Aerophagia waking me up every day

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Promsaco
Posts: 5
Joined: Tue Feb 17, 2026 10:26 pm

Aerophagia waking me up every day

Post by Promsaco » Tue Feb 17, 2026 11:31 pm

Good morning,

I have been using my BMC G2S A20 machine for a few months now, and I still can't get the right setting. My main problem is AEROPHAGIA -I wake up every day around 5am like a fully bloated balloon. It's incredibly uncomfortable and even painful.

My sleep is generally quite uneven: I wake up many times, turning around, readjusting my mask. I fall asleep on my side, but I guess I must end up on my back too and flip back to the side. I sleep better than when I didn't use the machine at all, but I can't say it's a fully restful night sleep.

I am using a PHILIPS RESPIRONICS full face mask (under nose), because the nasal mask wasn't working for me. I have read about aerophagia, tried different settings but nothing seems to work.

>> I'd really appreciate assistance telling me what's wrong with my current settings. Thank you in advance!
Attachments
CPAP Statistics.jpg
Includes usage details and settings changes
CPAP Statistics.jpg (332.12 KiB) Viewed 251341 times
CPAP General info.jpg
Includes: Events, flow rate, pressure, leak
CPAP General info.jpg (563.02 KiB) Viewed 251341 times

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Aerophagia waking me up every day

Post by zonker » Wed Feb 18, 2026 5:48 pm

Promsaco wrote: ↑
Tue Feb 17, 2026 11:31 pm


>> I'd really appreciate assistance telling me what's wrong with my current settings. Thank you in advance!
it's those settings that are causing YOU to have that dreaded aerophagia. and of course that's unwanted and interferes with your therapy. not everyone gets aerophagia. it's folk like you and me that have to get past it.

I know nothing about your machine. but it seems you may already know how to change your settings? if so maybe we can get this fixed. the first thing you want to do is lower your upper pressure. unfortunately, do the short term, it means you won't be getting proper therapy. but you aren't getting that now, so it may be worth the struggle.

what *I* did with my resmed machine was I kept lowering my upper pressure until my aerophagia went away. and yes that made my ahi jump up horribly. but once I found that spot (it'll be different for each of us.) I started slowly increasing that upper pressure by tiny increments. in the case of a resmed machine, I could raise mine minutely by .2 at a time. having left that at that settings for a couple of nights, i'd proceed to increase by another .2. this method requires patience as it's a very long process. I mean for me, weeks and weeks.

I have no idea what your machine increments are. I suspect it's .5 judging from you chart.

but my method worked for me. I dunno if it will work for you.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Promsaco
Posts: 5
Joined: Tue Feb 17, 2026 10:26 pm

Re: Aerophagia waking me up every day

Post by Promsaco » Thu Feb 19, 2026 11:42 am

Thank you so much for taking the time to reply. I am definitely going to try your method tonight. My machine moves in 0.5 increments, so I'll take it down to 9 instead of 9.5 and see what happens. I'll report back in a couple days. Thanks again.

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Aerophagia waking me up every day

Post by zonker » Thu Feb 19, 2026 4:42 pm

Promsaco wrote: ↑
Thu Feb 19, 2026 11:42 am
Thank you so much for taking the time to reply. I am definitely going to try your method tonight. My machine moves in 0.5 increments, so I'll take it down to 9 instead of 9.5 and see what happens. I'll report back in a couple days. Thanks again.
pls DO come back. I'm always curious to see if what works for me will work for others.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Promsaco
Posts: 5
Joined: Tue Feb 17, 2026 10:26 pm

Re: Aerophagia waking me up every day

Post by Promsaco » Thu Feb 26, 2026 9:30 pm

zonker wrote: ↑
Thu Feb 19, 2026 4:42 pm
Promsaco wrote: ↑
Thu Feb 19, 2026 11:42 am
Thank you so much for taking the time to reply. I am definitely going to try your method tonight. My machine moves in 0.5 increments, so I'll take it down to 9 instead of 9.5 and see what happens. I'll report back in a couple days. Thanks again.
pls DO come back. I'm always curious to see if what works for me will work for others.

good luck!
It's been a week and so far I haven't seen much improvement. I've taken the mask off a couple times while sleeping, I guess something was bothering me. I will lower it 0.5 more and see. Will keep you posted.

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Re: Aerophagia waking me up every day

Post by ozij » Fri Feb 27, 2026 4:52 am

Promsaco wrote: ↑
Thu Feb 26, 2026 9:30 pm
zonker wrote: ↑
Thu Feb 19, 2026 4:42 pm
Promsaco wrote: ↑
Thu Feb 19, 2026 11:42 am
Thank you so much for taking the time to reply. I am definitely going to try your method tonight. My machine moves in 0.5 increments, so I'll take it down to 9 instead of 9.5 and see what happens. I'll report back in a couple days. Thanks again.
pls DO come back. I'm always curious to see if what works for me will work for others.

good luck!
It's been a week and so far I haven't seen much improvement. I've taken the mask off a couple times while sleeping, I guess something was bothering me. I will lower it 0.5 more and see. Will keep you posted.
Don't reduce gradually from where your suffeing, start low, and rasise gradually. Your esophageal sphincter is opening and letting air in. Continued irrtation won't do it much good. Zonker point was to go gradually up from where thing are tolerable.

Can you explain why you made the changes you made, as seen in your overview chart?

Edited to remove an extra word

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Aerophagia waking me up every day

Post by zonker » Fri Feb 27, 2026 9:09 pm

ozij wrote: ↑
Fri Feb 27, 2026 4:52 am


Don't reduce gradually from where your suffeing, start low, and rasise gradually. Your esophageal sphincter is opening and letting air in. Continued irrtation won't do it much good. Zonker point was to go gradually up from where thing are tolerable.
OH! I scrolled up and I see I didn't make that clear. thanks for pointing that out.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Promsaco
Posts: 5
Joined: Tue Feb 17, 2026 10:26 pm

Re: Aerophagia waking me up every day

Post by Promsaco » Sun Mar 01, 2026 6:44 am

Understood. I am lowering my Max APAP to 6.5 and Min APAP to 6.0. I think this is as low as I should go. Let's see tomorrow!

I really hope I can fix it bc, as you know, waking up like a balloon about to burst is not fun.

Thanks guys, I appreciate your time.

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Aerophagia waking me up every day

Post by zonker » Sun Mar 01, 2026 1:45 pm

Promsaco wrote: ↑
Sun Mar 01, 2026 6:44 am


I really hope I can fix it bc, as you know, waking up like a balloon about to burst is not fun.
yeah it's my second favorite thing to do.

(the first favorite thing is to hurl a bowling ball into the air and catch it with my teeth.)
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Promsaco
Posts: 5
Joined: Tue Feb 17, 2026 10:26 pm

Re: Aerophagia waking me up every day

Post by Promsaco » Tue Mar 03, 2026 10:18 pm

ozij wrote: ↑
Fri Feb 27, 2026 4:52 am
Promsaco wrote: ↑
Thu Feb 26, 2026 9:30 pm
zonker wrote: ↑
Thu Feb 19, 2026 4:42 pm
Promsaco wrote: ↑
Thu Feb 19, 2026 11:42 am
Thank you so much for taking the time to reply. I am definitely going to try your method tonight. My machine moves in 0.5 increments, so I'll take it down to 9 instead of 9.5 and see what happens. I'll report back in a couple days. Thanks again.
pls DO come back. I'm always curious to see if what works for me will work for others.

good luck!
It's been a week and so far I haven't seen much improvement. I've taken the mask off a couple times while sleeping, I guess something was bothering me. I will lower it 0.5 more and see. Will keep you posted.
Don't reduce gradually from where your suffeing, start low, and rasise gradually. Your esophageal sphincter is opening and letting air in. Continued irrtation won't do it much good. Zonker point was to go gradually up from where thing are tolerable.

Can you explain why you made the changes you made, as seen in your overview chart?

Edited to remove an extra word
Good morning -after a not so good night. AHI is 5.4 this morning and I feel like taking a nap again -like you warned me @zonker. Max APAP was set at 6.5 and Min at 6.0. I will keep it like this one more night, just to make sure the high AHI is not due to something else. If I get the same read tmr I will raise it to 7. Will keep you posted. Have a nice day!

@ozij > I was just tweaking the settings on my own, based on what I read here and there and replies from chatgpt.

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Aerophagia waking me up every day

Post by zonker » Tue Mar 03, 2026 11:44 pm

Promsaco wrote: ↑
Tue Mar 03, 2026 10:18 pm
Will keep you posted. Have a nice day!
will keep my fingers crossed.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg