Resmed 11 wakes me up with strong pressure and leaks

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
nkbjct
Posts: 8
Joined: Tue May 02, 2023 1:55 pm

Resmed 11 wakes me up with strong pressure and leaks

Post by nkbjct » Tue May 02, 2023 8:30 pm

Hello everyone, hope you can shed some lights on my case. I was waked up by my Resmed 11 a couple of times a night, and each time when I waked up by it, I found the air pressure is high (14), and there is some leaks around my mask. AHI is 6.

If I setup the max pressure lower (12), above issue is almost gone, but my AHI goes up (10).

Any advice will be greatly appreciated, especially on how to prevent Resmed from waking me up during night with high pressure.

Jon

User avatar
Julie
Posts: 20019
Joined: Tue Feb 28, 2006 12:58 pm

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by Julie » Tue May 02, 2023 8:41 pm

Where is your minimum pressure set at?

nkbjct
Posts: 8
Joined: Tue May 02, 2023 1:55 pm

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by nkbjct » Tue May 02, 2023 9:11 pm

I set min pressure at 5.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by palerider » Tue May 02, 2023 9:22 pm

nkbjct wrote:
Tue May 02, 2023 9:11 pm
I set min pressure at 5.
Without any data, I'd guess your min should be 10, maybe more.

PS, 14 is far from 'strong', by lowering your max, you're preventing the machine from trying to stop your apneas, which is why your AHI is so bad.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

I've Been Tired
Posts: 56
Joined: Mon Feb 17, 2014 7:04 pm

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by I've Been Tired » Wed May 03, 2023 12:20 pm

a leak strap will seriously reduce your leaks
Max Pressure 14
Min Pressure 7

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by palerider » Wed May 03, 2023 12:49 pm

I've Been Tired wrote:
Wed May 03, 2023 12:20 pm
a leak strap will seriously reduce your leaks
Whatever a "leak strap" is, it probably won't.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

nkbjct
Posts: 8
Joined: Tue May 02, 2023 1:55 pm

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by nkbjct » Wed May 03, 2023 2:21 pm

As suggested, last night I turned up min pressure from 5 to 10, and max from 14 to 16, and attached is my Oscar diagram, not pretty. (somehow this board won't let me attach the diagram)

Maybe I should further change the min and max?

User avatar
Pugsy
Posts: 64926
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by Pugsy » Wed May 03, 2023 2:31 pm

nkbjct wrote:
Wed May 03, 2023 2:21 pm
(somehow this board won't let me attach the diagram)
Might help to read this sticky.

viewtopic/t183806/Attachment-storage-is ... MENTS.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Eddy
Posts: 34
Joined: Fri Feb 12, 2016 9:50 pm
Location: Richardson TX

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by Eddy » Wed May 03, 2023 3:37 pm

I have the same setup with recent upgrade to Airsense 10 to Aisense 11. I use RemZzzs Full Face Cpap Mask Liners to reduce the leaks. They are effective\ for me, but costly. When set a 5 min and 18 high ( recommended by new pulmonologist) I have leak issues. My median pressure is 12 and always used on Aircurve 10. If I set 12 min and 18 high leaks are minimum and my AHI is less than 2. At 5 -18 my AHI varies from 5 - 9 with excessive leaks. With Oscar you can see we what your median pressure looks like using the dotted line “med” … I then set my min at that or close to it. The constant upsurge to prevent the OA/CA tends to induce leaks … perhaps I move around more because of the constantly changing pressure. Pressure surges fro 5 to 12.. 15 the recedes back toward 5 then ramps up repeatedly. Set at 12 the pressure swings are only ~3 vs 10. I would give a higher min pressure a try… perhaps “sneekup” on it 8, then 10, then 12. Also try the Pressure relief of 1,2, or 3. You can change settings yourself.

I have observed an issue with the Airsense 11 that I had not seen with the 10. The F20 mask valve tends to open and shut with every breath when the min pressure is below 8-10… make a noise and makes the mask move up/down on my face. Seems a function of the valve on the f20 mask since I never had the Aircurve 10 at at pressure level lower than 10

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Jan 17, 2016 Split Sleep study results w/o CPAP AHI=83, W/CPAP AHI=29. Pressure in APAP 15-20 cm
The only way to keep your health is to eat what you don't want, drink what you don't like, and do what you'd rather not. Mark Twain

Was Mark Twain a CPAP user?

nkbjct
Posts: 8
Joined: Tue May 02, 2023 1:55 pm

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by nkbjct » Wed May 03, 2023 7:03 pm

very helpful information, and thanks. I have already set at 3 for inhale/exhale difference, and believe the same that,upsurge of air pressure, not constant high pressure, are likely causing mask leaks, which in turn causes even higher pressure. Will bump up minimum pressure from 10 to 12, max at 18 (my median aire pressure from Oscar map is 8), and see how it works.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by palerider » Wed May 03, 2023 7:55 pm

Eddy wrote:
Wed May 03, 2023 3:37 pm
I have the same setup with recent upgrade to Airsense 10 to Aisense 11. I use RemZzzs Full Face Cpap Mask Liners to reduce the leaks. They are effective\ for me, but costly.
What the sellers don't tell you is that you can just chuck 'em in the wash after use, and reuse them over and over and over.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by palerider » Wed May 03, 2023 7:57 pm

nkbjct wrote:
Wed May 03, 2023 7:03 pm
very helpful information, and thanks. I have already set at 3 for inhale/exhale difference, and believe the same that,upsurge of air pressure, not constant high pressure, are likely causing mask leaks, which in turn causes even higher pressure.
I don't know where you got the idea that leaks cause pressure to increase, because that is absolutely false.
Leaks don't affect the pressure the machine delivers until they get absolutely ludicrously large, at which point the machines will *REDUCE* the set pressure to try and get the mask to seal.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

I've Been Tired
Posts: 56
Joined: Mon Feb 17, 2014 7:04 pm

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by I've Been Tired » Wed May 03, 2023 8:12 pm

palerider wrote:
Wed May 03, 2023 12:49 pm
I've Been Tired wrote:
Wed May 03, 2023 12:20 pm
a leak strap will seriously reduce your leaks
Whatever a "leak strap" is, it probably won't.
A leak strap really reduced my leaks.

https://www.padacheek.com/anti-leak-and ... ity-straps
Max Pressure 14
Min Pressure 7

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by palerider » Wed May 03, 2023 10:22 pm

I've Been Tired wrote:
Wed May 03, 2023 8:12 pm
palerider wrote:
Wed May 03, 2023 12:49 pm
I've Been Tired wrote:
Wed May 03, 2023 12:20 pm
a leak strap will seriously reduce your leaks
Whatever a "leak strap" is, it probably won't.
A leak strap really reduced my leaks.

https://www.padacheek.com/anti-leak-and ... ity-straps
Well, if you'd SAID that you were talking about something Karen made instead of a generic "chin strap", then I'd probably said you might just be right.

She makes good stuff.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

I've Been Tired
Posts: 56
Joined: Mon Feb 17, 2014 7:04 pm

Re: Resmed 11 wakes me up with strong pressure and leaks

Post by I've Been Tired » Wed May 03, 2023 10:33 pm

palerider wrote:
Wed May 03, 2023 10:22 pm
I've Been Tired wrote:
Wed May 03, 2023 8:12 pm
palerider wrote:
Wed May 03, 2023 12:49 pm
I've Been Tired wrote:
Wed May 03, 2023 12:20 pm
a leak strap will seriously reduce your leaks
Whatever a "leak strap" is, it probably won't.
A leak strap really reduced my leaks.

https://www.padacheek.com/anti-leak-and ... ity-straps
Well, if you'd SAID that you were talking about something Karen made instead of a generic "chin strap", then I'd probably said you might just be right.

She makes good stuff.
You have 32,000 posts on here and you don't know what a leak strap is?!?! Good lord.
Max Pressure 14
Min Pressure 7