Difficulty removing hose from APAP
- MurrayNevada
- Posts: 196
- Joined: Sun May 20, 2018 8:25 am
Difficulty removing hose from APAP
I just completed my first week on APAP. Things seem to be going well thanks to assistance from Pugsy. My Problem: Removing the hose from the machine was very difficult. The soft rubber end of the hose was very difficult to move. I got it off but had to really pull and twist it with some force to do it. I'm afraid that doing this weekly will eventually do some damage to the APAP. Any ideas or tricks to make removal easier?
Thank you
Thank you
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
Additional Comments: ResMed S9 AutoSet for Travel & Backup |
Re: Difficulty removing hose from APAP
You should just be able to click it off very easy no force is nessescary... Push on either side and comes right off..
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ResMed S9 VPAP ST |
Re: Difficulty removing hose from APAP
OP is using a DreamStation machine...not ResMed...they use a different hose to humidifier connecting hose.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
-
- Posts: 290
- Joined: Mon Dec 12, 2016 11:35 pm
Re: Difficulty removing hose from APAP
The end of the hose that connects to the machine has a sort of clip arrangement: squeeze both the pads on the hose end to release it. (I'm making the assumption you also have the heated hose for the humidifier. . . )
However, the other end of the hose, that connects to the mask, is simply a rubber seal. Try this trick that my resp. tech. taught me: grip the hose end in one hand and the mask's connector in the other, palms facing down; now rotate both your wrists so palms face each other (as if you were snapping a twig in two). The rubber seal of the hose should come free of the mask's connector.
If you have the regular hose, though, try a similar technique to release it at the machine end by holding the machine firmly, with humidifier cover closed; grip the hose end and push it to one side (twist the wrist as in the abovenoted technique).
I hope this makes sense! It's harder to describe than actually do.
Jean
Edited to correct my assumption on initial composition.
However, the other end of the hose, that connects to the mask, is simply a rubber seal. Try this trick that my resp. tech. taught me: grip the hose end in one hand and the mask's connector in the other, palms facing down; now rotate both your wrists so palms face each other (as if you were snapping a twig in two). The rubber seal of the hose should come free of the mask's connector.
If you have the regular hose, though, try a similar technique to release it at the machine end by holding the machine firmly, with humidifier cover closed; grip the hose end and push it to one side (twist the wrist as in the abovenoted technique).
I hope this makes sense! It's harder to describe than actually do.
Jean
Edited to correct my assumption on initial composition.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
Humidifier: DreamStation Heated Humidifier |
Last edited by Soothest Sleep on Sat May 26, 2018 9:47 am, edited 1 time in total.
O soft embalmer of the still midnight,
Shutting, with careful fingers and benign,
Our gloom-pleas'd eyes, embower'd from the light,
Enshaded in forgetfulness divine
-- John Keats
Shutting, with careful fingers and benign,
Our gloom-pleas'd eyes, embower'd from the light,
Enshaded in forgetfulness divine
-- John Keats
Re: Difficulty removing hose from APAP
The rubber end of the hose that attaches to the humidifier connection will eventually become less stiff. I don't know of any way to speed up the process.
I assume you aren't using the heated hose and are just using the standard non heated hose that just has the rubber coupling thing.
And yeah, it can be pretty darn snug.
It is not necessary to remove the hose and wash it weekly which is what I assume you are wanting to do because someone told you to do that.
Especially since you are just using passover humidity. The inside of the hose really isn't going to get enough moisture in it to give any bugs a chance to grow.
I never wash my hose.. but that's me.
Once a month washing is more than sufficient if you feel the need to wash it.
I assume you aren't using the heated hose and are just using the standard non heated hose that just has the rubber coupling thing.
And yeah, it can be pretty darn snug.
It is not necessary to remove the hose and wash it weekly which is what I assume you are wanting to do because someone told you to do that.
Especially since you are just using passover humidity. The inside of the hose really isn't going to get enough moisture in it to give any bugs a chance to grow.
I never wash my hose.. but that's me.
Once a month washing is more than sufficient if you feel the need to wash it.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Difficulty removing hose from APAP
It's early .. just saw that ....
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ResMed S9 VPAP ST |
Re: Difficulty removing hose from APAP
Don't feel bad...I needed that second cup of coffee myself because my first response in my head was "just unhook it" and then woke up remembered this person.
He hadn't even been using a humidifier until the other day and he's using the DreamStation and just took delivery of the humidifier a few days ago and this is probably the first attempt to remove the hose to wash it.
Pretty sure a non heated hose because the DreamStation heated hose has something similar to the ResMed heated hose..not exactly the same but close enough.
The only way it would be hard to remove is if it is the standard rubber non heated hose and those can be a bitch sometimes.
I used a non heated hose not long ago to air out a machine that I was sending someone and I had a heck of a time getting it off and I barely slid it on the connection. It was hard to get on and hard to get off.
I don't know of anyway to soften the rubber on the end of the hose that attaches to the humidifier unless maybe some heat of some sort.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Difficulty removing hose from APAP
You didn't say whether you're using a heated hose (specific for Dream Station CPAPs), or "standard" hose (that will fit almost any CPAP).
If you're using the former, you should be gently pressing the two white release latches on the hose as you remove it. If it's the latter, you might try reversing it (put the mask end on the machine and vice-versa).
If you're using the former, you should be gently pressing the two white release latches on the hose as you remove it. If it's the latter, you might try reversing it (put the mask end on the machine and vice-versa).
- chunkyfrog
- Posts: 34391
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Difficulty removing hose from APAP
Non-heated hoses need to be wiggled to separate.
I learned early on to hold onto the rubber part only--and not the hose itself.
---The hard way. I destroyed my first hose.
I learned early on to hold onto the rubber part only--and not the hose itself.
---The hard way. I destroyed my first hose.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- MurrayNevada
- Posts: 196
- Joined: Sun May 20, 2018 8:25 am
Re: Difficulty removing hose from APAP
I'm using a standard hose. Thanks everyone for your help.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
Additional Comments: ResMed S9 AutoSet for Travel & Backup |
Re: Difficulty removing hose from APAP
What I do is wiggle the hose side to side, starting gently and working up, if needed, to a more forceful sideways force, and that will start to break the grip the hose has on the plastic nipple.... then it can be worked off.MurrayNevada wrote: ↑Sat May 26, 2018 12:00 pmI'm using a standard hose. Thanks everyone for your help.
I hardly ever remove my hose though, and in the months between changes, it gets a pretty good grip.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Difficulty removing hose from APAP
Think a kid's Chinese Handcuff's the harder you pull straight off the tighter it grips. Jimpalerider wrote: ↑Sat May 26, 2018 1:31 pmWhat I do is wiggle the hose side to side, starting gently and working up, if needed, to a more forceful sideways force, and that will start to break the grip the hose has on the plastic nipple.... then it can be worked off.MurrayNevada wrote: ↑Sat May 26, 2018 12:00 pmI'm using a standard hose. Thanks everyone for your help.
I hardly ever remove my hose though, and in the months between changes, it gets a pretty good grip.
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Difficulty removing hose from APAP
That's a good point.. wrote: ↑Sat May 26, 2018 1:54 pmThink a kid's Chinese Handcuff's the harder you pull straight off the tighter it grips. Jimpalerider wrote: ↑Sat May 26, 2018 1:31 pmWhat I do is wiggle the hose side to side, starting gently and working up, if needed, to a more forceful sideways force, and that will start to break the grip the hose has on the plastic nipple.... then it can be worked off.MurrayNevada wrote: ↑Sat May 26, 2018 12:00 pmI'm using a standard hose. Thanks everyone for your help.
I hardly ever remove my hose though, and in the months between changes, it gets a pretty good grip.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Difficulty removing hose from APAP
Usa a Quick Connect... https://www.cpap.com/productpage/hose-q ... nnect.html
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
Re: Difficulty removing hose from APAP
I use a little silicon grease. Smear a tiny dab inside the rubber end of the hose, and a little on the plastic connector on the machine. You can find it a pool supply stores and maybe home improvement stores. It has no smell and has not harmed my equipment.
Jim
Jim
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Heated Humidifer, Non-Heated tube, Oscar |
Resmed Airsense 10 Autoset, Swift FX nasal pillows mask