holes in my wisp white tubing
holes in my wisp white tubing
I have been a CPAP user for several years and have 95% compliance. I know that I sleep much better when I use my machine. Now for the question...The white tubing that connects to my wisp nasal mask has developed holes (for the second time). I have no animals and I can't figure out what is causing the problem. I have had to replace hose twice in 6 months. My pressure is 95% of the time below 15 and I can top out at 16. Is this too much pressure for this kind of mask? Does anyone have any ideas as to why this is happening? Thanks to Pugsy for showing me how to upload the picture
_________________
Machine: AirSense 11 Autoset |
Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame) |
Re: holes in my wisp white tubing
what are you cleaning it with?yulcdn wrote:I have been a CPAP user for several years and have 95% compliance. I know that I sleep much better when I use my machine. Now for the question...The white tubing that connects to my wisp nasal mask has developed holes (for the second time). I have no animals and I can't figure out what is causing the problem. I have had to replace hose twice in 6 months. My pressure is 95% of the time below 15 and I can top out at 16. Is this too much pressure for this kind of mask? Does anyone have any ideas as to why this is happening?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: holes in my wisp white tubing
I wash it with mild/ natural dishwashing soap. It is the same soap that I use for the long tubing, the mask and the water reservoir. This short tubing is the only one that develops holes.
_________________
Machine: AirSense 11 Autoset |
Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame) |
Re: holes in my wisp white tubing
I'd stop. I don't know what your 'mild/natural dishwashing soap' is, or even if that's what your tubing doesn't like, but perhaps you'd be better off just using warm water and nothing else.yulcdn wrote:I wash it with mild/ natural dishwashing soap. It is the same soap that I use for the long tubing, the mask and the water reservoir. This short tubing is the only one that develops holes.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: holes in my wisp white tubing
And maybe it's just defective crap.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
Re: holes in my wisp white tubing
Maybe explain more about your washing technique and any tools you use. For example, Do you use a brush?yulcdn wrote:I wash it with mild/ natural dishwashing soap. It is the same soap that I use for the long tubing, the mask and the water reservoir. This short tubing is the only one that develops holes.
Re: holes in my wisp white tubing
from the looks of it, I'm thinking half inch drill bit...Guest wrote:Maybe explain more about your washing technique and any tools you use. For example, Do you use a brush?yulcdn wrote:I wash it with mild/ natural dishwashing soap. It is the same soap that I use for the long tubing, the mask and the water reservoir. This short tubing is the only one that develops holes.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 2744
- Joined: Tue Oct 12, 2010 6:42 pm
Re: holes in my wisp white tubing
That looks very familiar. When I was using the Swift FX I had similar holes in the short tube.
After going through several hoses I finally did some testing. It appears that the hose material doesn't handle stretching. If you limit the amount the hose can expand along its length, the problem should go away.
I am an active sleeper and the short hose would constantly be stretched before the larger hose. This resulted in holes that were similar to what you are showing.
After going through several hoses I finally did some testing. It appears that the hose material doesn't handle stretching. If you limit the amount the hose can expand along its length, the problem should go away.
I am an active sleeper and the short hose would constantly be stretched before the larger hose. This resulted in holes that were similar to what you are showing.
_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier. |
SpO2 96+% and holding...
Re: holes in my wisp white tubing
Perhaps you are right about the stretching...I know that is not from the washing of the tubing. During the night I often tug on the tubing when I roll over. The real question is how can I limit the stretching of the short tubing while I am sleeping.
_________________
Machine: AirSense 11 Autoset |
Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame) |
- chunkyfrog
- Posts: 34424
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: holes in my wisp white tubing
Hose management.
Hang it from one of the many inventive devices created for that purpose or "rube goldberg" your own.
I used duck tape and scrunchies on a trip, and some drape their hose over tbe headboard.
Hang it from one of the many inventive devices created for that purpose or "rube goldberg" your own.
I used duck tape and scrunchies on a trip, and some drape their hose over tbe headboard.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- Okie bipap
- Posts: 3553
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: holes in my wisp white tubing
Perhaps you could try a short hose cover such as Karen at padacheek makes. You can snug up the ends of the cover with velcro and that may keep you from stretching the hose so much.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
- chunkyfrog
- Posts: 34424
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: holes in my wisp white tubing
Has this happened on more than one Wisp?
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- MaxDarkside
- Posts: 1199
- Joined: Sun Dec 18, 2011 4:21 pm
- Location: Minneapolis, MN
Re: holes in my wisp white tubing
Looks to me to be a bad tube design and materials. I'd change manufacturer.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead |
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs
- chunkyfrog
- Posts: 34424
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: holes in my wisp white tubing
I used the wisp for a couple of years, accumulating more than one.
For me, only the headgear cratered, and the fabric front frame was garbage.
I am looking at three, yes 3, front hoses that look like new.
Make me an offer. Via PM.
I have a box, containing the hoses, 3 used, and 1 unopened silicone front frames,
one used headgear, and six s/m cushions, still sealed in original packaging.
All or part--I'm happy either way.
For me, only the headgear cratered, and the fabric front frame was garbage.
I am looking at three, yes 3, front hoses that look like new.
Make me an offer. Via PM.
I have a box, containing the hoses, 3 used, and 1 unopened silicone front frames,
one used headgear, and six s/m cushions, still sealed in original packaging.
All or part--I'm happy either way.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- Michandola
- Posts: 20
- Joined: Wed Mar 08, 2017 2:25 pm
- Location: Tampa, FL
Re: holes in my wisp white tubing
How are you detaching the hose from the tubing that connects to cpap?yulcdn wrote:I have been a CPAP user for several years and have 95% compliance. I know that I sleep much better when I use my machine. Now for the question...The white tubing that connects to my wisp nasal mask has developed holes (for the second time). I have no animals and I can't figure out what is causing the problem. I have had to replace hose twice in 6 months. My pressure is 95% of the time below 15 and I can top out at 16. Is this too much pressure for this kind of mask? Does anyone have any ideas as to why this is happening? Thanks to Pugsy for showing me how to upload the picture
Mit - DreamMachine