Exercise and sleep apnea.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ChicagoGranny
Posts: 15431
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Exercise and sleep apnea.

Post by ChicagoGranny » Fri May 01, 2020 10:44 am

Garabeca wrote:
Fri May 01, 2020 9:39 am
The AHI is increasing because the pressure is too low ( due to an increased demand from the lungs for more air intake, due to metabolism change for the 24-48 window following exercise) Therefore the pressure is not sufficient, therefore the airways collapse.
Deeper breathing doesn't at all explain why the airway would be more subject to collapse and therefore need more pressure from the CPAP.

The movement of the human diaphragm controls the airflow as long as the airway is open. The CPAP is not used to deliver air to the lungs. It's only used to splint the airway open.

Do you or anyone else reading this thread have any scientific references that explain why the airway would be more subject to collapse with deeper breaths?

Garabeca
Posts: 15
Joined: Fri May 01, 2020 8:46 am

Re: Exercise and sleep apnea.

Post by Garabeca » Fri May 01, 2020 10:46 am

Goofproof wrote:
Fri May 01, 2020 10:32 am
And that's why one of the reasons they came up with APAP's, when properly set up, they will cover the increase. Jim

Changes in what we do daily affects our needs.
I appreciate the input but unless the dreamstation is doing a very bad job, the increases in pressure are too sporadic and doesn't fully cover for the baseline which is what you need in this case because your metabolism is on the other hand elevated to a higher but steady level. (steady = requiring baseline steady) On the days you need higher pressure (days you exercise) it is better to increase baseline yourself and let the apap figure out the rest but not the other way around.

Garabeca
Posts: 15
Joined: Fri May 01, 2020 8:46 am

Re: Exercise and sleep apnea.

Post by Garabeca » Fri May 01, 2020 11:06 am

ChicagoGranny wrote:
Fri May 01, 2020 10:44 am
Garabeca wrote:
Fri May 01, 2020 9:39 am
The AHI is increasing because the pressure is too low ( due to an increased demand from the lungs for more air intake, due to metabolism change for the 24-48 window following exercise) Therefore the pressure is not sufficient, therefore the airways collapse.
Deeper breathing doesn't at all explain why the airway would be more subject to collapse and therefore need more pressure from the CPAP.

The movement of the human diaphragm controls the airflow as long as the airway is open. The CPAP is not used to deliver air to the lungs. It's only used to splint the airway open.

Do you or anyone else reading this thread have any scientific references that explain why the airway would be more subject to collapse with deeper breaths?
I am speaking out of experience, scientifical evidence is not an absolute must if this can help others from what i've realized for myself, especially if it doesn't hurt to explore and try for yourself. Worst case, if higher pressure on those exercise days doesn't help, You can just go back to the pressure you were on. Also experience in some instances can speak volumes, sometimes more than scientifical evidence because often studies yield results and empirical evidence based on patient experience. I know that the diagphragm can be more easily scientifically studied due to an obvious mechanism that could be based on sleep depth (REM for example is known for airway collapse). But based on common sense, deeper breathing demands more air in the airways therefore subjecting diaphragm to more resistance prone reactions. Also you are correct to say that my theory is incomplete once again because indeed, deeper sleeping can or will make the diagphragm collapse more as well which is definitely contributing to the issue. So i will agree that it is not solely based on oxygen/airflow demand but also by the level of collapse depending on the depth of sleep induced by exercise. Mind you, overexercising might actually disrupt deep stages of sleep. In any case, for metabolism
related changes sensitive to various levels of exercise (the more the exercise intensity, the stronger the metabolism response) 24-48 hours is what i noticed for me, it could be more or less for others but it would be fair to assume for the sake of this suggestion, that at least a 24 hour period would be a good reference range for people experiencing this issue. And my observation in this situation is strictly on metabolism, this is regardless of the depth of sleep (overtraining could actually disrupt sleep depth) Again this is my experience and just hoping it can help people to whom this applies to as they will have nothing to lose by trying. Also i suggested the increase to be a cm more or less, nothing dramatic, just enough to see if it makes you more comfortable.
Last edited by Garabeca on Fri May 01, 2020 11:23 am, edited 1 time in total.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Exercise and sleep apnea.

Post by Julie » Fri May 01, 2020 11:17 am

Sorry...
Last edited by Julie on Fri May 01, 2020 12:47 pm, edited 1 time in total.

User avatar
Tricky Wash
Posts: 52
Joined: Sat Jul 11, 2015 11:57 am

Re: Exercise and sleep apnea.

Post by Tricky Wash » Fri May 01, 2020 11:21 am

His problem seems obvious:
Yayoe10 wrote:
Tue Feb 21, 2017 1:45 pm
feel sore and tired from it already at night ,
He's sore. He doesn't sleep as well and gets a lot of false event flags.

He would be better off cutting back on the exercise intensity/length and getting a good night's sleep. Physical conditioning is not just about exercise. Recovery is also needed.

Garabeca
Posts: 15
Joined: Fri May 01, 2020 8:46 am

Re: Exercise and sleep apnea.

Post by Garabeca » Fri May 01, 2020 11:27 am

Tricky Wash wrote:
Fri May 01, 2020 11:21 am
His problem seems obvious:
Yayoe10 wrote:
Tue Feb 21, 2017 1:45 pm
feel sore and tired from it already at night ,
He's sore. He doesn't sleep as well and gets a lot of false event flags.

He would be better off cutting back on the exercise intensity/length and getting a good night's sleep. Physical conditioning is not just about exercise. Recovery is also needed.
Yes i actually answered this in my last post as well, it could be related to overtraining and worse quality of sleep depth. The brain might be less able to regulate breathing mechanisms which would otherwise be smoother when person is not sore and CNS is not haywire. There is a lot going on but my suggestion remains, increase pressure and see if it helps you. Usually whether you are sore or overtrained or just had a good workout, all of these scenarios 90% of the time will still have a metabolism spike due to exercise in that 24 hour window.

User avatar
Tricky Wash
Posts: 52
Joined: Sat Jul 11, 2015 11:57 am

Re: Exercise and sleep apnea.

Post by Tricky Wash » Fri May 01, 2020 11:33 am

Garabeca wrote:
Fri May 01, 2020 11:27 am
Yes i actually answered this in my last post as well, it could be related to overtraining and worse quality of sleep depth. The brain might be less able to regulate breathing mechanisms which would otherwise be smoother when person is not sore and CNS is not haywire. There is a lot going on but my suggestion remains, increase pressure and see if it helps you. Usually whether you are sore or overtrained or just had a good workout, all of these scenarios 90% of the time will still have a metabolism spike due to exercise in that 24 hour window.
Do you use OSCAR's daily detail reports to check your CPAP performance?

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Exercise and sleep apnea.

Post by chunkyfrog » Fri May 01, 2020 11:46 am

Yayoe10 wrote:
Tue Feb 21, 2017 1:45 pm
Ive noticed every time I have a intense work out and feel sore and tired from it already at night , when I wake up the next day my AHInare always high. Not a drastic "high" but I get a change from 0.8-1.3 to a 3.5-4.5 and I can FEEL the difference which sucks because other people get way higher numbers so I should be happy with mine but i always feel way better with numbers before 1.5 especially below 1. But yea my question is can working out actually increase my AHI? I thought it was suppose to be the other way around .
Pain will disturb sleep--best to take an analgesic before bed to smooth out the bumps.
Recovery is key--do different muscle groups each time, and rest those areas for a couple days.
Less intense exercise, spread out, will produce results with less injury.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her
Last edited by chunkyfrog on Fri May 01, 2020 11:49 am, edited 1 time in total.

User avatar
ChicagoGranny
Posts: 15431
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Exercise and sleep apnea.

Post by ChicagoGranny » Fri May 01, 2020 11:49 am

Tricky Wash wrote:
Fri May 01, 2020 11:21 am
He's sore. He doesn't sleep as well and gets a lot of false event flags.
Probably. And, there is no need to raise the CPAP pressure to deal with false event flags.

His other problem is having a straight CPAP (according to his equipment profile).

Garabeca
Posts: 15
Joined: Fri May 01, 2020 8:46 am

Re: Exercise and sleep apnea.

Post by Garabeca » Fri May 01, 2020 6:56 pm

:)
I'm just trying to share what helped me because it actually worked for me and i no longer have issues with my cpap on days i exercise a little more intensely ! :D So i wanted to pass it on

Garabeca
Posts: 15
Joined: Fri May 01, 2020 8:46 am

Re: Exercise and sleep apnea.

Post by Garabeca » Fri May 01, 2020 6:59 pm

ChicagoGranny wrote:
Fri May 01, 2020 11:49 am
Tricky Wash wrote:
Fri May 01, 2020 11:21 am
He's sore. He doesn't sleep as well and gets a lot of false event flags.
Probably. And, there is no need to raise the CPAP pressure to deal with false event flags.

His other problem is having a straight CPAP (according to his equipment profile).
There is no harm in trying. One extra cm won't make such a difference anyways. I encourage people to "listen" to their bodies. That's what i did and that's how i finally got relief on exercise days. For most people, metabolism up regulates. Nothing wrong with trying to match that with an extra cm and see if it works.

Garabeca
Posts: 15
Joined: Fri May 01, 2020 8:46 am

Re: Exercise and sleep apnea.

Post by Garabeca » Fri May 01, 2020 7:03 pm

Tricky Wash wrote:
Fri May 01, 2020 11:33 am
Garabeca wrote:
Fri May 01, 2020 11:27 am
Yes i actually answered this in my last post as well, it could be related to overtraining and worse quality of sleep depth. The brain might be less able to regulate breathing mechanisms which would otherwise be smoother when person is not sore and CNS is not haywire. There is a lot going on but my suggestion remains, increase pressure and see if it helps you. Usually whether you are sore or overtrained or just had a good workout, all of these scenarios 90% of the time will still have a metabolism spike due to exercise in that 24 hour window.
Do you use OSCAR's daily detail reports to check your CPAP performance?
Good question, no i use sleepyhead. I also did a lot of analysis on sleepy head to determine my tidal volume and minute vent is higher on days i exercise therefore a slightly higher pressure is very welcome and comfortable.

Garabeca
Posts: 15
Joined: Fri May 01, 2020 8:46 am

Re: Exercise and sleep apnea.

Post by Garabeca » Fri May 01, 2020 7:05 pm

Garabeca wrote:
Fri May 01, 2020 6:59 pm
ChicagoGranny wrote:
Fri May 01, 2020 11:49 am
Tricky Wash wrote:
Fri May 01, 2020 11:21 am
He's sore. He doesn't sleep as well and gets a lot of false event flags.
Probably. And, there is no need to raise the CPAP pressure to deal with false event flags.

His other problem is having a straight CPAP (according to his equipment profile).
There is no harm in trying. One extra cm won't make such a difference anyways. I encourage people to "listen" to their bodies and if you can, get into analysing your own data on sleepyhead. That's what i did and that's how i finally got relief on exercise days. For most people, metabolism up regulates on exercise days. Tidal volume and minute vent increases. Nothing wrong with trying to match that with an extra cm to see if it helps things to be more comfortable.

Garabeca
Posts: 15
Joined: Fri May 01, 2020 8:46 am

Re: Exercise and sleep apnea.

Post by Garabeca » Fri May 01, 2020 7:20 pm

Garabeca wrote:
Fri May 01, 2020 7:05 pm
Garabeca wrote:
Fri May 01, 2020 6:59 pm
ChicagoGranny wrote:
Fri May 01, 2020 11:49 am
Tricky Wash wrote:
Fri May 01, 2020 11:21 am
He's sore. He doesn't sleep as well and gets a lot of false event flags.
Probably. And, there is no need to raise the CPAP pressure to deal with false event flags.

His other problem is having a straight CPAP (according to his equipment profile).
There is no harm in trying. One extra cm won't make such a difference anyways. I encourage people to "listen" to their bodies and if you can, get into analysing your own data on sleepyhead. That's what i did and that's how i finally got relief on exercise days. For most people, metabolism up regulates on exercise days. Tidal volume and minute vent increases. Nothing wrong with trying to match that with an extra cm to see if it helps things to be more comfortable.
Also wanted to point out, something is causing the false event flags and it surely isn't because the pressure is too high. His body is sore, that equals to his body being tired and in a recovery process. In no ordinary world will a tired and sore body due to exercise, require less oxygen to operate. And so, more oxygen demand equals more pressure demand as simple as that. Other wise he will mouth breathe or experience lack of oxygen which will cause him to snore and struggle as tidal volume rise from usual baseline is not being met by pressuse. Thus Creating false event flags and ultimately even wake up. unfortunately, the APAP is a good machine for sporadic need for pressure increase throughout the night. But the baseline pressure (also straight CPAP pressure) is what's important here on days the body is exhausted from exercise.

Garabeca
Posts: 15
Joined: Fri May 01, 2020 8:46 am

Re: Exercise and sleep apnea.

Post by Garabeca » Fri May 01, 2020 7:28 pm

Garabeca wrote:
Fri May 01, 2020 7:20 pm
Garabeca wrote:
Fri May 01, 2020 7:05 pm
Garabeca wrote:
Fri May 01, 2020 6:59 pm
ChicagoGranny wrote:
Fri May 01, 2020 11:49 am
Tricky Wash wrote:
Fri May 01, 2020 11:21 am
He's sore. He doesn't sleep as well and gets a lot of false event flags.
Probably. And, there is no need to raise the CPAP pressure to deal with false event flags.

His other problem is having a straight CPAP (according to his equipment profile).
There is no harm in trying. One extra cm won't make such a difference anyways. I encourage people to "listen" to their bodies and if you can, get into analysing your own data on sleepyhead. That's what i did and that's how i finally got relief on exercise days. For most people, metabolism up regulates on exercise days. Tidal volume and minute vent increases. Nothing wrong with trying to match that with an extra cm to see if it helps things to be more comfortable.
Also wanted to point out, something is causing the false event flags and it surely isn't because the pressure is too high. His body is sore, that equals to his body being tired and in a recovery process. In no ordinary world will a tired and sore body due to exercise, require less oxygen to operate. And so, more oxygen demands more pressure as simple as that. Other wise he will mouth breathe (false event flag) or experience lack of airflow/oxygen which will cause him to snore and struggle (false event flag) as his tidal volume rise from his usual baseline pressure is not being met anymore for that 24-48 hours window (depending on the intensity of the exercise) where the body is in need of more oxygen, thus Creating false event flags. Unfortunately, the APAP is a good machine for sporadic need for pressure increase throughout the night. But the baseline pressure (also straight CPAP pressure) is what's important here on days the body is exhausted from exercise. ALso keep in mind, athletes will recover faster and might need even less than 24 hours to recover their regular need for oxygen intake. So this is case by case and applies for people who brought up this topic in the first place. And lastly, using this logic, exercising earlier will help but probably not for the same following night of sleep but most probably for the next day's night of sleep.