Are machines numbers accurate?
Are machines numbers accurate?
My machine is reporting an average of 29 AHI’s per hour since early January. I had 2 sleep studies. The first was inconclusive as I wasn’t able to sleep long enough to obtain an accurate study. The second showed only 15 events/hour that were relieved with a CPAP pressure of 15 and I was subsequently started on CPAP. Mask leakage is not a problem. I’ve tried adjusting the pressure up and using the ‘Autoset’ mode with no real change. Just wondering what numbers I should trust and any recommendations you guys may have. Just FYI, neither the sleep center nor the DME provider have been willing to say anything. My doctor is a pulmonologist and doesn’t seem to understand my concern. Thanks in advance!
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Re: Are machines numbers accurate?
welcome to the zoo! if doctors would pay attention to their sleep apnea patients and offered good advice, there would be no need for this forum! seriously, you aren't alone in this as we've all been there.BillSc wrote: ↑Thu Feb 27, 2020 5:30 pmMy machine is reporting an average of 29 AHI’s per hour since early January. I had 2 sleep studies. The first was inconclusive as I wasn’t able to sleep long enough to obtain an accurate study. The second showed only 15 events/hour that were relieved with a CPAP pressure of 15 and I was subsequently started on CPAP. Mask leakage is not a problem. I’ve tried adjusting the pressure up and using the ‘Autoset’ mode with no real change. Just wondering what numbers I should trust and any recommendations you guys may have. Just FYI, neither the sleep center nor the DME provider have been willing to say anything. My doctor is a pulmonologist and doesn’t seem to understand my concern. Thanks in advance!
first up, we need to know if the machine you've selected in your profile is the you actually have. there are many different airsense models. could you check this list and verify?
wiki/index.php/CPAP_models
i'm not just being pedantic here. not all of these models are the same. and i want to be sure the model you have is fully data capable.
why? because then you can get the free software oscar and start monitoring your own data. even better, you can upload data here for the experts to look over and advise upon.
but to fully answer your question: yes, the numbers are accurate.
good luck!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Are machines numbers accurate?
No! But close enough to allow you to get your treatment correct, after all the human body isn't very accurate either. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Are machines numbers accurate?
He said he has tried the autoset mode..I guess he has an Autoset
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
Re: Are machines numbers accurate?
The answer is both Yes and No at the same time.
Your CPAP machine accurately reports what is going on during your sleep.
And, … the in-lab accurately records (after certain adjustments) what was going on during your sleep study.
The problem in comparing the two results comes from the baseline definition of a sleep event is different between the two.
For example, suppose both systems see an interruption in your breathing that is the result of an obstructive event.
Your home machine will record that as an OSA event (if your machine is a full data machine.)
The in-lab machine will record in the same way BUT when finally scored, the event sometimes will be as if it never happened. If your EEG recording indicates that an arousal or actual awakening is taking place at the same time, the event is not scored so the results will be different. The technicians can tell whether or not you are "asleep" and only score events then.
The implication is that the in-lab results will be lower than that reported at home. If your sleep is nice and steady with out many arousals / awakenings, then the results will be close.
If your sleep is quite disturbed, the results will be very different.
Your CPAP machine accurately reports what is going on during your sleep.
And, … the in-lab accurately records (after certain adjustments) what was going on during your sleep study.
The problem in comparing the two results comes from the baseline definition of a sleep event is different between the two.
For example, suppose both systems see an interruption in your breathing that is the result of an obstructive event.
Your home machine will record that as an OSA event (if your machine is a full data machine.)
The in-lab machine will record in the same way BUT when finally scored, the event sometimes will be as if it never happened. If your EEG recording indicates that an arousal or actual awakening is taking place at the same time, the event is not scored so the results will be different. The technicians can tell whether or not you are "asleep" and only score events then.
The implication is that the in-lab results will be lower than that reported at home. If your sleep is nice and steady with out many arousals / awakenings, then the results will be close.
If your sleep is quite disturbed, the results will be very different.
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Are machines numbers accurate?
Thanks for the responses. The machine is a brand new Airsense 10 Autoset. On the list it’s identified as #37028. I’m very interested in trying the software if it works with my machine. My doctor has recommended another CPAP Titration Study to see if the pressure needs further adjusting. I didn’t tell him that I had already changed it myself to 16 and later changed it to the Autoset mode. Nothing I’ve done so far has had any effect on the number of AHI’s. I’m really reluctant to do another sleep study. Not sure what these numbers represent, however, the machine shows the following under ‘Sleep Report’ (Technician mode):
AHI 28.6
Total AI 27.7
Central AI 26.0
If applicable, I would love to try the software and hopefully post my results.
Thanks again!
AHI 28.6
Total AI 27.7
Central AI 26.0
If applicable, I would love to try the software and hopefully post my results.
Thanks again!
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Re: Are machines numbers accurate?
do you have an sd card? if so, stick it in your machine and let it capture data overnight.
i wish one of those experts would show up now. i'm not entirely certain still that oscar will work in your machine.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Are machines numbers accurate?
No, the problem in comparing them comes from the fact that a sleep study is a brief snapshot of *one night's (or less) sleep*.JDS74 wrote: ↑Thu Feb 27, 2020 10:12 pmThe answer is both Yes and No at the same time.
Your CPAP machine accurately reports what is going on during your sleep.
And, … the in-lab accurately records (after certain adjustments) what was going on during your sleep study.
The problem in comparing the two results comes from the baseline definition of a sleep event is different between the two.
Sleep varies from night to night, and even hour to hour.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Are machines numbers accurate?
You have the wrong machine selected in your profile.
The one you want says "AutoSet", and it's under the R's, not the A's.
It's applicable.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Are machines numbers accurate?
The numbers are good "enough" to use in tweaking your therapy.
Lab sleep tests are a picture of only ONE night, but your cpap reports every night used.
Obviously, we have potentially superior reporting, by virtue of quantity alone.
Shhhhh, some lab owners will be upset--but who cares?
Lab sleep tests are a picture of only ONE night, but your cpap reports every night used.
Obviously, we have potentially superior reporting, by virtue of quantity alone.
Shhhhh, some lab owners will be upset--but who cares?
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Are machines numbers accurate?
What are your settings now.
What have you changed.
Are you using EPR ?
What have you changed.
Are you using EPR ?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: Are machines numbers accurate?
AHI 28.6
Total AI 27.7
Central AI 26.0
The majority of events are Central in nature your Autoset won’t help that you need to be on a different machine a BiLevel fixed it for some people but often an ASV will be required if you don’t have heart problems
Total AI 27.7
Central AI 26.0
The majority of events are Central in nature your Autoset won’t help that you need to be on a different machine a BiLevel fixed it for some people but often an ASV will be required if you don’t have heart problems
_________________
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
Re: Are machines numbers accurate?
Well, that just *screams* GET OSCAR
If the centrals a're real.Jas_williams wrote: ↑Fri Feb 28, 2020 7:03 amThe majority of events are Central in nature your Autoset won’t help that you need to be on a different machine a BiLevel fixed it for some people but often an ASV will be required if you don’t have heart problems
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Are machines numbers accurate?
Believe I have the correct machine now in my profile. Presently using the Autoset vice CPAP mode. Hope that’s okay. Will try to have an SD card placed into the machine tonight.
One thing I wanted to mention is my sleep doctor talked about sending me to a neurologist. He says they have meds available to help with people having centrals. Wasn’t aware of that. Wonder if that’s true?
Thanks for everything!
Bill
One thing I wanted to mention is my sleep doctor talked about sending me to a neurologist. He says they have meds available to help with people having centrals. Wasn’t aware of that. Wonder if that’s true?
Thanks for everything!
Bill
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Re: Are machines numbers accurate?
There are meds to help with central apneas but those are usually given when the people are having central apneas due to altitude issues.
So it's true but they aren't commonly given except in special situations.
So it's true but they aren't commonly given except in special situations.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.