Hello, I had been using my Resmed CPAP MAchine S9 series for 2 weeks and its great since its so quiet. My only problem is that the smartstart/stop was disable for some reasons I do not know. Before once I wear my mask the machine will start immediately but now I am doing manual by pressing the start/stop button. Is there anybody who can help me to enable the smartstop/start.
Thank you.
Thank you.
Re: Smart Start/Stop
Access the clinical menu (Press the setup button and the rotating button for 3 seconds) In there you can set Autostart to either Patient or On.
Resmed AutoSet Spirit II flow generator (Backup)
Resmed H3i Humidifier
Swift FX
Mirage Swift LT Nasal Pillows
Activa Nasal Mask
Resscan Software with USB + Data Card
SPO 7500 Pulse-Ox
MedicMon Blood Pressure Monitor
Aussie Heated Hose
Resmed H3i Humidifier
Swift FX
Mirage Swift LT Nasal Pillows
Activa Nasal Mask
Resscan Software with USB + Data Card
SPO 7500 Pulse-Ox
MedicMon Blood Pressure Monitor
Aussie Heated Hose
Re: Smart Start/Stop
The MASK ALERT will not alarm with the auto start/stop enabled. Personally, I'd rather the mask alert was working and pushing on/off is not a big deal for me as I use the mask fit feature every night.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.
Never, never, never, never say never.
Re: Smart Start/Stop
There is no such thing as mask alert on the S9.
Resmed AutoSet Spirit II flow generator (Backup)
Resmed H3i Humidifier
Swift FX
Mirage Swift LT Nasal Pillows
Activa Nasal Mask
Resscan Software with USB + Data Card
SPO 7500 Pulse-Ox
MedicMon Blood Pressure Monitor
Aussie Heated Hose
Resmed H3i Humidifier
Swift FX
Mirage Swift LT Nasal Pillows
Activa Nasal Mask
Resscan Software with USB + Data Card
SPO 7500 Pulse-Ox
MedicMon Blood Pressure Monitor
Aussie Heated Hose
Re: Smart Start/Stop
Are you using a full face mask?glen0901 wrote:Hello, I had been using my Resmed CPAP MAchine S9 series for 2 weeks and its great since its so quiet. My only problem is that the smartstart/stop was disable for some reasons I do not know. Before once I wear my mask the machine will start immediately but now I am doing manual by pressing the start/stop button. Is there anybody who can help me to enable the smartstop/start.
Thank you.
I don't know about the new S9 machines, but in the previous models, if the mask selector was set to MIR FULL, it disabled the smartstart feature.
From the troubleshooting section of the S8 Auto manual:
Use of a ResMed Full Facemask.
SmartStart does not work with a full-face mask as the
anti-asphyxia valve will not allow sufficiently high
pressure on exhalation.
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: Smart Start/Stop
That's right Den - S9 is the same as the S8.
Resmed AutoSet Spirit II flow generator (Backup)
Resmed H3i Humidifier
Swift FX
Mirage Swift LT Nasal Pillows
Activa Nasal Mask
Resscan Software with USB + Data Card
SPO 7500 Pulse-Ox
MedicMon Blood Pressure Monitor
Aussie Heated Hose
Resmed H3i Humidifier
Swift FX
Mirage Swift LT Nasal Pillows
Activa Nasal Mask
Resscan Software with USB + Data Card
SPO 7500 Pulse-Ox
MedicMon Blood Pressure Monitor
Aussie Heated Hose
Re: Smart Start/Stop
NO? Wow, mine's gone off several times when something was wrong -- that's a miss in my mind.bigk wrote:There is no such thing as mask alert on the S9.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.
Never, never, never, never say never.
Re: Smart Start/Stop
I found all the answers regarding my ResMed S9 Elite very helpful. Thank you to all who posted answers here.
Re: Smart Start/Stop
Thank you sooooo much! I pressed setting and held button down for 3 sec and it unlocked the specs. I found autostart and now all good. Thanks again.