for 21 days. (fits the page)
Day of week/month is followed by hours, AHI, pressure, leak, CI, and a few others,
as requested by my doctor.
At the end of each page, I swap pillows and filters.
Medicare allows for 2 a month, so I am a bit ahead.
Good for when I run out of credit--already out of money . . .
I draw up a manual spreadsheet, with 21 lines for 21 days. (fits the page)
Day of week/month is followed by hours, AHI, pressure, leak, CI, and a few others,
as requested by my doctor.
At the end of each page, I swap pillows and filters.
Medicare allows for 2 a month, so I am a bit ahead.
Good for when I run out of credit--already out of money . . .
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |