I am not sure if my humidifier is working correctly on my Dreamstation, I have it set at 5 and the amount of water it uses doesn’t change , I recently just spent a week with a friend of mine in Arizona who has the same machine and has his set at 5 also , in the morning his chamber was almost empty and mine was half full, have any other people run into this ?
Sent from my iPhone using Tapatalk
Humidifier chamber
Re: Humidifier chamber
If it isn't using water in Arizona then something is probably up.
Turn it on and let it run for about 15 minutes and then turn it off and immediately remove the water chamber and check the bottom of the chamber for warmth. It should be at least warm to the touch. Not necessarily horribly hot but good and warm.
If it's not...you have a problem and it needs to go back to where you got it.
I don't know if the DreamStation offers Classic Mode or not...the prior model System One did.
If it does (need to be in the clinical setup menu area to changed humidification modes) then change it to Classic Mode at the setting of 5 and see what happens. If it doesn't use much water in Classic mode...most definitely there is a problem somewhere in the unit.
Might be in the humidifier but also might be in the blower unit and a connection problem in the blower.
If you don't know about clinical setup mode and how to get there go here and request the manual via email and it's all explained in the manual.
https://www.apneaboard.com/adjust-cpap- ... tup-manual
Turn it on and let it run for about 15 minutes and then turn it off and immediately remove the water chamber and check the bottom of the chamber for warmth. It should be at least warm to the touch. Not necessarily horribly hot but good and warm.
If it's not...you have a problem and it needs to go back to where you got it.
I don't know if the DreamStation offers Classic Mode or not...the prior model System One did.
If it does (need to be in the clinical setup menu area to changed humidification modes) then change it to Classic Mode at the setting of 5 and see what happens. If it doesn't use much water in Classic mode...most definitely there is a problem somewhere in the unit.
Might be in the humidifier but also might be in the blower unit and a connection problem in the blower.
If you don't know about clinical setup mode and how to get there go here and request the manual via email and it's all explained in the manual.
https://www.apneaboard.com/adjust-cpap- ... tup-manual
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Humidifier chamber
Thanks I will try this .Pugsy wrote:If it isn't using water in Arizona then something is probably up.
Turn it on and let it run for about 15 minutes and then turn it off and immediately remove the water chamber and check the bottom of the chamber for warmth. It should be at least warm to the touch. Not necessarily horribly hot but good and warm.
If it's not...you have a problem and it needs to go back to where you got it.
I don't know if the DreamStation offers Classic Mode or not...the prior model System One did.
If it does (need to be in the clinical setup menu area to changed humidification modes) then change it to Classic Mode at the setting of 5 and see what happens. If it doesn't use much water in Classic mode...most definitely there is a problem somewhere in the unit.
Might be in the humidifier but also might be in the blower unit and a connection problem in the blower.
If you don't know about clinical setup mode and how to get there go here and request the manual via email and it's all explained in the manual.
https://www.apneaboard.com/adjust-cpap- ... tup-manual
Sent from my iPhone using Tapatalk
Re: Humidifier chamber
You didn't say that it is using almost no water, just that it is using less than you think it should, or less than somebody else's does under what appears to be the same settings and conditions.
Were you actually sleeping in the same room (not just the same building)? If not, the conditions (especially the temperature) might not be the same.
Also, other things can affect how much water you use such as your pressure setting, what type of mask you're using, your leak rate (the DreamStation compensates for leaks), and several other variables.
Were you actually sleeping in the same room (not just the same building)? If not, the conditions (especially the temperature) might not be the same.
Also, other things can affect how much water you use such as your pressure setting, what type of mask you're using, your leak rate (the DreamStation compensates for leaks), and several other variables.
- chunkyfrog
- Posts: 34452
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Humidifier chamber
Just check if the metal bottom of the tank gets warm.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Humidifier chamber
Thanks good info ! This is probably the reason .D.H. wrote:You didn't say that it is using almost no water, just that it is using less than you think it should, or less than somebody else's does under what appears to be the same settings and conditions.
Were you actually sleeping in the same room (not just the same building)? If not, the conditions (especially the temperature) might not be the same.
Also, other things can affect how much water you use such as your pressure setting, what type of mask you're using, your leak rate (the DreamStation compensates for leaks), and several other variables.
Sent from my iPhone using Tapatalk