respironics remstar servicing required

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
craigclive
Posts: 8
Joined: Sun Nov 12, 2017 4:03 pm
Location: Chestnut Hill Cove, MD

respironics remstar servicing required

Post by craigclive » Fri Nov 17, 2017 12:44 pm

I had posted this before but didn't get any responses. I have a Respironics Remstar cpap I use for traveling. Last trip I received a message "servicing required" and the machine shut off. I could not get it to work.

I have contacted a repair company about my issue and they responded as follows:

"Service required" error code means your CPAP needs to be sent in for repair or replacement, if it is still under warranty. The code usually means a fatal error.

Is this true? I noted on their website that there is a $100 deposit which can be used for a new machine.

Thanks,
Craig

_________________
Machine: DreamStation CPAP Machine
Humidifier: DreamStation Heated Humidifier
Additional Comments: the dreamstation machine is a DX500xxxx

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: respironics remstar servicing required

Post by Goofproof » Fri Nov 17, 2017 1:01 pm

You haven't posted enough info to get a answer. Like who did you contact?? How many hours are on the machine, when did you buy it. And please use the same post not starting a new one over. It may be a better deal to get a new unit. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

Guest

Re: respironics remstar servicing required

Post by Guest » Fri Nov 17, 2017 1:13 pm

craigclive wrote:I had posted this before but didn't get any responses. I have a Respironics Remstar cpap I use for traveling. Last trip I received a message "servicing required" and the machine shut off. I could not get it to work.

I have contacted a repair company about my issue and they responded as follows:

"Service required" error code means your CPAP needs to be sent in for repair or replacement, if it is still under warranty. The code usually means a fatal error.

Is this true? I noted on their website that there is a $100 deposit which can be used for a new machine.

Thanks,
Craig
I inherited a remstar pro.
Right! That's like saying "I have a Ford and the "Check Engine Light" came on in the dash."

You haven't told us which generation of "RemStar Pro" you have. They've used that model designation in almost every generation they've made. It's likely to be one of the following.
Is it a "M Series", a "System One 50 Series" or a "System One 60 Series"?
It was more prevalent in the M Series machines.

Which repair shop did you contact?

That message has been noted in these forum pages for quite a few years, so some additional information would help.
The company you contacted (if they're reputable) should have been able to give you some indication of what the nature of the problem is and a range of costs to fix it.

The dollar amount mentioned is probably normal.

Woody
Posts: 591
Joined: Sat Mar 10, 2007 7:59 pm

Re: respironics remstar servicing required

Post by Woody » Fri Nov 17, 2017 8:15 pm

I had one that set some error code and all I did was unplug it, let it sit for a minute.
This cleared the error and I used it another 6 months till I got a new machine.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: SleapyHead software

craigclive
Posts: 8
Joined: Sun Nov 12, 2017 4:03 pm
Location: Chestnut Hill Cove, MD

Re: respironics remstar servicing required

Post by craigclive » Thu Nov 23, 2017 9:27 am

The cpap is a Remstar Auto A-Flex. Usage hours are 1066.5 The device is about 2 years old. I inherited it brand new from a relative who died before using the machine.

I contacted the following:
Respshop
9215 151st Ave NE
Redmond, WA, 98052

Help!!!

_________________
Machine: DreamStation CPAP Machine
Humidifier: DreamStation Heated Humidifier
Additional Comments: the dreamstation machine is a DX500xxxx

ractar28
Posts: 222
Joined: Wed Oct 01, 2008 3:06 pm

Re: respironics remstar servicing required

Post by ractar28 » Thu Nov 23, 2017 9:43 am

First thing to note is how much a new machine will cost. From Respironics, I suspect it will be EXPENSIVE. You can find a suitable replacement for $400 (although you'll have to program it yourself). I posted such a deal recently on here.


User avatar
Julie
Posts: 20017
Joined: Tue Feb 28, 2006 12:58 pm

Re: respironics remstar servicing required

Post by Julie » Thu Nov 23, 2017 9:51 am

You might want to look at Acbio.com who repair lots and lots of machines.

User avatar
WRX03
Posts: 137
Joined: Fri Apr 29, 2016 9:37 am
Location: Roanoke, Virginia

Re: respironics remstar servicing required

Post by WRX03 » Fri Nov 24, 2017 6:51 am

Craig did you do what I recommended on your other post about this? If a unit seats a while the blower sometimes it gets stuck. If there was high humidity when it was last used the blower sometimes gets stuck. If you spin the blower a little it "might" start working. I see this a lot with error codes 20, 22, 24, & 30. Also Respironics self measures the ohms of the blower, if the ohms are out of range it will throw a code (dirt, rust, moisture). 1000hrs is low, so this may work for you, give it a spin.

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Actually PCA is a "For Her". It's DIY piece together from spare parts CPAP. It's multi color! LOL