Confused about run times for CPAP battery banks
-
- Posts: 104
- Joined: Tue May 14, 2013 6:57 pm
Confused about run times for CPAP battery banks
I have a Dreamstation 1 CPAP that says it uses 6.67 amps at 12 volts, or about about 80 watts of power. It appears this must include the humidifier.
How does a CPAP battery with 160 watt hours of power manage to power a Dreamstation 1 for 8 hours? Does the CPAP really not take 80 watts? My minimum pressure is 15 so my power usage would be on the higher end. At 80 watts I would need 640 watt hours minimum to power my CPAP for 8 hours. I am thinking about getting a Kill-A-Watt meter to see what it really uses.
How does a CPAP battery with 160 watt hours of power manage to power a Dreamstation 1 for 8 hours? Does the CPAP really not take 80 watts? My minimum pressure is 15 so my power usage would be on the higher end. At 80 watts I would need 640 watt hours minimum to power my CPAP for 8 hours. I am thinking about getting a Kill-A-Watt meter to see what it really uses.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: SleepWeaver Anew™ Full Face Mask with Headgear |
- ChicagoGranny
- Posts: 14545
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Confused about run times for CPAP battery banks
The blower alone draws less than 5 watts at 15 cmH2O.reader2580 wrote: ↑Wed May 03, 2023 10:57 amI have a Dreamstation 1 CPAP that says it uses 6.67 amps at 12 volts, or about about 80 watts of power. It appears this must include the humidifier.
Are you using a machine that was replaced under the DreamStation One recall?
Re: Confused about run times for CPAP battery banks
You turn off the humidifier when you're on battery.reader2580 wrote: ↑Wed May 03, 2023 10:57 amI have a Dreamstation 1 CPAP that says it uses 6.67 amps at 12 volts, or about about 80 watts of power. It appears this must include the humidifier.
How does a CPAP battery with 160 watt hours of power manage to power a Dreamstation 1 for 8 hours? Does the CPAP really not take 80 watts? My minimum pressure is 15 so my power usage would be on the higher end. At 80 watts I would need 640 watt hours minimum to power my CPAP for 8 hours. I am thinking about getting a Kill-A-Watt meter to see what it really uses.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- loggerhead12
- Posts: 561
- Joined: Fri Sep 01, 2017 4:00 pm
Re: Confused about run times for CPAP battery banks
Get the kill-a-watt. It's not using 6.67 amps.reader2580 wrote: ↑Wed May 03, 2023 10:57 amI have a Dreamstation 1 CPAP that says it uses 6.67 amps at 12 volts, or about about 80 watts of power. It appears this must include the humidifier.
-
- Posts: 104
- Joined: Tue May 14, 2013 6:57 pm
Re: Confused about run times for CPAP battery banks
My regular machine is a Resmed. Yes, this Dreamstation 1 was replaced under recall. I want to use the Dreamstation 1 for away from home and keep the Resmed for home use.ChicagoGranny wrote: ↑Wed May 03, 2023 11:50 amAre you using a machine that was replaced under the DreamStation One recall?
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: SleepWeaver Anew™ Full Face Mask with Headgear |
-
- Posts: 104
- Joined: Tue May 14, 2013 6:57 pm
Re: Confused about run times for CPAP battery banks
I am definitely planning to turn off the humidifier. I think I need to get the Kill-A-Watt to see what it really uses. I don't want to have to spend more than necessary on a battery. I can charge up every day.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: SleepWeaver Anew™ Full Face Mask with Headgear |
Re: Confused about run times for CPAP battery banks
Generally machines use less than 100wh a night, depending on pressure.reader2580 wrote: ↑Wed May 03, 2023 1:48 pmI am definitely planning to turn off the humidifier. I think I need to get the Kill-A-Watt to see what it really uses. I don't want to have to spend more than necessary on a battery. I can charge up every day.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Confused about run times for CPAP battery banks
If you're talking about a lead acid chemistry battery, you want to avoid discharging it too deeply, like 50% at most. You'll shorten its useful life span significantly if you deep cycle it. Lithium ion, probably not so much of an issue. I would still opt for some extra capacity regardless of the type of battery.reader2580 wrote: ↑Wed May 03, 2023 1:48 pmI am definitely planning to turn off the humidifier. I think I need to get the Kill-A-Watt to see what it really uses. I don't want to have to spend more than necessary on a battery. I can charge up every day.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Oracle HC452 Oral CPAP Mask |
Re: Confused about run times for CPAP battery banks
I don't know by the numbers, but I have a 200W portable "power station" (FlashFish Brand). I don't use a humidifier and I use about 2/3 of the power in 8 hours with my CPAP. This is a nice size (about the size of a shoebox), fairly lightweight, and it can be recharged with AC, the car battery, or solar (if you have a solar panel--they bundle a 50W portable panel for about $75 more). It would be OK for 1 night of camping, but not two. They have larger ones, too.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Re: Confused about run times for CPAP battery banks
afaik the Kill-A-Watt only measures AC - you really want to know what the DC power draw is.reader2580 wrote: ↑Wed May 03, 2023 1:48 pmI think I need to get the Kill-A-Watt to see what it really uses.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: ↑What exactly do you think you're adding to this thread?
- chunkyfrog
- Posts: 34449
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Confused about run times for CPAP battery banks
Plug the Kill-a-Watt between the main power cord and the wall outlet.
Once you know the AC wattage, then DC is also measured in watts.
It will at least give you a rough idea.
Once you know the AC wattage, then DC is also measured in watts.
It will at least give you a rough idea.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Confused about run times for CPAP battery banks
Seems like you'll get pretty close with a kill-a-watt, in the ballpark anyhow. This battery camping business is a little tricky with the power usage. You'll use more watt hours if you're doing an inverter vs direct DC, worse than a direct with 12V DC to 24V DC converter. The inverter IDK how much extra you would need but you definitely lose power in the conversion. Steer clear of that option if you can. That was one nice feature of the PRS1 series, the 12V direct connection with no conversion. Don't forget to turn off the heated humidifier and hose; you'll blow thru 100% of a modest battery in a few hours and regret it.SleepGeek wrote: ↑Thu May 04, 2023 7:09 pmafaik the Kill-A-Watt only measures AC - you really want to know what the DC power draw is.reader2580 wrote: ↑Wed May 03, 2023 1:48 pmI think I need to get the Kill-A-Watt to see what it really uses.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Oracle HC452 Oral CPAP Mask |
Re: Confused about run times for CPAP battery banks
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: ↑What exactly do you think you're adding to this thread?
-
- Posts: 104
- Joined: Tue May 14, 2013 6:57 pm
Re: Confused about run times for CPAP battery banks
I will be using Lifepo4, not lead acid. The AC adapter uses some power, but measuring AC usage would get pretty close. I am planning on a direct DC connection to the Dreamstation.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: SleepWeaver Anew™ Full Face Mask with Headgear |
- loggerhead12
- Posts: 561
- Joined: Fri Sep 01, 2017 4:00 pm
Re: Confused about run times for CPAP battery banks
If you want to get real precise, this is what I use to monitor my battery setup.reader2580 wrote: ↑Fri May 05, 2023 7:21 pmI will be using Lifepo4, not lead acid. The AC adapter uses some power, but measuring AC usage would get pretty close. I am planning on a direct DC connection to the Dreamstation.
https://www.amazon.com/dp/B079JVGRSL/re ... RhaWw&th=1