Weird sensation breathing in

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
pandatx
Posts: 378
Joined: Sat Mar 16, 2013 8:07 am

Weird sensation breathing in

Post by pandatx » Tue Feb 27, 2018 4:23 pm

Hi all! I was on here a few years ago but never got used to my machine and fell off the wagon. But I am extremely motivated to try again. Today I set up my machine and settled in for a practice and I had a weird sensation, that as soon as it happened I remembered this is what drove me nuts before. So I am wondering if anyone is familiar and might have suggestions. What happens is this: I'm breathing in and out fine (not used to it but its fine) but intermittently I get this weird "skipping" on the inhale. Almost how it would sound/feel if there was a slight clog. Then it stops until the next ti e, which theres no telling when that will be. So instead of a steady inhale, it sort of feels like iii-n-ha-llll-e. Not sure if this makes any sense, but if you have had this happen you would definitely know it.

Thanks for any input at all. Keeping my fingers crossed that I can do it this time!

User avatar
Pugsy
Posts: 64161
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Weird sensation breathing in

Post by Pugsy » Tue Feb 27, 2018 4:29 pm

Verify that the diameter of the hose you are using matches what the machine thinks it is using.
You will need to go into the clinical menu setup area to see this information.
In the manual is the explanation for how to get into the clinical menu setup area if you don't know how.
https://forum.apneuvereniging.nl/pdf/S9 ... manual.pdf
The above idea is a shot in the dark if you are using a non heated hose.
If you are using the ClimateLine heated hose the machine should recognize it and this wouldn't apply.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
pandatx
Posts: 378
Joined: Sat Mar 16, 2013 8:07 am

Re: Weird sensation breathing in

Post by pandatx » Tue Feb 27, 2018 4:36 pm

Hi Pugsy! The tubing I am using is: ClimateLine™ Tubing for S9™ and H5i™ Climate Control System

rcobourn
Posts: 207
Joined: Mon Apr 03, 2017 8:25 am

Re: Weird sensation breathing in

Post by rcobourn » Tue Feb 27, 2018 5:01 pm

pandatx wrote:Hi Pugsy! The tubing I am using is: ClimateLineImage Tubing for S9Image and H5iImage Climate Control System
It's probably not the tubing then. Try removing filter to see if that is the issue. If you are using nasal pillows, try a different size. Work on mask adjustment; many beginners make the mask too tight, you could be collapsing the nasal pillows. Most likely you need to take the machine in to your provider and have them test it while in use.

User avatar
Pugsy
Posts: 64161
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Weird sensation breathing in

Post by Pugsy » Tue Feb 27, 2018 5:05 pm

Well darn. I was hoping it would be something simple like that.
You might check the filter as someone mentioned.
Other than that I have no other ideas. Sorry.

Mighty handsome super pugs in your avatar...are they yours?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
pandatx
Posts: 378
Joined: Sat Mar 16, 2013 8:07 am

Re: Weird sensation breathing in

Post by pandatx » Tue Feb 27, 2018 9:19 pm

Thank you for the replies! I will definitely check the filter. I'm not sure what to do if I need to take it to my provider. They are in Texas but I moved to NC. Hopefully, its the filter :wink:

Yes, those are my pugs. Or were. My boy (the bigger one) passed away. It was heartbreaking. But now we have our little pug, Shanti, and a scruffy rescue girl, Mika. I need to change my avatar.

rcobourn
Posts: 207
Joined: Mon Apr 03, 2017 8:25 am

Re: Weird sensation breathing in

Post by rcobourn » Tue Feb 27, 2018 9:48 pm

Probably just start over with a new provider and new equipment, if you have insurance coverage.

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Weird sensation breathing in

Post by TedVPAP » Tue Feb 27, 2018 9:57 pm

If you want you can perform a basic test by shaping the hose as a manometer and filling it with some water. Run the hose down towards the floor and then up making a "U". Add some water. The level of the water will be the same in both sides of the "U" when the machine is off. When on, the water level will go down in one side and it will go up the same amount in the other. For example, if your pressure is 10cm-H2O then the difference in water levels will now be 10cm. The machine should be able to deliver this constant pressure indefinitely so the water levels should stay constant.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Weird sensation breathing in

Post by palerider » Tue Feb 27, 2018 10:55 pm

bad info deleted.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Thu Mar 01, 2018 3:05 am, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

rcobourn
Posts: 207
Joined: Mon Apr 03, 2017 8:25 am

Re: Weird sensation breathing in

Post by rcobourn » Tue Feb 27, 2018 11:21 pm

palerider wrote:
TedVPAP wrote:
Tue Feb 27, 2018 9:57 pm
if your pressure is 10cm-H2O then the difference in water levels will now be 10cm.
This is potentially confusing.

The difference in water level on either tube, from where it started at equilibrium, will be 10cm.

The difference between the two levels when the machine is running, will be 20cm, one +10cm, one -10cm.
Thanks, Heisenberg![IMG]//uploads.tapatalk-cdn.com/201802 ... 17c645.jpg[/IMG]

User avatar
pandatx
Posts: 378
Joined: Sat Mar 16, 2013 8:07 am

Re: Weird sensation breathing in

Post by pandatx » Wed Feb 28, 2018 7:15 pm

Thank you so much for the suggestions. Very much appreciated :)

TedVPAP
Posts: 975
Joined: Sat Jun 12, 2010 10:29 am

Re: Weird sensation breathing in

Post by TedVPAP » Wed Feb 28, 2018 9:36 pm

rcobourn wrote:
Tue Feb 27, 2018 11:21 pm
palerider wrote:
TedVPAP wrote:
Tue Feb 27, 2018 9:57 pm
if your pressure is 10cm-H2O then the difference in water levels will now be 10cm.
This is potentially confusing.

The difference in water level on either tube, from where it started at equilibrium, will be 10cm.

The difference between the two levels when the machine is running, will be 20cm, one +10cm, one -10cm.
Thanks, Heisenberg![IMG]//uploads.tapatalk-cdn.com/201802 ... 17c645.jpg[/IMG]
Let's get the science right. The difference in level in my example is 10. 5cm down + 5 cm up = the applied pressure (10cm-H2O).
https://en.wikipedia.org/wiki/Pressure_measurement

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/