ResMed Air Sense 10 won't power up

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
sleepez
Posts: 3
Joined: Sat Feb 09, 2019 8:53 am

ResMed Air Sense 10 won't power up

Post by sleepez » Sat Feb 09, 2019 9:07 am

Just got this machine, less than one month old. The machine was working fine. Took it on a week-long road trip. It worked fine in the hotel. Returned home, set it back up in my bedroom and turned it on. The machine initially began to power up, but then died. The air stopped and the screen went dark. Tried resetting by holding the power button for three seconds....no help. Ensured nothing was blocking the air tube. Tried resetting the air tube and SD card....nothing helps, cannot get the machine to power up. Any ideas???????? Thanks in advance

User avatar
Okie bipap
Posts: 3566
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: ResMed Air Sense 10 won't power up

Post by Okie bipap » Sat Feb 09, 2019 9:57 am

Have you checked all of the electrical connections to ensure they are securely seated? Have you tried to plug it into a different wall socket? The contacts within the wall socket can spread apart and not make proper contact with the plug. Does the green light on the power brick light up? If not, you may have a bad power brick. Your machine may have died on you. Sometimes machines just quit working for one reason or another. One of mine died while we were out of town. I set it up in the hotel and it refused to come on. I checked everything I mentioned above and it still would not work. My DME changed it out for me once I got back home and took it in to them.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

sleepez
Posts: 3
Joined: Sat Feb 09, 2019 8:53 am

Re: ResMed Air Sense 10 won't power up

Post by sleepez » Sun Feb 10, 2019 10:29 am

Your reply is spot on. I am embarrassed to admit it, but the problem was a loose connection at the brick. I am embarrassed bc I am a 22-year retired Aviation Electrician, USN. Been troubleshooting electrical problems all my life. The brick has a shroud around it where the plug connects, so the physical connection is somewhat hidden, but I did push on the plug. Only thing I can figure out is that I must not have pushed hard enough. Found the problem as I retraced my steps for the 4th time. I was cringing at the thought of having to make an appointment with the VA bc that is a long lead time.

Anyhow, your post is spot on and I greatly appreciate it. The ResMed is a beautiful machine and, so far, is working just fine. I do not do well without the CPAP machine and usually begin to panic when something goes wrong with it. Thanks again for your reply

Kind Regards,
Glenn

User avatar
Okie bipap
Posts: 3566
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: ResMed Air Sense 10 won't power up

Post by Okie bipap » Sun Feb 10, 2019 6:12 pm

Glad it worked for you. I understand about the VA being a long lead time on anything. I have not had to have much interface with them since I retired. The companies I worked for had very good medical insurance, and now I am on Medicare and Tricare. I am currently trying to have my disability rating adjusted. Don't know if it will work or not, but I am trying.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

sleepez
Posts: 3
Joined: Sat Feb 09, 2019 8:53 am

Re: ResMed Air Sense 10 won't power up

Post by sleepez » Tue Feb 12, 2019 8:41 pm

Good luck on your disability review. Just finished up with that process. Took me 2 1/2 years. Not sure what caused more stress, my disabilities or the review process. Anyhow, waiting on my new award level now. Very relieved it is over with, however it turns out. Looking forward to retirement.........soon.

Take Care,
Glenn

holisticdmd
Posts: 1
Joined: Sun Apr 12, 2020 3:54 am

Re: ResMed Air Sense 10 won't power up

Post by holisticdmd » Sun Apr 12, 2020 4:04 am

My cpap goes dead every 1-3 months, it's new and unfortunately I am out of the country for a while due to the Covid-19 situation so can not get back to the service place. If i play around with the connections over a 2-3 day period it can start working again. Does this machine have an inherent connection condition or a reset/ overload button? the power can be dicey down her and this usually happens in the middle of the night when the power goes out. I have a generator so the power always comes back on in a minute or two.Very strange and very annoying. Hard to believe it is a connection that goes bad just because the power goes out and comes back on again.It has happened 2-3 times now all under this same circumstance. Also, I have voltage regulators and surge protection for my entire home down here in Costa Rica due to energy issues so I doubt that is the problem.

Nthernlgts
Posts: 15
Joined: Fri Nov 22, 2019 11:15 am

Re: ResMed Air Sense 10 won't power up

Post by Nthernlgts » Tue Apr 21, 2020 10:27 am

I had the same issue happen last night. We also had a rare power interruption at the house late yesterday afternoon, and only for a few seconds. I went to bed, the unit was totally dead. I checked all the connections unplugged it plugged it back in, etc... . Nothing. I grabbed a different power supply from one of my laptops that had the same style connector and same wattage. The light flashed briefly on the top of the unit, then nothing. Went back to the original power supply and tried it one more time. Voila! It lit up like a Christmas tree and ran all night. Go figure. There’s got to be something very touchy in the electronics.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Air Sense 10 won't power up

Post by palerider » Tue Apr 21, 2020 12:43 pm

Nthernlgts wrote:
Tue Apr 21, 2020 10:27 am
I had the same issue happen last night. We also had a rare power interruption at the house late yesterday afternoon, and only for a few seconds. I went to bed, the unit was totally dead. I checked all the connections unplugged it plugged it back in, etc... . Nothing. I grabbed a different power supply from one of my laptops that had the same style connector and same wattage. The light flashed briefly on the top of the unit, then nothing. Went back to the original power supply and tried it one more time. Voila! It lit up like a Christmas tree and ran all night. Go figure. There’s got to be something very touchy in the electronics.
"The same connector and wattage" isn't good enough, it needs the right voltage (24vdc) too.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

redrocklobster
Posts: 1
Joined: Sun Jul 03, 2022 6:24 am

Re: ResMed Air Sense 10 won't power up

Post by redrocklobster » Sun Jul 03, 2022 6:36 am

My AirSense 10 bilevel is on the fritz in a similar way. Power supply seems to check out. The screen flashes white twice, a second apart, every couple minutes. Otherwise black. The top start/stop light flickers green sometimes. Working fine 25 hours ago. Traveled, might have got knocked on luggage. (Yes I sometimes need to pack it).

Is it dead or any hope?

User avatar
SleepGeek
Posts: 1055
Joined: Mon Feb 09, 2015 2:56 pm

Re: ResMed Air Sense 10 won't power up

Post by SleepGeek » Sun Jul 03, 2022 10:30 am

redrocklobster wrote:
Sun Jul 03, 2022 6:36 am
Traveled, might have got knocked on luggage. (Yes I sometimes need to pack it).
Did water get spilled inside the electronics by any chance?
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: What exactly do you think you're adding to this thread?

Grumpy48
Posts: 411
Joined: Tue May 12, 2020 7:57 am

Re: ResMed Air Sense 10 won't power up

Post by Grumpy48 » Sun Jul 03, 2022 11:40 am

I've noticed the plug/jack on my AS10 is a pretty snug fit and it does need an extra bit of push to get it seated all the way in. Quite possibly it won't make a good connection if not all the way in.

_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion)
Additional Comments: Resmed AirSense 10 Auto with Humidifier

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: ResMed Air Sense 10 won't power up

Post by Wulfman... » Sun Jul 03, 2022 1:30 pm

sleepez wrote:
Sun Feb 10, 2019 10:29 am
Your reply is spot on. I am embarrassed to admit it, but the problem was a loose connection at the brick. I am embarrassed bc I am a 22-year retired Aviation Electrician, USN. Been troubleshooting electrical problems all my life. The brick has a shroud around it where the plug connects, so the physical connection is somewhat hidden, but I did push on the plug. Only thing I can figure out is that I must not have pushed hard enough. Found the problem as I retraced my steps for the 4th time. I was cringing at the thought of having to make an appointment with the VA bc that is a long lead time.

Anyhow, your post is spot on and I greatly appreciate it. The ResMed is a beautiful machine and, so far, is working just fine. I do not do well without the CPAP machine and usually begin to panic when something goes wrong with it. Thanks again for your reply

Kind Regards,
Glenn
Does that mean it is actually working again? If it is, I haven't been able to find that specific part in your post.

Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

jsmit86
Posts: 253
Joined: Wed Mar 20, 2019 2:43 pm
Location: Conroe, TX (Suburban Houston TX)

Re: ResMed Air Sense 10 won't power up

Post by jsmit86 » Sun Jul 03, 2022 3:26 pm

I have had a possibly similar problem with my unit. The brick is plugged in tight, but sometimes the plug going into the yellow connector on the machine does not make a good connection. I had a time when I thought I could not get it back working, but then I took a small metal nail file, and ran it around the round collar of that jack where the metal barrel makes contact with the inside of the jack. That fixed it. Sometimes when the unit gets unplugged, I have to do that again.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto

Architect
Posts: 168
Joined: Sun Jan 31, 2021 5:09 pm
Location: Central Tx

Re: ResMed Air Sense 10 won't power up

Post by Architect » Sun Jul 03, 2022 5:19 pm

sleepez wrote:
Sun Feb 10, 2019 10:29 am
I am embarrassed to admit it, but the problem was a loose connection at the brick.
Happy, you were able to find solutoin, thanks to Okie. Years ago, wasn't a member to sleep/pap forums. My pap did same thing, my pap wouldn't come on, called VA made appt to check pap. I packed my pap up in bag, waited 4 day's for appt. Got to clnic, sleep tech plugged my unit up, to clinic connection, pap machine worked fine.

Do I feel stupid :roll: plug was in at home, but not pushed all the way in. That's what apnea brain fog does :wink: