why?D.H. wrote:Don't forget to use it (or run it without using it) for a few hours every week.nanwilson wrote:Oops... just read this thread and thought I had better check my old F & P 604 brick that I got in 2010. It hasn't been used in over a year and very infrequently before that... to my surprise it started up just fine!!!!!!!! So, I guess I am fine with my trusty old back up brick..... I don't need a power outage to crop up and not know it won't power on Better stop calling it my "old brick" too, it just might decide to not work if I'm nasty to it
Cheers and thanks for the heads up guys.
Na
fisher paykel icon no power
Re: fisher paykel icon no power
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
Re: fisher paykel icon no power
D.H. wrote:nanwilson wrote:Oops... just read this thread and thought I had better check my old F & P 604 brick that I got in 2010. It hasn't been used in over a year and very infrequently before that... to my surprise it started up just fine!!!!!!!! So, I guess I am fine with my trusty old back up brick..... I don't need a power outage to crop up and not know it won't power on Better stop calling it my "old brick" too, it just might decide to not work if I'm nasty to it
Cheers and thanks for the heads up guys.
Na
Don't forget to use it (or run it without using it) for a few hours every week.
WHY As I said it only gets used maybe once a year for 4 or 5 days in my motorhome. When I first got it in 2010 I only used it for about 6 months until one of the members here sold me an auto, then I'm I have only used it as my backup... total, maybe... less than 100 hours as back up, including power outages .
Good luck with telling my to run it a few hours every week, it hasn't died yet, so why should it now. It has had LITTLE use in 5 years....do you run your back up a few hours every week. I sincerely doubt it.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Started cpap in 2010.. still at it with great results.
Re: fisher paykel icon no power
It's just like a refrigerator or dishwasher. You have to run it every so often or the seals dry out. Also, should the machine cease to function, you'll know about it at a good time, not have a crisis on vacation or right before vacation.nanwilson wrote: . . .
[/color]WHY As I said it only gets used maybe once a year for 4 or 5 days in my motorhome. When I first got it in 2010 I only used it for about 6 months until one of the members here sold me an auto, then I'm I have only used it as my backup... total, maybe... less than 100 hours as back up, including power outages .
Good luck with telling my to run it a few hours every week, it hasn't died yet, so why should it now. It has had LITTLE use in 5 years....do you run your back up a few hours every week. I sincerely doubt it.
Re: fisher paykel icon no power
DH...... You can't read can you. I said..... it has only been used about 100 hours over the past 5 years... and still works as soon as I turned it on. You either need to slow down your reading and catch every word or TAKE READING LESSONS. I have noticed in the threads that you have answered, it is always the same, you only "get" half of what has been written and then give out erroneous information on what you "think" has been asked.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Started cpap in 2010.. still at it with great results.
Re: fisher paykel icon no power
you're so full of shit.D.H. wrote:It's just like a refrigerator or dishwasher. You have to run it every so often or the seals dry out. Also, should the machine cease to function, you'll know about it at a good time, not have a crisis on vacation or right before vacation.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: fisher paykel icon no power
I don't think he *can* read... or comprehend, maybe the words just go in his eyes and bounce around in that big hollow space between his ears... then fall out his mouth onto the floor, or if we're unfortunate, the keyboard.nanwilson wrote:DH...... You can't read can you. I said..... it has only been used about 100 hours over the past 5 years... and still works as soon as I turned it on. You either need to slow down your reading and catch every word or TAKE READING LESSONS. I have noticed in the threads that you have answered, it is always the same, you only "get" half of what has been written and then give out erroneous information on what you "think" has been asked.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: fisher paykel icon no power
Hi All,
Just spoke with someone that knows the problem....(working @ F&P).......
Looks like it is the CAPACITOR that does not charge....
That's why, after a long time, it starts....
Is it possible to change it????....That I don't know.....
Just spoke with someone that knows the problem....(working @ F&P).......
Looks like it is the CAPACITOR that does not charge....
That's why, after a long time, it starts....
Is it possible to change it????....That I don't know.....
Re: fisher paykel icon no power
how are you with a soldering iron and screwdriver?JeePee wrote:Hi All,
Just spoke with someone that knows the problem....(working @ F&P).......
Looks like it is the CAPACITOR that does not charge....
That's why, after a long time, it starts....
Is it possible to change it????....That I don't know.....
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34451
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: fisher paykel icon no power
I wonder if that is the only brand with that "feature" . . .
[[[I would add to PR's suggestion, "INSULATED screwdriver"]]]
[[[I would add to PR's suggestion, "INSULATED screwdriver"]]]
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: fisher paykel icon no power
well, the screwdriver was only to open it up, not poke around in the unplugged innerds.chunkyfrog wrote:I wonder if that is the only brand with that "feature" . . .
[[[I would add to PR's suggestion, "INSULATED screwdriver"]]]
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34451
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: fisher paykel icon no power
Just covering my little green butt (and your rosy one).
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: fisher paykel icon no power
.
Dropping another of his filthy Piles on the street of decent discourse
............. This Asshole who styles himself The Angel of Death
I suppose if he can call one of our members as being full of shit I can fairly call him
........................................................ an Asshole without being deleted
Abusing some Newbie who has just made a reasonable observation
telling him that he is full of "Shit " - Seemingly unaware that all can see
exactly what sort of "Excrement" it is that comes out the hole he calls
................................................................................a mouth
And of coarse this Yellow Dog ran with his tail to hide under his Foe/Button
as soon as I kicked the crippled SOB down a flight of rough Metaphorical Stairs -
............. He's so lame - so funny !
............. & neat to see the way the
............. froggy jumps to him too
.............
................ Quite a Clique
Someone should tell this THING that
people come here for Help & Advice
................. Not to be Insulted
.
.
D.H. wrote:It's just like a refrigerator or dishwasher. You have to run it every so often or the seals dry out.
Also, should the machine cease to function, you'll know about it at a good time, not have a crisis on vacation
or right before vacation.
Well there's that lowlife DOG again - - The Pileriderpalerider wrote: you're so full of shit.
Dropping another of his filthy Piles on the street of decent discourse
............. This Asshole who styles himself The Angel of Death
I suppose if he can call one of our members as being full of shit I can fairly call him
........................................................ an Asshole without being deleted
Abusing some Newbie who has just made a reasonable observation
telling him that he is full of "Shit " - Seemingly unaware that all can see
exactly what sort of "Excrement" it is that comes out the hole he calls
................................................................................a mouth
And of coarse this Yellow Dog ran with his tail to hide under his Foe/Button
as soon as I kicked the crippled SOB down a flight of rough Metaphorical Stairs -
............. He's so lame - so funny !
............. & neat to see the way the
............. froggy jumps to him too
.............
................ Quite a Clique
Someone should tell this THING that
people come here for Help & Advice
................. Not to be Insulted
.
.
Re: fisher paykel icon no power
Can we get back on topic? Last week, I complained to F&P about this. They told me to take my Premo (bought in May 2013) to my homecare provider/CPAP supplier which I did today. Turns out the capacitor issue is real for many of the machines manufactured 2012-2014. They gave me a new one, plus a hose and carrying case, no questions asked.JeePee wrote:Hi All,
Just spoke with someone that knows the problem....(working @ F&P).......
Looks like it is the CAPACITOR that does not charge....
That's why, after a long time, it starts....
Is it possible to change it????....That I don't know.....
BTW: There were a bunch of recalls in 2012-2013. My serial number was not on the list but it should have been. I have to say F&P stepped up.