Hi,
Asking for my partner, new to CPAP, bought one without diagnosis.
We set the pressure at 6.6 and it worked great, he says he felt great after 4hrs night. However I worry about the CA’s. Since we don’t know the type of apnea he has, is the fixed pressure good or making it potentially worse, and what level is safe for ‘experimenting’?
Can you suggest something based on the OSCAR data?
Thank you!
new user- help with adjusting pressure
new user- help with adjusting pressure
- Attachments
-
- Image 23-05-2025 at 23.40.jpeg (1013.79 KiB) Viewed 7904 times
Last edited by zuzijanka on Sat May 24, 2025 11:29 am, edited 1 time in total.
Re: new user- help with adjusting pressure
Before worrying about the centrals seen you need to determine if they are real asleep centrals or post arousal centrals or maybe just a normal sleep transition central.
Zoom in in a 4 minute time frame on that little group of centrals happening just prior to the 3:20 time mark.
So that it looks like these examples here.
viewtopic.php?f=1&t=187767&p=1451526#p1451526
Zoom in in a 4 minute time frame on that little group of centrals happening just prior to the 3:20 time mark.
So that it looks like these examples here.
viewtopic.php?f=1&t=187767&p=1451526#p1451526
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: new user- help with adjusting pressure
Thank you! It looks like the centrals appear after some breath disturbance, like in the screenshots.Pugsy wrote: ↑Sat May 24, 2025 7:59 amBefore worrying about the centrals seen you need to determine if they are real asleep centrals or post arousal centrals or maybe just a normal sleep transition central.
Zoom in in a 4 minute time frame on that little group of centrals happening just prior to the 3:20 time mark.
]
Does it mean that it should be safe to up the pressure?
- Attachments
-
- Screenshot 2025-05-25 at 14.23.06.png (450.39 KiB) Viewed 5706 times
-
- Screenshot 2025-05-25 at 14.34.15.png (450.63 KiB) Viewed 5706 times
-
- Screenshot 2025-05-25 at 14.14.50.png (62.2 KiB) Viewed 5706 times
Re: new user- help with adjusting pressure
Why were you wanting to increase the pressure anyway?
More pressure won't help with the centrals and you really aren't having enough of the OAs or hyponeas to warrant more pressure.