Recording pulse oximeters that stay on the finger well?
Recording pulse oximeters that stay on the finger well?
I'd like reviews of of inexpensive recording pulse oximeters that stay on the finger better, than finger clamp devices like the Contec CMS50DA+ I tried, even if I turn over while alseep.
In addition to staying on the finger, I prefer units that time-tag data, in a format that can be input into OSCAR.
Alternately, I would be interested in inexpensive units that also detect breathing.
Reasons:
I still want to find out if a wedge pillow and neck brace are as effective for me as my CPAP machine. (They are sufficient for me to stop snoring.) This is partly curiosity, partly I want to go back country camping. I'd rather not have to carry the machine and batteries.
Side Issue:
I sometimes breath deep while asleep, raising O2 blood levels, then stop breathing for a while. The machine initiates a Clear Airways event. The O2 levels usually don't drop below 92%. This is sometimes repeated a number of times. Can there be an adverse health effect from stopping breathing even if the O2 levels are reasonably high?
(OSCAR plot clarification: There is a time lag between breathing patterns and O2 levels. This might partly be a problem with the way I synced things: my pulse oximeter does not record time.)
In addition to staying on the finger, I prefer units that time-tag data, in a format that can be input into OSCAR.
Alternately, I would be interested in inexpensive units that also detect breathing.
Reasons:
I still want to find out if a wedge pillow and neck brace are as effective for me as my CPAP machine. (They are sufficient for me to stop snoring.) This is partly curiosity, partly I want to go back country camping. I'd rather not have to carry the machine and batteries.
Side Issue:
I sometimes breath deep while asleep, raising O2 blood levels, then stop breathing for a while. The machine initiates a Clear Airways event. The O2 levels usually don't drop below 92%. This is sometimes repeated a number of times. Can there be an adverse health effect from stopping breathing even if the O2 levels are reasonably high?
(OSCAR plot clarification: There is a time lag between breathing patterns and O2 levels. This might partly be a problem with the way I synced things: my pulse oximeter does not record time.)
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I used to use a (defective) PR DreamStation Auto; I switched. |
- Attachments
-
- oscar.png (21.1 KiB) Viewed 6607 times
Re: Recording pulse oximeters that stay on the finger well?
I use the Wellue SleepU and it stays on my finger fine all night. The Contec CMS50F I also have that uses the "duckbill" type sensor. I tend to knock that one around a bit and get odd reading now and then. Neither are really inexpensive though.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear |
MacbookAir (Intel) macOS Monterey (12.6.8)
OSCAR 1.5.1
Resmed S10 Autoset w/heated hose
F30 Mask
Pressure 9-14
OSCAR Team
OSCAR 1.5.1
Resmed S10 Autoset w/heated hose
F30 Mask
Pressure 9-14
OSCAR Team
Re: Recording pulse oximeters that stay on the finger well?
The sleep clinic told me to tape it on…worked fine.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
ResMed AirSense Autoset FOR HER/F30i
Re: Recording pulse oximeters that stay on the finger well?
Thanks for the suggestions!
Oh! So obvious. I am going to try this. If it works, I will feel like such an idiot. Even if it doesn't, I should have tried it. Thanks!
In retrospect, my OSCAR example was bad - only in one of the cases shown did the oxygen level drop. A combination treatment - CPAP machine selection, mask selection, pillow selection, neck brace, sleep position choice, dental mouth guard, and what I think of as a slow gradual training process (even though many experts say you can't learn during sleep) - has been successful for me - about 1 AHI / night recorded by my CPAP machine (even though it is set on the lowest possible initial pressure), which is within the normal zone. So I have trouble finding good examples where the machine itself detects an event. I can set OSCAR to be more sensitive. Also, the problem with getting the oximeter to stay on my finger gets in the way.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I used to use a (defective) PR DreamStation Auto; I switched. |
Re: Recording pulse oximeters that stay on the finger well?
camper wrote: ↑Fri Jan 07, 2022 6:14 pmI'd like reviews of of inexpensive recording pulse oximeters that stay on the finger better, than finger clamp devices like the Contec CMS50DA+ I tried, even if I turn over while alseep.
In addition to staying on the finger, I prefer units that time-tag data, in a format that can be input into OSCAR.
Alternately, I would be interested in inexpensive units that also detect breathing.
Reasons:
I still want to find out if a wedge pillow and neck brace are as effective for me as my CPAP machine. (They are sufficient for me to stop snoring.) This is partly curiosity, partly I want to go back country camping. I'd rather not have to carry the machine and batteries.
Side Issue:
I sometimes breath deep while asleep, raising O2 blood levels, then stop breathing for a while. The machine initiates a Clear Airways event. The O2 levels usually don't drop below 92%. This is sometimes repeated a number of times. Can there be an adverse health effect from stopping breathing even if the O2 levels are reasonably high?
(OSCAR plot clarification: There is a time lag between breathing patterns and O2 levels. This might partly be a problem with the way I synced things: my pulse oximeter does not record time.)
I use a Contec CMS50F. No trouble syncing if you start CPAP and CMS50 at same time. No need to input anything, Oscar can read the CMS50 file after you import it into the supplied software. Only problem is that it records only 1 day at a time.
https://www.ebay.com/itm/143574800753?e ... BM_uWHuMdf
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Amara View Full Face CPAP Mask with Headgear |
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose |
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm
Amara View mask
O2 - 2-4 lpm
- SleepyCPAP
- Posts: 333
- Joined: Wed Dec 08, 2010 6:01 am
Re: Recording pulse oximeters that stay on the finger well?
I use the O2Ring, worn on my thumb. It stays on very well.
- SleepyCPAP
- SleepyCPAP
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Use OSCAR. Combine AlaxoStent with VAuto for perfect 0.0 AHI at PS 3.6 over 4cm EPAP |
-- SleepyCPAP
Sleep study in 2010 (11cm CPAP). Pillows (Swift FX>TAP PAP >Bleep). PRS1 “Pro” 450/460 until recall, now Aircurve 10 VAuto. Tape mouth. Palatal Prolapse solved by AlaxoStent & VAuto EPAP 4cm, PS 3.6cm = 0.0 AHI
Sleep study in 2010 (11cm CPAP). Pillows (Swift FX>TAP PAP >Bleep). PRS1 “Pro” 450/460 until recall, now Aircurve 10 VAuto. Tape mouth. Palatal Prolapse solved by AlaxoStent & VAuto EPAP 4cm, PS 3.6cm = 0.0 AHI
Re: Recording pulse oximeters that stay on the finger well?
Could you please explain how you set OSCAR to be more sensitive? I don't understand.
_________________
Machine: AirSense 11 Autoset |
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
Re: Recording pulse oximeters that stay on the finger well?
Edit Profile -> Select Profile
File -> Edit -> Preferences
CPAP -> Turn on Custom CPAP User Event Flagging
Note: This detects "User events", detected by OSCAR, not CPAP-machine-detected-events.
The menu explicitly says
For example user event type 1 might be 20% over 8 seconds , user event type 2 might be 50% over 8 seconds.Custom flagging is an experimental method of detecting events missed by the machine. They are NOT included in AHI.
BTW, if you hover over the fields you will see descriptions. E.g., it says the option must be enabled before importing data.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I used to use a (defective) PR DreamStation Auto; I switched. |
Last edited by camper on Fri Jan 21, 2022 12:43 pm, edited 1 time in total.
Re: Recording pulse oximeters that stay on the finger well?
Thanks for the info.SleepyCPAP wrote: ↑Sat Jan 08, 2022 5:16 pmI use the O2Ring, worn on my thumb. It stays on very well.
You "tape mouth"?- SleepyCPAP
Sleep study in 2010 (11cm CPAP). Adjusted to 10cm with pillows (Swift FX>TAP PAP >Bleep). PRS1 “Pro” 450/460 until recall, now Aircurve 10 VAuto. Tape mouth. Palatal Prolapse solved by AlaxoStent.
Is that common?
What happens if your nose gets congested? Could you die? I imagine someone waking, choking, and being too confused to remove the tape.
You can find lots of online info that questions the safety of this. E.g.,
https://www.forbes.com/sites/ninashapir ... 05683566e9
https://www.sleepfoundation.org/snoring ... -for-sleep
Both of those sites do say some people do it, and that some sleep specialists advocate it - but it would be scary to me.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I used to use a (defective) PR DreamStation Auto; I switched. |
- SleepyCPAP
- Posts: 333
- Joined: Wed Dec 08, 2010 6:01 am
Re: Recording pulse oximeters that stay on the finger well?
camper,
I’ve been taping, on and off, for years to control mouth leaks during CPAP. I started with the 3M Micropore tape, but found I would drool and it would lift. I’m now using Poligrip strips between my lips and KT Tape over the top. I fold a corner on one or both sides of the tape to grab to take it off.
I don’t claim this is safe for others, but I was glad to learn how it works for me.
I might find, at my new low pressures, that Poligrip would work by itself or combined with the weak 3M Micropore tape. I haven’t tried.
- SleepyCPAP
I’ve been taping, on and off, for years to control mouth leaks during CPAP. I started with the 3M Micropore tape, but found I would drool and it would lift. I’m now using Poligrip strips between my lips and KT Tape over the top. I fold a corner on one or both sides of the tape to grab to take it off.
I don’t claim this is safe for others, but I was glad to learn how it works for me.
I might find, at my new low pressures, that Poligrip would work by itself or combined with the weak 3M Micropore tape. I haven’t tried.
- SleepyCPAP
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Use OSCAR. Combine AlaxoStent with VAuto for perfect 0.0 AHI at PS 3.6 over 4cm EPAP |
-- SleepyCPAP
Sleep study in 2010 (11cm CPAP). Pillows (Swift FX>TAP PAP >Bleep). PRS1 “Pro” 450/460 until recall, now Aircurve 10 VAuto. Tape mouth. Palatal Prolapse solved by AlaxoStent & VAuto EPAP 4cm, PS 3.6cm = 0.0 AHI
Sleep study in 2010 (11cm CPAP). Pillows (Swift FX>TAP PAP >Bleep). PRS1 “Pro” 450/460 until recall, now Aircurve 10 VAuto. Tape mouth. Palatal Prolapse solved by AlaxoStent & VAuto EPAP 4cm, PS 3.6cm = 0.0 AHI
Re: Recording pulse oximeters that stay on the finger well?
Oh.SleepyCPAP wrote: ↑Sun Jan 23, 2022 10:36 amI’ve been taping, on and off, for years to control mouth leaks during CPAP.
I use a full face mask (Resmed F20) that completely covers my mouth and nose. It also creates a little tension that keeps my mouth closed.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I used to use a (defective) PR DreamStation Auto; I switched. |