zonker
- Posts: 11048
- Joined: Fri Jun 19, 2015 4:36 pm
Post
by zonker » Mon Apr 22, 2019 5:22 pm
Usman wrote: ↑Mon Apr 22, 2019 8:08 am
do i have to install any software on my laptop for this purpose ?
and what is the use of bluetooth on dreamstation
please click on the last link at the bottom of this post to download the sleepyhead software.
the bluetooth is there to get the data generated by the machine to the doctor remotely. most of us turn it off because we don't want any doctors making changes to our data remotely.
good luck!
-
Chris8243
- Posts: 419
- Joined: Sun Apr 08, 2018 10:42 pm
- Location: Wyoming USA
Post
by Chris8243 » Mon Apr 22, 2019 7:24 pm
zonker wrote: ↑Mon Apr 22, 2019 5:22 pm
Usman wrote: ↑Mon Apr 22, 2019 8:08 am
do i have to install any software on my laptop for this purpose ?
and what is the use of bluetooth on dreamstation
please click on the last link at the bottom of this post to download the sleepyhead software.
the bluetooth is there to get the data generated by the machine to the doctor remotely. most of us turn it off because we don't want any doctors making changes to our data remotely.
good luck!
Zonker,
I thought the modem was for the DME or doctor to see results & the bluetooth was just for the app. Am I wrong?
-
zonker
- Posts: 11048
- Joined: Fri Jun 19, 2015 4:36 pm
Post
by zonker » Mon Apr 22, 2019 7:38 pm
Chris8243 wrote: ↑Mon Apr 22, 2019 7:24 pm
zonker wrote: ↑Mon Apr 22, 2019 5:22 pm
Usman wrote: ↑Mon Apr 22, 2019 8:08 am
do i have to install any software on my laptop for this purpose ?
and what is the use of bluetooth on dreamstation
please click on the last link at the bottom of this post to download the sleepyhead software.
the bluetooth is there to get the data generated by the machine to the doctor remotely. most of us turn it off because we don't want any doctors making changes to our data remotely.
good luck!
Zonker,
I thought the modem was for the DME or doctor to see results & the bluetooth was just for the app. Am I wrong?
<sigh> it's been a long day and i'm not sure that's what i meant to type. i think i was thinking of some of the past conspiracy theories brought up and that's what i typed.
you are right, it's for apps.
at about the same time, i just pmed pugsy and palerider to look at some charts, but then linked to the wtong post.
i may have to go lay down for awhile.
-
palerider
- Posts: 32300
- Joined: Wed Dec 16, 2009 5:43 pm
- Location: Dallas(ish).
Post
by palerider » Mon Apr 22, 2019 7:42 pm
Chris8243 wrote: ↑Mon Apr 22, 2019 7:24 pm
I thought the modem was for the DME or doctor to see results & the bluetooth was just for the app. Am I wrong?
And you're going to trust them to tell you what's really going on?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
Chris8243
- Posts: 419
- Joined: Sun Apr 08, 2018 10:42 pm
- Location: Wyoming USA
Post
by Chris8243 » Mon Apr 22, 2019 8:48 pm
palerider wrote: ↑Mon Apr 22, 2019 7:42 pm
Chris8243 wrote: ↑Mon Apr 22, 2019 7:24 pm
I thought the modem was for the DME or doctor to see results & the bluetooth was just for the app. Am I wrong?
And you're going to trust them to tell you what's really going on?
Ahhhh, I do like my doctor. Last time I saw him, he said "I see you've changed your pressure. Good job. I would have done the same." I asked if he'd heard of Sleepyhead, he did, and I told him how I learned of it on this forum, blah blah, blah.
I turned the Bluetooth off on my machine long ago, the app is meaningless. The modem can only be turned off for 3 days at a time (it's set to come back on after 3 days.) I would remove it or turn it off permanently but apparently not an option.
-
Usman
- Posts: 15
- Joined: Sun Apr 21, 2019 5:54 am
Post
by Usman » Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
-
palerider
- Posts: 32300
- Joined: Wed Dec 16, 2009 5:43 pm
- Location: Dallas(ish).
Post
by palerider » Tue Apr 23, 2019 12:33 pm
Usman wrote: ↑Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
There isn't one.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
Dan_McD
- Posts: 146
- Joined: Sat Dec 22, 2018 1:15 pm
Post
by Dan_McD » Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.
"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
-
palerider
- Posts: 32300
- Joined: Wed Dec 16, 2009 5:43 pm
- Location: Dallas(ish).
Post
by palerider » Tue Apr 23, 2019 1:24 pm
Dan_McD wrote: ↑Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.
"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
That won't tell you anything at all about your leaks during the night.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
Pugsy
- Posts: 64156
- Joined: Thu May 14, 2009 9:31 am
- Location: Missouri, USA
Post
by Pugsy » Tue Apr 23, 2019 1:46 pm
There is some sort of leak number that gets shown with the "Show AHI" turned on in the clinical setup menu area....so that people can get the limited data that the machine's LCD screen shows.
It's all explained in the clinical manual...that I guess someone hasn't bothered to get and/or read yet.
I forget off the top of my head which leak number it gives though....and no time to go look it up when people need to get their own manuals and do their own reading.
I provided the link...they have to do their own part of the work.
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
-
Dan_McD
- Posts: 146
- Joined: Sat Dec 22, 2018 1:15 pm
Post
by Dan_McD » Tue Apr 23, 2019 2:03 pm
palerider wrote: ↑Tue Apr 23, 2019 1:24 pm
Dan_McD wrote: ↑Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.
"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
That won't tell you anything at all about your leaks during the night.
*
What will tell you "about your leaks during the night"?
-
Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Post
by Jas_williams » Tue Apr 23, 2019 2:28 pm
Dan_McD wrote: ↑Tue Apr 23, 2019 2:03 pm
palerider wrote: ↑Tue Apr 23, 2019 1:24 pm
Dan_McD wrote: ↑Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.
"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
That won't tell you anything at all about your leaks during the night.
*
What will tell you "about your leaks during the night"?
Sleepyhead
-
palerider
- Posts: 32300
- Joined: Wed Dec 16, 2009 5:43 pm
- Location: Dallas(ish).
Post
by palerider » Tue Apr 23, 2019 2:30 pm
Dan_McD wrote: ↑Tue Apr 23, 2019 2:03 pm
palerider wrote: ↑Tue Apr 23, 2019 1:24 pm
Dan_McD wrote: ↑Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.
"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
That won't tell you anything at all about your leaks during the night.
*
What will tell you "about your leaks during the night"?
Looking at the data that's on the SD card.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
Chris8243
- Posts: 419
- Joined: Sun Apr 08, 2018 10:42 pm
- Location: Wyoming USA
Post
by Chris8243 » Tue Apr 23, 2019 2:44 pm
Pugsy wrote: ↑Tue Apr 23, 2019 1:46 pm
There is some sort of leak number that gets shown with the "Show AHI" turned on in the clinical setup menu area....so that people can get the limited data that the machine's LCD screen shows.
It's all explained in the clinical manual...that I guess someone hasn't bothered to get and/or read yet.
I forget off the top of my head which leak number it gives though....and no time to go look it up when people need to get their own manuals and do their own reading.
I provided the link...they have to do their own part of the work.
The LCD screen shows "Mask Fit""as a percentage for one day, 7 days, 30 days. Not particularly helpful as it is after the fact. There ia a "Check Mask Fit" on the clinical menu which runs for 40 seconds. I don't use it, I guess because I can tell immediately if the mask is leaking as soon as I turn the machine on.
-
Dan_McD
- Posts: 146
- Joined: Sat Dec 22, 2018 1:15 pm
Post
by Dan_McD » Tue Apr 23, 2019 2:55 pm
palerider wrote: ↑Tue Apr 23, 2019 2:30 pm
Dan_McD wrote: ↑Tue Apr 23, 2019 2:03 pm
palerider wrote: ↑Tue Apr 23, 2019 1:24 pm
Dan_McD wrote: ↑Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.
"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
That won't tell you anything at all about your leaks during the night.
*
What will tell you "about your leaks during the night"?
Looking at the data that's on the SD card.
That's kinda what I thought, to check the usage, however long it is in use, on sleepyhead "Leak Rate". I thought the OP asked "is there any way to check mask leakage?" Dreamstation has that "Mask Fit" option. They quote "This feature allows the patient to check the fit of their mask
prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit".
You are right, at night if you knock your mask loose or pull the hose off, the "Mask Fit" feature is of no use.