I have been using an AirSense 10 CPAP, Pressure 9 for the past seven years with excellent results. AHI events have consistently averaged below 2, and leakage usually at or near zero. This system was recently replaced with a new AirSense 11 with the same settings. My AHI event readings on the AirSense 11 have consistently been much higher, averaging around 8 and range from 6-12. I am using the same mask, Philips Nuance Pro on both systems. If I switch back to the old AirSense 10, my numbers drop back to AHI less than 2, and no leakage.
My DME swapped out the first new Airsense 11 for a new one. The AHI events averaged 8 on both AirSense11 machines, but the leakage decreased from an average of 15 to 10.
Working with my DME and my doctor, we have tried adjusting the pressure setting, as well as humidity and hose temperature, but these changes have not helped. My doctor said that he had another patient with the same issue, and ended up replacing the AirSense 11 with a new AirSense 10 to solve it.
Has anyone else experienced this and found a cause or solution?
Higher AHI Readings on AirSense 11 than AirSense 10
Re: Higher AHI Readings on AirSense 11 than AirSense 10
Without looking at the daily detailed data, we (and your doctors and the DME) are just guessing about what might be happening.
It would be worth uploading data to SleepHQ from both machines and putting the public "share" links to the data over here. Or importing data from both machines into OSCAR and taking some screen shots of the daily data.
Because without that daily data, all we can do is speculate.
It would be worth uploading data to SleepHQ from both machines and putting the public "share" links to the data over here. Or importing data from both machines into OSCAR and taking some screen shots of the daily data.
Because without that daily data, all we can do is speculate.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Also use a P10 mask |
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Re: Higher AHI Readings on AirSense 11 than AirSense 10
Good advice. I used SleepHQ and posted results. I was able to dial in my therapy and get AHI down under .5 AHI pretty consistently. Usually around .2 AHI.
I found SleepHQ somewhat easier to use than Oscar. When I had my Airsense 10, I did use Oscar and posted to this board. I did have improved results using that method too. The setting tweaks were much better than what I got from my original sleep doctor/report.
I found SleepHQ somewhat easier to use than Oscar. When I had my Airsense 10, I did use Oscar and posted to this board. I did have improved results using that method too. The setting tweaks were much better than what I got from my original sleep doctor/report.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset |
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto