Machine runs out of water.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Mr.Mar10s
Posts: 10
Joined: Tue Aug 30, 2016 7:16 pm

Machine runs out of water.

Post by Mr.Mar10s » Tue Sep 13, 2016 1:49 am

First night using CPAP in months. Ran out of water in about 5 hours. New hose. New mask. Part of the reason I stopped. Can't last all night. Feels pointless. What would you guys recommend for making the water last? Thanks for the help. This community is giving me the courage and wisdom to get back on treatment. It honesty has always felt overwhelming to try and figure out and it seems like Drs can't tell you much bout sleep apnea treatment except go get a sleep test again. This community is the best. Thank you again for the help.
Phillips Respironics system one. Ref560p. Full face mask.

Mr.Mar10s
Posts: 10
Joined: Tue Aug 30, 2016 7:16 pm

Re: Machine runs out of water.

Post by Mr.Mar10s » Tue Sep 13, 2016 1:51 am

Btw it's a Phillips Respironics system one I believe.
Phillips Respironics system one. Ref560p. Full face mask.

Guest

Re: Machine runs out of water.

Post by Guest » Tue Sep 13, 2016 2:40 am

The most likely cause of excess water use is a leak. The more air flowing over the water the more water that will be wicked away. Download Sleepyhead and check for leaks during the nite. Search this forum for more info.

User avatar
LSAT
Posts: 13316
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Machine runs out of water.

Post by LSAT » Tue Sep 13, 2016 5:28 am

Several possible reasons for high water use....(1) You have the humidifier set too high...(2) If you have Air Conditioning /heat running it is dry in your house and more water will be used (don't know where you are located)...(3) You are mouth breathing which acts like a leak.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Machine runs out of water.

Post by JDS74 » Tue Sep 13, 2016 6:26 am

Welcome to the forum.
A couple of bookkeeping things first.

Could you go to the User Control Panel (upper left) and post your equipment? You can get the model number from the bottom of the machine (blower unit not the humidifier part). Check the display on this forum by clicking Preview to make sure your machine is displaying properly. If it is an older machine, it may not display properly and then you will need to just type in the description in the signature box.

Do you have a heated hose - it has electrical connections where it plugs into the humidifier.

finally, it will be helpful to you to get a copy of all sleep studies so that you can have available to you the answers to any more questions from forum members.

End of Bookkeeping.

Is your mask a nasal a mask? If so, and you are mouth breathing, then excessive water use might result.
If your mask is a full face mask, then mouth breathing is normal and does use more water.

Is your mouth dry when you wake up in the night? That would be an indicator of mouth breathing.
So, some things to try:
Get a soft cervical collar - Walmart, CVS, etc. and use it. The collar will support your jaw and stop your mouth from falling open when your facial muscles relax with sleep. When using a full face mask, having the jaw drop, even just a little, can cause mask leaks and cause the problem you are having.

Check to see the level of humidity that your machine is providing. It may be that you can lower it a little and solve this particular problem.

Get a copy of SleepyHead software and begin to track your therapy. This is only helpful if your machine does collect the data needed. Not all CPAP machines do collect the data and that is the reason that filling out the equipment list in the User Control Panel is helpful.

If you have a full data machine, and these other suggestions don't solve the problem, there are other methods that can help. But first things first, try the simple ones and maybe all will be well.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Pugsy
Posts: 64922
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Machine runs out of water.

Post by Pugsy » Tue Sep 13, 2016 8:27 am

Which model System One machine?
Look on the bottom of the blower unit for a 3 digit number which may have DS or REF in front or behind it.
Tell us those 3 digits.
It's important to know because you want check for large leaks and not all machines will gather leak data.

And does your hose have a black coupling on the end that goes to the humidifier lid? This is important to know.

What humidifier setting are you using?
What pressure are you using?

It shouldn't use all the water in 5 hours unless something is going on and the first suspect would be large leak.
Even at the max setting in the driest of environments it should last the biggest part of the night.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

D.H.
Posts: 3522
Joined: Sun Sep 04, 2016 7:07 pm

Re: Machine runs out of water.

Post by D.H. » Tue Sep 13, 2016 9:10 am

I have this problem with the Phillips Dream Station. For the record, my pressure is 13½ , and I have the humidity on high. Since I get up at least once during the night to use the toilet, I've learned to refill it at that time.

If you can otherwise sleep through the night, or doing this would severely disrupt your sleep, you can try a lower humidity setting.

The heated hose makes it possible to deliver more humidity. Thus, if the tank has the same capacity as the older machines, it stands to reason that the tank would empty sooner.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max
Last edited by D.H. on Mon Feb 06, 2017 10:31 pm, edited 1 time in total.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Machine runs out of water.

Post by chunkyfrog » Tue Sep 13, 2016 9:19 am

My water lasts from 7-10 hours, depending on the humidity in the room.
Get a cheapie hygrometer, and observe the correlation yourself.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Mr.Mar10s
Posts: 10
Joined: Tue Aug 30, 2016 7:16 pm

Re: Machine runs out of water.

Post by Mr.Mar10s » Tue Sep 13, 2016 5:52 pm

Pugsy wrote:Which model System One machine?
Look on the bottom of the blower unit for a 3 digit number which may have DS or REF in front or behind it.
Tell us those 3 digits.
It's important to know because you want check for large leaks and not all machines will gather leak data.

And does your hose have a black coupling on the end that goes to the humidifier lid? This is important to know.

What humidifier setting are you using?
What pressure are you using?

It shouldn't use all the water in 5 hours unless something is going on and the first suspect would be large leak.
Even at the max setting in the driest of environments it should last the biggest part of the night.

ref 560p. Yes it has a black coupling. It was on humidity 3 I just changed it to and I usually you the 5 setting. Thank you everyone.
Phillips Respironics system one. Ref560p. Full face mask.

User avatar
Bill44133
Posts: 1087
Joined: Tue Jul 09, 2013 8:34 pm
Location: North Royalton, OH

Re: Machine runs out of water.

Post by Bill44133 » Tue Sep 13, 2016 6:29 pm

I have the same problem.. My Vicks cold air humidifier decided it was going to work any more. Without the humidifier running in the bedroom my machine water is emptied after 4 to 5 hours, and I wake up all congested its terrible.

I like to keep my setting for humidity to 4 on my 760 machine. I have found changing the setting to 3 gets me closer to 7 hours of sleep before the water runs dry.

I would recommend running a room humidifier in your bedroom to not run out of water in your pap machine. If you like the extra humidity the machine offers but runs out of water.

I wish you well..

_________________
Machine: DreamStation BiPAP® Auto Machine
Humidifier: DreamStation Heated Humidifier
Additional Comments: Settings are IPap 23 EPap 19

Isnore2

Re: Machine runs out of water.

Post by Isnore2 » Fri Feb 03, 2017 8:53 pm

I have the Philips DreamStation and it consistantly is out of water when I wake up after 6 hours of use. the supplier gave me a new humidifier after my first week of use. The same problem still exists. It does have the heated hose connection. I'm running at 2 humidity. It's out of water whether I have the preheat on or not. My wife wakes up when it runs out of water because it makes a gurgling noise. I'm considering asking my insurance if there's another brand they'll cover. Having dry air blown into my nose & throat is painful. It is dry in the high desert, but that 's all the reason for a more humid air supply. And I'm not a mouth breather. 2/3/2017.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Machine runs out of water.

Post by Wulfman... » Fri Feb 03, 2017 9:02 pm

Isnore2 wrote:I have the Philips DreamStation and it consistantly is out of water when I wake up after 6 hours of use. the supplier gave me a new humidifier after my first week of use. The same problem still exists. It does have the heated hose connection. I'm running at 2 humidity. It's out of water whether I have the preheat on or not. My wife wakes up when it runs out of water because it makes a gurgling noise. I'm considering asking my insurance if there's another brand they'll cover. Having dry air blown into my nose & throat is painful. It is dry in the high desert, but that 's all the reason for a more humid air supply. And I'm not a mouth breather. 2/3/2017.
Have you checked your software reports for large leaks? You ARE using software, aren't you? Excessive leakage can cause more water being expelled in various ways. If you use a nasal mask, it could be coming out your mouth (even though you claim not to be a mouth-breather). Mouth-breathing and mouth-leaking are different problems from the same source.

I also live in a high, dry altitude and use "passover" humidification......no heat but water in the tank, year-round......and have no problems with that.

Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

D.H.
Posts: 3522
Joined: Sun Sep 04, 2016 7:07 pm

Re: Machine runs out of water.

Post by D.H. » Fri Feb 03, 2017 10:08 pm

Wulfman... wrote:
Isnore2 wrote:I have the Philips DreamStation and it consistantly is out of water when I wake up after 6 hours of use. the supplier gave me a new humidifier after my first week of use. The same problem still exists. It does have the heated hose connection. I'm running at 2 humidity. It's out of water whether I have the preheat on or not. My wife wakes up when it runs out of water because it makes a gurgling noise. I'm considering asking my insurance if there's another brand they'll cover. Having dry air blown into my nose & throat is painful. It is dry in the high desert, but that 's all the reason for a more humid air supply. And I'm not a mouth breather. 2/3/2017.
Have you checked your software reports for large leaks? You ARE using software, aren't you? Excessive leakage can cause more water being expelled in various ways. If you use a nasal mask, it could be coming out your mouth (even though you claim not to be a mouth-breather). Mouth-breathing and mouth-leaking are different problems from the same source.

I also live in a high, dry altitude and use "passover" humidification......no heat but water in the tank, year-round......and have no problems with that.

Den

.
As I previously posted, my DreamStation also runs out of water before morning if I don't refill during the night. My pressure is 13.5 min/20 max (it has never gone above 17 and it rarely goes above 15.5). This is a change in my setting from my last posting to this thread in September. I set the humidity on max and use a heated hose. Sleepyhead does report significant leakage, but Dreammapper (the software provided by Phillips) report no significant leakage. I believe that this is due to known issue with Sleepyhead (i.e. the leaks are "fake news").

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max
Last edited by D.H. on Fri Feb 03, 2017 10:19 pm, edited 1 time in total.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Machine runs out of water.

Post by Wulfman... » Fri Feb 03, 2017 10:19 pm

Higher pressures will also use up water faster.......along with higher HH settings.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
jtravel
Posts: 298
Joined: Sun May 22, 2016 11:45 am
Location: Tucson,AZ

Re: Machine runs out of water.

Post by jtravel » Sun Feb 05, 2017 9:33 am

Isnore2 wrote:I have the Philips DreamStation and it consistantly is out of water when I wake up after 6 hours of use. the supplier gave me a new humidifier after my first week of use. The same problem still exists. It does have the heated hose connection. I'm running at 2 humidity. It's out of water whether I have the preheat on or not. My wife wakes up when it runs out of water because it makes a gurgling noise. I'm considering asking my insurance if there's another brand they'll cover. Having dry air blown into my nose & throat is painful. It is dry in the high desert, but that 's all the reason for a more humid air supply. And I'm not a mouth breather. 2/3/2017.
anyone know Why this users Post Name show "also Posted as jtravel" ?

my User Name is Jtravel and it's not a post from me.

_________________
MachineMask
Additional Comments: Apap 10-15 EPR setting of 3