ResMed AirSense 10 No Cellular Connection

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
scottkeen
Posts: 1
Joined: Wed Jun 12, 2019 9:03 am

ResMed AirSense 10 No Cellular Connection

Post by scottkeen » Wed Jun 12, 2019 9:10 am

My AirSense 10 stopped transmitting maybe a year or more ago. The connection icon has a circle-with-a-line-through-it icon (not bars, not airplane) which indicates the cellular modem is on but not connecting.

It worked fine for a year, but then I stopped going to the sleep center doctor and stopped returning their calls. I'm wondering if they disabled the cellular modem connection? I've tried my AirSense in several locations, where I live in Virginia, when I'm traveling, and my Dad's house in Florida (he has an AirSense 10 too and it connects fine), but mine won't get a signal and just shows the circle-with-a-line-through-it icon.

Is this because I stopped doing business with the sleep center?

User avatar
Dog Slobber
Posts: 4157
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: ResMed AirSense 10 No Cellular Connection

Post by Dog Slobber » Wed Jun 12, 2019 10:40 am

Your DME can't disable your modem

Your modem may have just stopped functioning or it might be an older system that used a now unsupported Cell system. ie CDMA.

Go into About and compare your Cell Type to that of your fathers.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: ResMed AirSense 10 No Cellular Connection

Post by Goofproof » Wed Jun 12, 2019 11:17 am

Most people would call stopping the call home a a service not a problem. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: ResMed AirSense 10 No Cellular Connection

Post by chunkyfrog » Wed Jun 12, 2019 11:35 am

Goofproof wrote:
Wed Jun 12, 2019 11:17 am
Most people would call stopping the call home a a service not a problem. Jim
Exactly!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
ChicagoGranny
Posts: 15071
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: ResMed AirSense 10 No Cellular Connection

Post by ChicagoGranny » Wed Jun 12, 2019 12:51 pm

scottkeen wrote:
Wed Jun 12, 2019 9:10 am
My AirSense 10 stopped transmitting maybe a year or more ago.
So far, you seem to have done OK without it.

NoOnesPerfect
Posts: 163
Joined: Thu May 09, 2019 7:29 am
Location: Washington, USA

Re: ResMed AirSense 10 No Cellular Connection

Post by NoOnesPerfect » Wed Jun 12, 2019 1:06 pm

Other than making insurance related compliance easier to monitor, the modem serves no useful purpose. I don’t see the problem here.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: ResMed AirSense 10 No Cellular Connection

Post by Goofproof » Wed Jun 12, 2019 1:40 pm

NoOnesPerfect wrote:
Wed Jun 12, 2019 1:06 pm
Other than making insurance related compliance easier to monitor, the modem serves no useful purpose. I don’t see the problem here.
And when you own your machine compliance isn't a factor, plus you can use Real Software, and get the Real Data, so you can tweak your treatment for the best results. Win, Win, and Win! Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

EROCK
Posts: 1
Joined: Wed Dec 14, 2022 8:59 pm

Re: ResMed AirSense 10 No Cellular Connection

Post by EROCK » Wed Dec 14, 2022 9:04 pm

Forget what Dog Slobber said. I am guessing he hasn't messed with a Cell Modem ever. If the configuration is wrong you will get the line circle. You should also get that if you are out of range of a tower. I am going to guess they had a configuration issue or error. If you unplug the modem or it died you wont see anything at all in the area.

User avatar
Jlfinkels
Posts: 468
Joined: Sun Jun 20, 2021 3:51 pm

Re: ResMed AirSense 10 No Cellular Connection

Post by Jlfinkels » Wed Dec 14, 2022 9:29 pm

Do as Dog Slobber said, go to “My Options”, scroll to the bottom for “About”, then scroll down until you see “Internal Modem”. If there is no data then the modem is gone, bereft of life, it is no more. If there is data then there should be something in the “Provider” field (ATT, Sprint, T-Mobile, Verizon, etc). Also look at the signal level, which has to be above -110 dBm for reliable data transmission. And then look at “Type”, which should say 4G (if it says 3G then it may have been disabled by the providers where you tested).

Do the same with your fathers machine. Compare the results. Even better, take a photo of it and post it here.
Sometimes it is the very people who no one imagines anything of who do the things that no one imagines

User avatar
ozij
Posts: 10429
Joined: Fri Mar 18, 2005 11:52 pm

Re: ResMed AirSense 10 No Cellular Connection

Post by ozij » Wed Dec 14, 2022 10:59 pm

Zomby thread, 2019, doesn't look in need of further responses to the OP.

But just to clarify for those who may have a similar concern and read
Jlfinkels wrote:
Wed Dec 14, 2022 9:29 pm
Do as Dog Slobber said, go to “My Options”, scroll to the bottom for “About”, then scroll down until you see “Internal Modem”. If there is no data then the modem is gone, bereft of life, it is no more. If there is data then there should be something in the “Provider” field (ATT, Sprint, T-Mobile, Verizon, etc). Also look at the signal level, which has to be above -110 dBm for reliable data transmission. And then look at “Type”, which should say 4G (if it says 3G then it may have been disabled by the providers where you tested).
If it says 3G then 3G cell transmission may have been disabled by the Cell provider in the areas where you tested.
Older machines have 3G modems, newer ones have 4G.

Edited to add the word "further" to my warning. No criticism of Jlfinkels's response intended.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
Last edited by ozij on Thu Dec 15, 2022 12:03 am, edited 1 time in total.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Jlfinkels
Posts: 468
Joined: Sun Jun 20, 2021 3:51 pm

Re: ResMed AirSense 10 No Cellular Connection

Post by Jlfinkels » Wed Dec 14, 2022 11:11 pm

ozij wrote:
Wed Dec 14, 2022 10:59 pm
Zomby thread, 2019, doesn't look in need of responses to the OP.
Fair enough mate. I should have read more closely, but was concerned that the response about cell configs needed an accurate response.
Sometimes it is the very people who no one imagines anything of who do the things that no one imagines

User avatar
ozij
Posts: 10429
Joined: Fri Mar 18, 2005 11:52 pm

Re: ResMed AirSense 10 No Cellular Connection

Post by ozij » Thu Dec 15, 2022 12:01 am

Jlfinkels wrote:
Wed Dec 14, 2022 11:11 pm
ozij wrote:
Wed Dec 14, 2022 10:59 pm
Zomby thread, 2019, doesn't look in need of responses to the OP.
Fair enough mate. I should have read more closely, but was concerned that the response about cell configs needed an accurate response.
Not meant for you. Meant for others who may have considered responding. I though your response informative, and good and have edited my warning.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Dog Slobber
Posts: 4157
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: ResMed AirSense 10 No Cellular Connection

Post by Dog Slobber » Thu Dec 15, 2022 12:58 pm

Although this is a zombie thread, it's still very relevant. Actually some might argue it's even more relevant now as carriers have been shutting down 3g in 2022.

In any event what was posted in 2019 was true then and still true now.
  • DME's can not disable the modem
  • The timing of the OP suggests he had an early Air10 with a CDMA modem, with CDMA being shutdown
  • It could be malfunctioning hardware
Why EROCK necro'd a zombie post to demonstrate he's an idiot is beyond me, ResMed Air10 modems aren't user or DME configurable.