AHI Computation

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Allessio77
Posts: 48
Joined: Sun Nov 26, 2017 11:38 am
Location: Arizona

AHI Computation

Post by Allessio77 » Thu Nov 30, 2017 12:50 pm

I was wondering if the AHI is computed on the time the machine is on, or
the time you are sleeping or the some other indicator.

I ask this since my Resmed seems to know the time i go to sleep. (SmartStart function does not turn on your pressures till it senses that you are asleep).
Does it also factor out any wake up times? Could it know that in the same manner as your sleep commencement?

If you read in bed with the machine on for any length of time, would that affect the AHI?

_________________
MachineMask
Additional Comments: Pressure 7-11, EPR = 1, Hum = 4, Hose Temp = 67

User avatar
Pugsy
Posts: 64922
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI Computation

Post by Pugsy » Thu Nov 30, 2017 1:02 pm

These machines have no clue if we are asleep or not.
SmartSense is only for breathing...has nothing to do with sleep. The machine senses a human is breathing on the other end and it turns itself on and has nothing to do at all with sleep vs wake.
Now the Auto ramp thing...maybe it can but that's just for ramp.
In general the AHI you see is base on an average for when the machine is turned on and then off. You could lay awake for 4 hours using the mask and machine and the machine still will factor those 4 hours into the calculations.

Sometimes awake breathing is irregular and these machines only measure air flow and sometimes our irregular awake breathing will trigger a flag from the machine. If you are reading and you hold your breath for 10 seconds (and we pause our breathing all the time and not realize it) the machine will likely give you a 10 second central apnea flag. So yes, reading in bed with mask on might impact the AHI a little...probably not a lot though.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
ChicagoGranny
Posts: 15076
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AHI Computation

Post by ChicagoGranny » Thu Nov 30, 2017 1:04 pm

AHI
Instead of using AHI, it's best to use Sleepyhead to view individual events, how long they are and when they happen.

User avatar
LSAT
Posts: 13316
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AHI Computation

Post by LSAT » Thu Nov 30, 2017 1:06 pm

Allessio77 wrote:I was wondering if the AHI is computed on the time the machine is on, or
the time you are sleeping or the some other indicator.

I ask this since my Resmed seems to know the time i go to sleep. (SmartStart function does not turn on your pressures till it senses that you are asleep).
Does it also factor out any wake up times? Could it know that in the same manner as your sleep commencement?

If you read in bed with the machine on for any length of time, would that affect the AHI?
The AHI is calculated by dividing the time the machine is on divided by the number of events. The machine does not know if you are sleeping or not. It just knows you are breathing. Example....If you have the machine running for 8 hours and you were reading for 2 of those hours, with 16 events the AHI calculation would be 2.0. In reality you were sleeping for only 6 hours and the true AHI is 2.66.

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: AHI Computation

Post by TASmart » Thu Nov 30, 2017 1:09 pm

I just wanted to add that the PA at my sleep Dr. told me that the hypopnea identification is not that great on xPAP's. Her advice was to take the hypopnea number with a grain of salt.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
ChicagoGranny
Posts: 15076
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AHI Computation

Post by ChicagoGranny » Thu Nov 30, 2017 1:26 pm

LSAT wrote:It just knows you are breathing.
... or not breathing:

Image


User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: AHI Computation

Post by Jay Aitchsee » Thu Nov 30, 2017 3:42 pm

TASmart wrote:I just wanted to add that the PA at my sleep Dr. told me that the hypopnea identification is not that great on xPAP's. Her advice was to take the hypopnea number with a grain of salt.
ResMed S8's notoriously over counted hypopneas. I haven't heard the same about S9's and later.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video