Backup battery for unexpected power outages Resmed Airsense 10

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
gardengirl77
Posts: 3
Joined: Sun Jul 12, 2020 12:46 pm

Backup battery for unexpected power outages Resmed Airsense 10

Post by gardengirl77 » Sun Jul 12, 2020 12:55 pm

I would like to get a backup battery for my Resmed Airsense 10 autoset for power outages. I hope the machine will be able to hold a charge so I can use it if the power goes out unexpectedly.

My husband suggests this machine - the Jackery Portable Power Station Generator but I notice the comments sound like it can turn off after 3 hours if not drawing enough wattage.

https://www.amazon.com/Jackery-Portable ... 881&sr=8-3

This one looks good but some comments say it didn't hold a charge. MAXOAK CPAP Battery Backup CPAP Power Bank for Airsense 10 Resmed S9 Philip REMstar System one 60 50 Dreamstation Z1 Aircurve Transcend Mini Travel Camping Emergency Power DC12/15/24V&USB5V

https://www.amazon.com/MAXOAK-Airsense- ... 966&sr=8-3

I am hoping the machine can just sit by my bed and if the power goes out, I can plug my machine into it and go right back to sleep. Do you know if either of the above will do that? If not, do you have any other backup batteries you recommend? I like in New England if that makes a difference.

Thank you!!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by palerider » Sun Jul 12, 2020 4:30 pm

Unless your husband wants to blow a ton of money on little capacity, he should help you with a 35ah mobility scooter battery, a battery tender, battery box, and the 12v power cable for your cpap.

That'll last you 5 nights or so without heated humidity, and cost a fraction of those expensive things you're looking at.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

gardengirl77
Posts: 3
Joined: Sun Jul 12, 2020 12:46 pm

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by gardengirl77 » Wed Jul 15, 2020 3:10 pm

Thank you for the reply. I am afraid that suggestion is a bit beyond my level of technical ability. Maybe my husband can do what you recommend.

I am wondering if you or anyone knows if the Jackery Portable Power Station Generator is hard to keep on when using and if it is something people who have CPAPs often use as a backup battery for times the power suddenly goes out. We are thinking of times when we are sleeping and the power goes out, something that happened recently. It would be very nice at those times to be able to simply unplug our current machines and plug them into a backup battery without having to get out of bed and then go back to sleep. Here is the link to the Jackery Portable Power Station Generator.
https://www.amazon.com/Jackery-Portable ... 881&sr=8-3

I wonder if this other one works better for CPAPs. MAXOAK CPAP Battery Backup CPAP Power Bank for Airsense 10 Resmed S9 Philip REMstar System one 60 50 Dreamstation Z1 Aircurve Transcend Mini Travel Camping Emergency Power DC12/15/24V&USB5V

https://www.amazon.com/MAXOAK-Airsense- ... 966&sr=8-3

We have Resmed Air Sense 10s Autoset and are in New England.

Thanks for any replies.

User avatar
khauser
Posts: 1308
Joined: Fri Feb 01, 2013 9:31 am
Location: Nashua, NH

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by khauser » Wed Jul 15, 2020 3:28 pm

gardengirl77 wrote:
Wed Jul 15, 2020 3:10 pm
Thank you for the reply. I am afraid that suggestion is a bit beyond my level of technical ability. Maybe my husband can do what you recommend.

I am wondering if you or anyone knows if the Jackery Portable Power Station Generator is hard to keep on when using and if it is something people who have CPAPs often use as a backup battery for times the power suddenly goes out. We are thinking of times when we are sleeping and the power goes out, something that happened recently. It would be very nice at those times to be able to simply unplug our current machines and plug them into a backup battery without having to get out of bed and then go back to sleep. Here is the link to the Jackery Portable Power Station Generator.
https://www.amazon.com/Jackery-Portable ... 881&sr=8-3

I wonder if this other one works better for CPAPs. MAXOAK CPAP Battery Backup CPAP Power Bank for Airsense 10 Resmed S9 Philip REMstar System one 60 50 Dreamstation Z1 Aircurve Transcend Mini Travel Camping Emergency Power DC12/15/24V&USB5V

https://www.amazon.com/MAXOAK-Airsense- ... 966&sr=8-3

We have Resmed Air Sense 10s Autoset and are in New England.

Thanks for any replies.
I do not think the Jackery units will work well for you. First, they are not designed for full-time always-on (so that it would just switch over in the midst of a power outage), and second (and related) they are Li-Ion based. Li-Ion batteries *hate* being fully charged all the time. They will be useless within a few years at best, and then you will need a new unit, or a replacement battery pack, at significant expense again.

I feel similarly about the MAXOAK unit, especially after looking at the user manual. There is a disclaimer in it that you should re-charge the unit to 60% before storing ... that is precisely for the reason I've been mentioning.

Although Li-Ion is more 'current' if you will (Zonker, that was NOT a pun!) there's a reason people here press for lead acid batteries. To my knowledge there is no Li-Ion based device that solves this issue ... to do so it would have to manage the charge level and keep it below around 80%. While that is certainly possible, it then means that you get less runtime for the same investment in battery.

Bummer, I know...

_________________
MachineMask
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.

Woody
Posts: 591
Joined: Sat Mar 10, 2007 7:59 pm

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by Woody » Wed Jul 15, 2020 5:18 pm

The simple 12v lead acid battery may not be sutiable either as I
think your Cpap needs 24v dc. But a lead acid batery and a cheap
dc power inverter would do the job.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: SleapyHead software

User avatar
khauser
Posts: 1308
Joined: Fri Feb 01, 2013 9:31 am
Location: Nashua, NH

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by khauser » Wed Jul 15, 2020 5:20 pm

Woody wrote:
Wed Jul 15, 2020 5:18 pm
The simple 12v lead acid battery may not be sutiable either as I
think your Cpap needs 24v dc. But a lead acid batery and a cheap
dc power inverter would do the job.
The ResMed 12v power cable takes care of the voltage conversion. It's designed for use in a vehicle.

_________________
MachineMask
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.

User avatar
wm_hess
Posts: 334
Joined: Sat Oct 16, 2010 2:39 am

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by wm_hess » Wed Jul 15, 2020 11:29 pm

gardengirl77 wrote:
Wed Jul 15, 2020 3:10 pm
Thank you for the reply. I am afraid that suggestion is a bit beyond my level of technical ability. Maybe my husband can do what you recommend.
Hi gardengirl77

Being extremely charitable, I'm not comfortable or experienced with this type of stuff either. If you or your husband reads this thread, viewtopic/t114012/Choosing-a-Battery.ht ... =capn+loki, it talks very clearly about exactly what is needed and provides links as well. It's written by Cap'n Loki who is our resident expert on battery power.

For a total of around 150$ (2017 pricing) he has a pretty complete battery backup system. That beats the $250 Jackery system (smallest unit/smallest wattage), or the $300 maxoak options.

-Bill

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: CMS-50I Wrist Pulse Oximeter, SP02 Review, Sleepyhead

1.41
Posts: 7
Joined: Mon Feb 27, 2012 7:15 pm

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by 1.41 » Sat Jul 18, 2020 12:49 am

I've read through the linked post and I'm afraid I'm still not clear as to the setup that's recommended. Would it be possible to post pictures of the actual set up with an explanation of what's what? I understand the idea of a scooter battery for example, I understand the idea of a battery charger, but I cannot understand what the CPAP machine would plug into, or how to connect it to the battery. Any help in that regard would be most appreciated.

_________________
MachineMask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by palerider » Sat Jul 18, 2020 5:42 am

1.41 wrote:
Sat Jul 18, 2020 12:49 am
I've read through the linked post and I'm afraid I'm still not clear as to the setup that's recommended. Would it be possible to post pictures of the actual set up with an explanation of what's what? I understand the idea of a scooter battery for example, I understand the idea of a battery charger, but I cannot understand what the CPAP machine would plug into, or how to connect it to the battery. Any help in that regard would be most appreciated.
Look at the cpap 12v adapter for your machine, they come with a standard '12v plug' (used to be called a cigarette lighter plug) and clips, so you can just clip it onto the battery if you don't do a more permanent installation.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
khauser
Posts: 1308
Joined: Fri Feb 01, 2013 9:31 am
Location: Nashua, NH

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by khauser » Sat Jul 18, 2020 10:48 am

To add to palerider's answer, look at the resmed power adapter: https://www.cpap.com/productpage/resmed ... 0-machines. You can see that one end has a DC power outlet plug (aka cigarette lighter plug) and the other end plugs in to your cpap. The box in the middle converts the 12v dc to what the machine needs (24v dc).

And as palerider pointed out below, it comes with battery clips and a power outlet.

Does that help?

_________________
MachineMask
Last edited by khauser on Sat Jul 18, 2020 6:56 pm, edited 1 time in total.
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by palerider » Sat Jul 18, 2020 5:43 pm

khauser wrote:
Sat Jul 18, 2020 10:48 am
To add to palerider's answer, look at the resmed power adapter: https://www.cpap.com/productpage/resmed ... 0-machines. You can see that one end has a DC power outlet plug (aka cigarette lighter plug) and the other end plugs in to your cpap. The box in the middle converts the 12v dc to what the machine needs (24v dc).

So you'll need a dc power outlet for the battery ... https://www.amazon.com/dp/B075KDB7Y5/re ... eFbW5GH361 is a good example. I think the fuse is mandatory for safety.

Does that help?
Look again, it comes with alligator clips and a dc power socket.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
khauser
Posts: 1308
Joined: Fri Feb 01, 2013 9:31 am
Location: Nashua, NH

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by khauser » Sat Jul 18, 2020 6:54 pm

palerider wrote:
Sat Jul 18, 2020 5:43 pm
Look again, it comes with alligator clips and a dc power socket.
Indeed it does! I will edit my reply and remove the second part. Thx

_________________
MachineMask
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.

1.41
Posts: 7
Joined: Mon Feb 27, 2012 7:15 pm

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by 1.41 » Tue Jul 21, 2020 4:45 pm

Gentlemen: Thank you very much indeed for you assistance with this. It's exactly the information I was looking for! A simple solution to a vexing problem, but one needs to know where to look. Thanks again!

_________________
MachineMask

gardengirl77
Posts: 3
Joined: Sun Jul 12, 2020 12:46 pm

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by gardengirl77 » Sun Aug 09, 2020 2:36 pm

Thank you all for the great advice! We really appreciate it!!

We are wondering what you think of the following options for 2 CPAP machines; about $110 - $120 for each machine for a total of about $220 - $240:

Batteries (2 pack of 35AH, 12 V; $130): https://smile.amazon.com/dp/B07VFT6BQT/ ... _lig_dp_it

Car Charger DC Adapter for ResMed Air 10 (2x$33=$66): https://smile.amazon.com/dp/B07P6H994Q/ ... _lig_dp_it

Option1 (shorter wire Ring Terminal; $11) Female Cigarette Lighter 1.6FT Outlet Ring Terminal Plug Power Supply Cord https://smile.amazon.com/dp/B07DH8D4FR/ ... _lig_dp_it

Option 2 (longer wire with Battery Clamp; 2x10=$20) 2V/ 24V Extension Cord Plug Socket with Battery Clamp 4.9FT https://smile.amazon.com/dp/B072MZM74S/ ... _lig_dp_it

We already have a car battery charger: https://www.amazon.com/Schumacher-SC-60 ... op?ie=UTF8

2 battery boxes 2x10=$20): https://smile.amazon.com/NOCO-HM318BKS- ... 057&sr=8-3

Thank you for any feedback!

User avatar
SleepGeek
Posts: 1055
Joined: Mon Feb 09, 2015 2:56 pm

Re: Backup battery for unexpected power outages Resmed Airsense 10

Post by SleepGeek » Sun Aug 09, 2020 8:01 pm

*WARNING* - No one has answered in 7hrs but the insecure one will now.

IMO Option1 (shorter wire Ring Terminal) should be long enuff. Don't forget the cpap cord will give you additional length.

Also I think the charger you already have is over kill. By all means you can use it because you already have it. What I have used are both of these for diff applications
https://www.harborfreight.com/search?q= ... %20charger

The smaller one I have seen on sale for $5 so keep an eye out if that interest you. You can use these with the alligator clips to try things out or crimp on some rings lugs.

I have used mine for almost 10 yrs like a UPS with the charger always plugged in - while using DC to power my cpap. Who wants to wake up in the middle of the nite to fuss around with a plug then 'try' to get back to sleep? Not I.
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: What exactly do you think you're adding to this thread?