Airsense 11 Water Leak
Airsense 11 Water Leak
I have an Airsense 11 Autoset that is about 1 1/2 years old.
For the most part it's been a good machine.
For that last 8 months or so, I am having repeated issues with the humidifier tub leaking.
When it leaks I see some water puddling around the bottom of the place where the tub is inserted, and the water has a brown tinge to it.
I am filling the tub well under the max line. I have used both distilled and softened and filtered tap water, and it still occurs. It doesn't happen every night.
I have replaced tubs about 10 times. The most recent one is about 10 days old, dishwasher safe, but I only hand wash it with Dawn soap about once a week. I usually use level 3 for my humidity setting. This happens with both the ClimateLine tube, and a non-heated ResMed 15mm tube. The machine is not being tipped or jostled to make the water leak our.
I am wondering if any of the forum members have heard of this issue, and have any suggestions for a solve.
For the most part it's been a good machine.
For that last 8 months or so, I am having repeated issues with the humidifier tub leaking.
When it leaks I see some water puddling around the bottom of the place where the tub is inserted, and the water has a brown tinge to it.
I am filling the tub well under the max line. I have used both distilled and softened and filtered tap water, and it still occurs. It doesn't happen every night.
I have replaced tubs about 10 times. The most recent one is about 10 days old, dishwasher safe, but I only hand wash it with Dawn soap about once a week. I usually use level 3 for my humidity setting. This happens with both the ClimateLine tube, and a non-heated ResMed 15mm tube. The machine is not being tipped or jostled to make the water leak our.
I am wondering if any of the forum members have heard of this issue, and have any suggestions for a solve.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset |
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
- vandownbytheriver
- Posts: 388
- Joined: Sat Feb 03, 2024 11:42 pm
Re: Airsense 11 Water Leak
You should replace the tub. You can test the old tub by filling with water and letting it sit... but it may not leak until heated. Tubs are supposed to be replaced every six months according to RM. Brownish water is a sign of rust or corrosion and is not good.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
Re: Airsense 11 Water Leak
I have replaced several tubs. You are correct that heat seems to be necessary.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset |
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
- vandownbytheriver
- Posts: 388
- Joined: Sat Feb 03, 2024 11:42 pm
Re: Airsense 11 Water Leak
I tried repairing the old-style AS10 tub and wound up cracking the plastic... otherwise I think it would have worked... just very difficult to get apart.
I have the tube of food-grade silicone left over from that experiment. The new AS10 tubs have the circular bottom. Should mine start leaking I'll attempt the same repair, it should be much easier with the circular plate.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
Re: Airsense 11 Water Leak
I had the same issue with a leaking AirSense 11. Had to replace it at 6 months. I think the heating results in cracks in the plastic with resulting leaks
_________________
Machine: Airsense 11 w/ ClimateLine |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
CPAP user since Summer 2023
Re: Airsense 11 Water Leak
Checking... just replacing the tub?
I have replaced many tubs. Either there is a big problem with the tubs, or something in the machine is causing the leak.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset |
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Re: Airsense 11 Water Leak
I agree -
It seems strange that every tub leaks within a few weeks, and that this does not seem to be a rampant problem with the tubs.
I am having the machine checked out again. My DME is a good guy. I will use my backup unit for a couple of weeks until the unit is checked out again.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset |
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Re: Airsense 11 Water Leak
Have you watched the Australian guy's videos. He has one about leaks on 11 models. Not from the tub but the hose connection.
_________________
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: SleepyHead , P10 and Dreamwear FFM, Airfit N20 |
Re: Airsense 11 Water Leak
I just got the air curve 11 and I have no leaks and I didn't even have any leaks when I took the ac11 off the headboard cabinet which is behind my bed obviously to change some of the settings. So was resting on a pillow, turned it to the side, it was rotated up to a 90° angle and nothing came out. The GM at Lincare said to me you don't have to bang it in but you just have to put the water tank in firmly. However, I just got mine two days ago and maybe they changed it since the air sense 11 came out.