holes in my wisp white tubing

Post by yulcdn » Sun Apr 09, 2017 4:05 pm

I have been a CPAP user for several years and have 95% compliance. I know that I sleep much better when I use my machine. Now for the question...The white tubing that connects to my wisp nasal mask has developed holes (for the second time). I have no animals and I can't figure out what is causing the problem. I have had to replace hose twice in 6 months. My pressure is 95% of the time below 15 and I can top out at 16. Is this too much pressure for this kind of mask? Does anyone have any ideas as to why this is happening? Thanks to Pugsy for showing me how to upload the picture

Image

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: holes in my wisp white tubing

Post by palerider » Sun Apr 09, 2017 4:09 pm

yulcdn wrote:I have been a CPAP user for several years and have 95% compliance. I know that I sleep much better when I use my machine. Now for the question...The white tubing that connects to my wisp nasal mask has developed holes (for the second time). I have no animals and I can't figure out what is causing the problem. I have had to replace hose twice in 6 months. My pressure is 95% of the time below 15 and I can top out at 16. Is this too much pressure for this kind of mask? Does anyone have any ideas as to why this is happening?
what are you cleaning it with?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
yulcdn
Posts: 65
Joined: Mon Mar 19, 2012 10:35 pm
Location: Southern California

Re: holes in my wisp white tubing

Post by yulcdn » Sun Apr 09, 2017 4:23 pm

I wash it with mild/ natural dishwashing soap. It is the same soap that I use for the long tubing, the mask and the water reservoir. This short tubing is the only one that develops holes.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: holes in my wisp white tubing

Post by palerider » Sun Apr 09, 2017 4:37 pm

yulcdn wrote:I wash it with mild/ natural dishwashing soap. It is the same soap that I use for the long tubing, the mask and the water reservoir. This short tubing is the only one that develops holes.
I'd stop. I don't know what your 'mild/natural dishwashing soap' is, or even if that's what your tubing doesn't like, but perhaps you'd be better off just using warm water and nothing else.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Julie
Posts: 19931
Joined: Tue Feb 28, 2006 12:58 pm

Re: holes in my wisp white tubing

Post by Julie » Sun Apr 09, 2017 5:09 pm

And maybe it's just defective crap.

Guest

Re: holes in my wisp white tubing

Post by Guest » Sun Apr 09, 2017 5:33 pm

yulcdn wrote:I wash it with mild/ natural dishwashing soap. It is the same soap that I use for the long tubing, the mask and the water reservoir. This short tubing is the only one that develops holes.
Maybe explain more about your washing technique and any tools you use. For example, Do you use a brush?

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: holes in my wisp white tubing

Post by palerider » Sun Apr 09, 2017 6:21 pm

Guest wrote:
yulcdn wrote:I wash it with mild/ natural dishwashing soap. It is the same soap that I use for the long tubing, the mask and the water reservoir. This short tubing is the only one that develops holes.
Maybe explain more about your washing technique and any tools you use. For example, Do you use a brush?
from the looks of it, I'm thinking half inch drill bit...

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: holes in my wisp white tubing

Post by HoseCrusher » Sun Apr 09, 2017 9:55 pm

That looks very familiar. When I was using the Swift FX I had similar holes in the short tube.

After going through several hoses I finally did some testing. It appears that the hose material doesn't handle stretching. If you limit the amount the hose can expand along its length, the problem should go away.

I am an active sleeper and the short hose would constantly be stretched before the larger hose. This resulted in holes that were similar to what you are showing.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
yulcdn
Posts: 65
Joined: Mon Mar 19, 2012 10:35 pm
Location: Southern California

Re: holes in my wisp white tubing

Post by yulcdn » Sun Apr 09, 2017 10:43 pm

Perhaps you are right about the stretching...I know that is not from the washing of the tubing. During the night I often tug on the tubing when I roll over. The real question is how can I limit the stretching of the short tubing while I am sleeping.

User avatar
chunkyfrog
Posts: 34452
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: holes in my wisp white tubing

Post by chunkyfrog » Mon Apr 10, 2017 10:32 am

Hose management.
Hang it from one of the many inventive devices created for that purpose or "rube goldberg" your own.
I used duck tape and scrunchies on a trip, and some drape their hose over tbe headboard.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Okie bipap
Posts: 3556
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: holes in my wisp white tubing

Post by Okie bipap » Mon Apr 10, 2017 12:18 pm

Perhaps you could try a short hose cover such as Karen at padacheek makes. You can snug up the ends of the cover with velcro and that may keep you from stretching the hose so much.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
chunkyfrog
Posts: 34452
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: holes in my wisp white tubing

Post by chunkyfrog » Mon Apr 10, 2017 1:44 pm

Has this happened on more than one Wisp?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: holes in my wisp white tubing

Post by MaxDarkside » Mon Apr 10, 2017 4:39 pm

Looks to me to be a bad tube design and materials. I'd change manufacturer.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

User avatar
chunkyfrog
Posts: 34452
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: holes in my wisp white tubing

Post by chunkyfrog » Mon Apr 10, 2017 5:37 pm

I used the wisp for a couple of years, accumulating more than one.
For me, only the headgear cratered, and the fabric front frame was garbage.
I am looking at three, yes 3, front hoses that look like new.
Make me an offer. Via PM.
I have a box, containing the hoses, 3 used, and 1 unopened silicone front frames,
one used headgear, and six s/m cushions, still sealed in original packaging.
All or part--I'm happy either way.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Michandola
Posts: 20
Joined: Wed Mar 08, 2017 2:25 pm
Location: Tampa, FL

Re: holes in my wisp white tubing

Post by Michandola » Mon Apr 10, 2017 6:47 pm

yulcdn wrote:I have been a CPAP user for several years and have 95% compliance. I know that I sleep much better when I use my machine. Now for the question...The white tubing that connects to my wisp nasal mask has developed holes (for the second time). I have no animals and I can't figure out what is causing the problem. I have had to replace hose twice in 6 months. My pressure is 95% of the time below 15 and I can top out at 16. Is this too much pressure for this kind of mask? Does anyone have any ideas as to why this is happening? Thanks to Pugsy for showing me how to upload the picture

Image
How are you detaching the hose from the tubing that connects to cpap?

_________________
MachineMaskHumidifier
Mit - DreamMachine