Cold Nose on CPAP
Cold Nose on CPAP
Hi everyone,
I started CPAP on Friday night. I did great on Saturday night, slept through the night, though I woke up once.
Last night (Sunday night), I only made it until 1am. I can't remember every reason I had trouble. There was more mask leaking, I think because I washed out the mask and headgear in the shower and the parts of the mask got pulled apart. I thought it was dry and properly assembled when I put it on, but maybe not. So I kept having leaks, but finally must have fallen asleep.
I woke up feeling the need to take it off. The thing I most noticed was that my nose was very cold. The area surrounding my nose was cold too but it was mostly my nose.
Today (Monday) I took a good long nap - lucky that I was able to do that - and I did well. I slept well for maybe three or four hours. So that is promising. My nose was cold when I took off the mask, but not as cold and not difficult to tolerate.
Does anyone know what would cause that? I do have a humidifier. I set it at 0.5, because the DME rep told me that - ironically - people tend to have problems with dryness when they turn up the heat on the humidifier, as turning up the heat is drying.
I did fine at 0.5 on Saturday night. And I think I may have actually turned it up to 1.0 on Sunday night, the night when my nose was too cold. It was back down to 0.5 today when I napped.
Thanks for your help!
Janet
I started CPAP on Friday night. I did great on Saturday night, slept through the night, though I woke up once.
Last night (Sunday night), I only made it until 1am. I can't remember every reason I had trouble. There was more mask leaking, I think because I washed out the mask and headgear in the shower and the parts of the mask got pulled apart. I thought it was dry and properly assembled when I put it on, but maybe not. So I kept having leaks, but finally must have fallen asleep.
I woke up feeling the need to take it off. The thing I most noticed was that my nose was very cold. The area surrounding my nose was cold too but it was mostly my nose.
Today (Monday) I took a good long nap - lucky that I was able to do that - and I did well. I slept well for maybe three or four hours. So that is promising. My nose was cold when I took off the mask, but not as cold and not difficult to tolerate.
Does anyone know what would cause that? I do have a humidifier. I set it at 0.5, because the DME rep told me that - ironically - people tend to have problems with dryness when they turn up the heat on the humidifier, as turning up the heat is drying.
I did fine at 0.5 on Saturday night. And I think I may have actually turned it up to 1.0 on Sunday night, the night when my nose was too cold. It was back down to 0.5 today when I napped.
Thanks for your help!
Janet
_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
- BlackSpinner
- Posts: 9742
- Joined: Sat Apr 25, 2009 5:44 pm
- Location: Edmonton Alberta
- Contact:
Re: Cold Nose on CPAP
Another DME employee who failed elementary school science? No - warm air can hold more moisture. Start playing with your humidifier settings, there is no reason to have a cold nose.Does anyone know what would cause that? I do have a humidifier. I set it at 0.5, because the DME rep told me that - ironically - people tend to have problems with dryness when they turn up the heat on the humidifier, as turning up the heat is drying.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Additional Comments: Quatro mask for colds & flus S8 elite for back up |
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal
Re: Cold Nose on CPAP
My humdifier has settings for 1-4. My nose was super cold during the day after using setting 2 for a week. I switched to setting 3 and it's been better for me.
_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear |
Additional Comments: Wisp Nasal Mask |
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Cold Nose on CPAP
I am so spoiled by my Climateline tubing, that if I sleep without it, I get a nearly frigid beak.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Cold Nose on CPAP
Turning up the heat if you aren't using the ClimateLine hose might be more drying (though in all honesty I don't believe what you were told is the whole truth) but it also allows for more humidity to be delivered...so it should offset each other.
Are you using the ClimateLine heated hose or a regular slimline hose?
A setting of 0.5 or 1.0 with a regular hose isn't going to add much warmth at all and yes, your nose will get cold...I call it ice cube nose. Very annoying.
If you aren't getting an option to increase the temperature separately...you may be using a regular non heated hose.
Your choices are to either increase the bedroom temp or increase the humidity setting to add more warmth hopefully to the air.
You might get enough relief with a hose cozy acting as insulation to prevent the air in the hose from cooling off so much.
You could make a hose cozy with some old socks to see if it helps. Just cut out the toes and slide over the hose and secure with something like rubber bands or even scotch tape.
If you aren't using the ClimateLine hose....get one. It keeps the air warmer all the way up to the mask hose or attachment.
You won't be sorry.
Are you using the ClimateLine heated hose or a regular slimline hose?
A setting of 0.5 or 1.0 with a regular hose isn't going to add much warmth at all and yes, your nose will get cold...I call it ice cube nose. Very annoying.
If you aren't getting an option to increase the temperature separately...you may be using a regular non heated hose.
Your choices are to either increase the bedroom temp or increase the humidity setting to add more warmth hopefully to the air.
You might get enough relief with a hose cozy acting as insulation to prevent the air in the hose from cooling off so much.
You could make a hose cozy with some old socks to see if it helps. Just cut out the toes and slide over the hose and secure with something like rubber bands or even scotch tape.
If you aren't using the ClimateLine hose....get one. It keeps the air warmer all the way up to the mask hose or attachment.
You won't be sorry.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Cold Nose on CPAP
Totally worth every dollar of it. Then you'll be able to try different combinations of temp and humidity as they can be manipulated independently.Pugsy wrote:If you aren't using the ClimateLine hose....get one. It keeps the air warmer all the way up to the mask hose or attachment.
You won't be sorry.
Family member of mine, new to cpap, was told by her DME that there was no such thing as a heated hose for her S9 -- Oy! Once I told her about it and she got one, she no longer has ice cube nose and is comfy at night.
p.s. Froggie -- you have a beak ?
_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle |
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura |
Re: Cold Nose on CPAP
Thanks, guys! I do have the slimline hose. I'll try playing with the settings. If that doesn't work, I'll try the cozy or the climateline hose.
So glad I have this forum, thanks for the advice!
Janet
So glad I have this forum, thanks for the advice!
Janet
_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Re: Cold Nose on CPAP
I'd try putting the humidifier up to 3 or 4. You have it really low right now. Higher number on humidifier = greater moisture. Will help with dryness and the cold air.
Re: Cold Nose on CPAP
Thanks. I'll try it at 3 tonight and see what happens. Already looked into buying the climateline hose if needed, and the Snugglehose fleece cover.
_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Re: Cold Nose on CPAP
Happy to say I slept fairly well last night! Had the humidity on 3 as advised. I woke at 2am - seems to be a usual thing for me. Took me a long time to get back to sleep. But other than that, I slept through the night with the machine on.
I think I will treat myself to the snugglehose, as it would simply be more comfortable when the hose gets in my way during sleep, AND it might help.
If I have more problems, I'll ask for my next hose to by a Climateline.
Thanks so much for all your help and support!!
Janet
I think I will treat myself to the snugglehose, as it would simply be more comfortable when the hose gets in my way during sleep, AND it might help.
If I have more problems, I'll ask for my next hose to by a Climateline.
Thanks so much for all your help and support!!
Janet
_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Re: Cold Nose on CPAP
Should have added - no cold nose!
_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Re: Cold Nose on CPAP
Great to hear
Re: Cold Nose on CPAP
By the way, I think give the climate hose a go too. Might as well if you have it
Re: Cold Nose on CPAP
Thanks, Sleepstar! I don't have it. I was seriously considering ordering it. But since I did well last night, I've decided to see how things work out. I'm supposed to have a hose replacement every 3 months (I'm surprised at how often they will replace this expensive equipment!).
I'm going to wait and see how I do. If I think I need the Climatehose, I'll request it. And if they say no, and I still think I need it, I'll buy it myself.
It is so great to have this forum! I'll bet it's helping many people stay in compliance and get well! All of you who are helping us are a true blessing. I just hope that when I adjust, I don't forget to stick around and pay it forward (which I know I sometimes do, inadvertently).
Janet
I'm going to wait and see how I do. If I think I need the Climatehose, I'll request it. And if they say no, and I still think I need it, I'll buy it myself.
It is so great to have this forum! I'll bet it's helping many people stay in compliance and get well! All of you who are helping us are a true blessing. I just hope that when I adjust, I don't forget to stick around and pay it forward (which I know I sometimes do, inadvertently).
Janet
_________________
Mask: Mirage™ FX For Her Nasal CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Cold Nose on CPAP
Kaiasgram: No beak, but it's been cold, stiff, and numb enough I could have believed I had a beak!
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |