Philips Respironics REMstar Auto A Flex
Philips Respironics REMstar Auto A Flex
I'm having a problem with the air pressure. I have no air pressure when I put on the nose mask. However, when the mask is off, the pressure is strong flowing. It's the reverse of what I want it to do. Any suggestions? I have the Philips Respironics REMstar Auto A Flex model.
Re: Philips Respironics REMstar Auto A Flex
what happens when you take the mask off is totally meaningless, the machine is programmed to create a specific pressure, it does that by creating enough airflow to raise the pressure against the built in leak rate of the mask. when you take the mask off, there's no resistance, so the pressure drops to zero, and the machine ramps up to full speed, trying to maintain the pressure. when you put the mask on, the blower slows *way* down.sirguy77 wrote:I'm having a problem with the air pressure. I have no air pressure when I put on the nose mask. However, when the mask is off, the pressure is strong flowing. It's the reverse of what I want it to do. Any suggestions? I have the Philips Respironics REMstar Auto A Flex model.
you need to figure out *which* of the *many* Philips Respironics REMstar Auto A Flex model you have, and if it has an SD card, get some software, like sleepyhead, to read the data stored on it.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 1562
- Joined: Tue May 06, 2008 5:57 pm
- Location: Murrysville, PA
Re: Philips Respironics REMstar Auto A Flex
Good advise above. Your machine may be set to default pressures starting at 4 cmH2O. It would help if you could check what your settings are. A low pressure setting can feel like you're not getting enough air. Increasing the minimum pressure and/or turning off the ramp function can quickly solve this problem.
Here is a link on how to access those settings. Please record the existing settings before making changes. http://www.apneaboard.com/pr-system-one ... structions
Here is a link on how to access those settings. Please record the existing settings before making changes. http://www.apneaboard.com/pr-system-one ... structions
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS |