Banned, I personally haven't seen a provider's manual for AVAPS. However, 15 BPM calculates to one complete respiratory cycle every 4 seconds (both inhale and exhale). Resmed VPAP manuals would recommend that half of those 4 seconds should normally be allotted to inspiration, for a Ti Max of 2 seconds. However, Resmed would have Ti Max slightly "short-cycled" or capped on an as-needed basis for COPD (to cope with possible "breath stacking" which can pose increasingly difficult expiratory work and thus undue cardiopulmonary distress for COPD patients).Banned wrote: Breath Rate (BPM): 15
Inspiratory Time (Ti): 1.8 (seconds)
Rise Time: 6 (0.6 second)
If Respironics' recommended inspiratory times concur with Resmed, then a 15 BPM setting just may warrant a slightly longer Ti than 1.8 seconds. However, you may have already methodically experimented with these parameters, only to discover that a 1.8 second inspiratory time is best for you. That 1.8 second setting (doubled for both I and E) works out to a natural or spontaneous BPM of roughly 16.67 (and latency backed-up at a slower 15 rate with your setting). And that may be just right for you based on your previous BiLevel trials.
Banned, since you're an engineer I'll toss out three ideas that you probably already considered:Banned wrote:The only downside to this particular copy of the AVAPS is when you turn the machine 'on' the motor slams against the inner housing. As long as it isn't going through the housing, it should be good to go. There is also no way to get the machine out of 'Provider mode' no matter when or how long you press the 'Silence' button.

If the inability to exit provider mode is a function of corrupt firmware or other control-circuitry malfunction (as opposed to a faulty switch), then the machine's entire functional integrity should be questioned. IMO the machine should be considered highly suspect based on the condition you described above. So please be very careful, sir...