F&P Evora Full Face Mask

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
ajsmith
Posts: 5
Joined: Fri May 31, 2024 1:17 pm

F&P Evora Full Face Mask

Post by ajsmith » Sun Jun 23, 2024 5:20 pm

I got the mask last week and had zero leaks each day. I washed it this weekend and now it is leaking badly. I have read the manual and watched videos on it. I will try to contact the medical device store tomorrow but often they don't answer the phone. I will use the nasal mask which leaks terribly and keeps me awake. Any suggestions? I find this mask extremely hard to put on and take off - and now leaks!

User avatar
ChicagoGranny
Posts: 15078
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: F&P Evora Full Face Mask

Post by ChicagoGranny » Sun Jun 23, 2024 6:12 pm

ajsmith wrote:
Sun Jun 23, 2024 5:20 pm
I washed it this weekend and now it is leaking badly.
Have you been able to identify the position of the leak?

When you washed the mask, did you damage the Anti-Asphyxia Valve?
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
vandownbytheriver
Posts: 542
Joined: Sat Feb 03, 2024 11:42 pm

Re: F&P Evora Full Face Mask

Post by vandownbytheriver » Sun Jun 23, 2024 10:28 pm

The Evora FF is what's known as a 'hybrid' mask, nasal with mouth covering. I find if I undo both side hooks, put the mask under my nose, then pull the headgear over, I can hook the side pieces and be going.

Just fit the mask without the headgear, you should be able to find the leak while holding the cushion to your face.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion)
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape.

ajsmith
Posts: 5
Joined: Fri May 31, 2024 1:17 pm

Re: F&P Evora Full Face Mask

Post by ajsmith » Mon Jun 24, 2024 11:58 am

ChicagoGranny wrote:
Sun Jun 23, 2024 6:12 pm
ajsmith wrote:
Sun Jun 23, 2024 5:20 pm
I washed it this weekend and now it is leaking badly.
Have you been able to identify the position of the leak? No

When you washed the mask, did you damage the Anti-Asphyxia Valve?
Tell me more about this valve, please. It might be the problem.

User avatar
vandownbytheriver
Posts: 542
Joined: Sat Feb 03, 2024 11:42 pm

Re: F&P Evora Full Face Mask

Post by vandownbytheriver » Mon Jun 24, 2024 12:37 pm

ajsmith wrote:
Mon Jun 24, 2024 11:58 am
Tell me more about this valve, please. It might be the problem.
I don't believe the Evora FF has this valve. The Vitera has one.

https://www.fphcare.com/us/homecare/sle ... vora-full/

Examine where the small tube snaps onto the mask spigot.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion)
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape.

User avatar
Okie bipap
Posts: 3566
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: F&P Evora Full Face Mask

Post by Okie bipap » Mon Jun 24, 2024 1:58 pm

There is flexible flap located in the mask frame right above where the hose connects. This is the antiaphyxiation valve.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
ChicagoGranny
Posts: 15078
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: F&P Evora Full Face Mask

Post by ChicagoGranny » Mon Jun 24, 2024 2:36 pm

Okie bipap wrote:
Mon Jun 24, 2024 1:58 pm
There is flexible flap located in the mask frame right above where the hose connects. This is the antiaphyxiation valve.
Yep.
ajsmith wrote:
Mon Jun 24, 2024 11:58 am
Tell me more about this valve, please. It might be the problem.
The flap needs to close when the circuit is pressurized. When the machine is not running, the flap opens to prevent CO2 buildup in the mask.
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
ChicagoGranny
Posts: 15078
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: F&P Evora Full Face Mask

Post by ChicagoGranny » Mon Jun 24, 2024 2:59 pm

vandownbytheriver wrote:
Mon Jun 24, 2024 12:37 pm
I don't believe the Evora FF has this valve.
Members don't have to comment when they don't know what they are talking about.
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
vandownbytheriver
Posts: 542
Joined: Sat Feb 03, 2024 11:42 pm

Re: F&P Evora Full Face Mask

Post by vandownbytheriver » Mon Jun 24, 2024 10:00 pm

ChicagoGranny wrote:
Mon Jun 24, 2024 2:59 pm
vandownbytheriver wrote:
Mon Jun 24, 2024 12:37 pm
I don't believe the Evora FF has this valve.
Members don't have to comment when they don't know what they are talking about.
Well, made me look! Yep, mine has one. It's hidden behind the oval hose connection.

Thanks for your gentle correction.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion)
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape.

ajsmith
Posts: 5
Joined: Fri May 31, 2024 1:17 pm

Re: F&P Evora Full Face Mask

Post by ajsmith » Thu Jun 27, 2024 7:43 am

I fixed it. It was the little flexible anti-asphyxiation valve that is located between the mask and the tube coming out of the mask. Somehow it had gotten misaligned. No leaks now but, for me, an uncomfortable mask.