ResMed N30i Nasal/Pillow Mask

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
FLORINE
Posts: 1
Joined: Thu Mar 12, 2020 2:28 pm

ResMed N30i Nasal/Pillow Mask

Post by FLORINE » Thu Mar 12, 2020 3:25 pm

Should settings on ResMed 10 machine be set at NASAL or PILLOW for the N30i mask ? Please advise.

User avatar
Pugsy
Posts: 64923
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed N30i Nasal/Pillow Mask

Post by Pugsy » Thu Mar 12, 2020 6:22 pm

According to the user guide paperwork that should have come with the mask.....setting should be "pillows"

https://www.resmed.com/us/dam/documents ... er_eng.pdf

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

SleepyPaolo
Posts: 57
Joined: Sat Nov 23, 2019 1:08 am

Re: ResMed N30i Nasal/Pillow Mask

Post by SleepyPaolo » Fri Mar 13, 2020 12:26 am

Yep, pillows.
The nasal setting is for those nasal masks that cover the nose over the bridge.
The nasal cushions that sit under the nose only like the N30i use the pillow setting.